(Almost) Everything You Need to Know About Apex Salinity Calibration and TC Factor

FlyingOctopi

Community Member
View Badges
Joined
Feb 25, 2018
Messages
79
Reaction score
35
Rating - 0%
0   0   0
My readings from the probe are slowly climbing even after I top off. Is this something fixed by the tc factor or is this due to air bubbles.
 

keithIHS

Active Member
View Badges
Joined
Oct 13, 2016
Messages
128
Reaction score
57
Rating - 0%
0   0   0
Could be either. Need more info. Can you create a chart of salinity and temperature and post it? Over what timeframe is it rising and how much is it rising? Is temp rising similarly to salinity? Or is it falling opposite to salinity? Are you using an auto top off? If topping off by hand, do you mean it doesnt drop back down after you add the fresh water? If by hand, did you wait 10-20 mins for the fresh to mix in and the Apex to update the data? Is your probe upside down in the tank or sump?
 

FlyingOctopi

Community Member
View Badges
Joined
Feb 25, 2018
Messages
79
Reaction score
35
Rating - 0%
0   0   0
Screenshot_20210423-094024_APEX Fusion.jpg


I'm topping off in the morning and night every day back to where the Hanna checker says it's 35ppt. I'll be setting up an auto top off soon. The probe is right side up but on about a 45 degree angle. I left the calibration solution in the tank but pulled it out to calibrate. I know now to leave both in the sump but just got new solution yesterday so I plan on recalibration this weekend.
 

powers2001

started reefing 1999
View Badges
Joined
May 2, 2013
Messages
5,812
Reaction score
5,508
Location
LIVING WELLS MONTANA
Rating - 0%
0   0   0
I have a question for everyone. What are your experiences on drying the salinity probe when removing from the sump, rinsing clean in RO water, and patting dry with paper towels? How much time are you allowing for a complete dry and how do you know you have a complete dry before submersing in calibration solution to ensure no contamination of solution in close proximity of the probe tip?
 

keithIHS

Active Member
View Badges
Joined
Oct 13, 2016
Messages
128
Reaction score
57
Rating - 0%
0   0   0
It doesn't need to be completely dry. Rinse in RODI and you're good to go (the conductivity of RODI is tiny).
The precision of the Hannah checker is 1 ppt. The accuracy is unfortunately +/-2 (I think). I'd wager though that the repeatability is much better (let me know if you need an explanation of those three terms). Your Hannah might be drifting but I doubt it. To be sure, if you have a handheld refractometer, you could make up some of Randy's calibration recipe and double check it. BUT my vote is for air bubbles. Turn it upside down and tap it a few times and leave it in there upside down for a day.
On another topic, why is your temp swinging so frequently? Is your heater close to your temp probe?
 

keithIHS

Active Member
View Badges
Joined
Oct 13, 2016
Messages
128
Reaction score
57
Rating - 0%
0   0   0
I should have said that if you look closely at your chart you might be able to convince yourself that the salinity reading goes up a tiny bit when temp goes up and vice versa. If true, this would indicate the TC is not quite right, but the amount is tiny and the correlation isn't high so I wouldn't mess w it. The important point is that salinity is drifting up but temperature is not drifting up (or down) over the same time frame, so TC is not the issue.
Regarding topping off, if you have a sump, and even better if you have a return pump compartment in your sump with an overflow baffle, make a mark on your sump at the water level of the return compartment and refill to that line. It will be easier and more consistent over the course of a week or two or maybe longer than checking salinity with refraction index.
 

keithIHS

Active Member
View Badges
Joined
Oct 13, 2016
Messages
128
Reaction score
57
Rating - 0%
0   0   0
I apologize, I didn't really answer your question about drying off the probe. I just tapped it downward against a paper towel or cloth to dislodge any drops. Have done this a number of times and seen no effect from RODI remaining in the probe. No need to dry thoroughly. Any remaining water in the tip will be negligible and if you manage to get the air out any remaining RODI will be flushed.
 

FlyingOctopi

Community Member
View Badges
Joined
Feb 25, 2018
Messages
79
Reaction score
35
Rating - 0%
0   0   0
My heater was in the tank and the temp probe is in the return pump compartment. Today I put the heater in the skimmer compartment.
 

SpiceReefer90

New Member
View Badges
Joined
Apr 15, 2021
Messages
5
Reaction score
2
Location
Dallas
Rating - 0%
0   0   0
I was trying to calibrate my new salinity probe a few weeks ago, and ran into all sorts of problems. Calibrations that were way off the mark, big fluctuations in readings that couldn't possibly be correct (and were verifiably not), etc. After searching around the internet, it was clear that a lot of folks have problems with this sensor. Some have written it off completely, and others insist that such problems are the result of calibration error. Information on correct calibration and configuration was present but scattered. Some of it was incomplete or incorrect. After having read many forum threads, spoken to Apex customer service, and experimented with different approaches, I am here to tell you that the sensor does work reliably when correctly configured and calibrated. This post is intended to capture everything I've learned in a single place, hopefully making this easier for someone else in the future.

Apex Configurables
There are only two configurable aspects of the Apex salinity sensor that can affect the reading. The first is, of course, the calibration point. The second is the temperature compensation factor (or TC factor). The salinity probe's readings are significantly affected by temperature - the TC factor setting allows you to compensate for this if you also have a temperature probe configured. The rest of this post captures the concepts and steps necessary to configure these inputs correctly.

TC Factor and Correctness
The TC factor configurable can be found on the advanced tab of the conductivity probe configuration page. It is a percentage, set manually - apparently different tanks land on different values. The percentage controls exactly how much compensation should be applied (i.e. x% compensation per degree).

One way to see if your TC factor is correct is by looking at your salinity graph with your temperature readings overlaid. If the TC factor needs to be adjusted, you will be able to see that the salinity readings track fluctuations in temperature. If the TC factor is correct(-ish), this correlation will not be present.

Here is what this looks like when TC is disabled (i.e. 0%):
Screenshot 2020-05-10 at 15.17.10.png


Here is the same system with TC configured:
Screenshot 2020-05-10 at 15.18.09.png


You can see from the first screenshot a clear correlation between temperature and salinity. Compare to the second screenshot with TC applied, and the salinity is relatively stable. I'm still not quite sure about those little bumps though. I continue to play around fine tuning the TC factor to see how they react (if at all - it's possible that the fluctuation is related to some funny business with my overflow).

Configuring the TC Factor
The TC factor defaults to 0. This is confusing for new Apex owners, as the salinity probe is all but worthless without applying temperature compensation. After speaking with customer service, and reading through some documentation, it appears that the default is 0 because setting it to anything other than zero causes problems in systems that don't have a temperature probe installed (the documentation includes a large WARNING section about this). Since Neptune cannot know or guarantee that any given system will have its temperature probe installed, the default is 0 to be safe.

If you have a temperature probe installed, you will definitely want to adjust the default TC factor. It is important to note that the temperature probe must be plugged into the same module as the salinity probe. In other words, if your salinity probe is plugged into your base unit, then you must also have a temperature probe plugged into the base unit. Similarly, if your salinity probe is plugged into a PM2 salinity module, then the temperature probe must be plugged into the same PM2.

TC Factor Value Range
It may take some experimentation to find the correct TC factor for your system. The Apex documentation states that seawater normally has a TC factor between 2.1% and 2.3%, thus recommending 2.2% as a good starting point. By comparison, a Neptune representative told me that in their experience, correct values are often found in the 1.9% to 2.1% range. Your mileage may vary.

The Apex documentation states that the salinity probe must be re-calibrated when changing the TC factor. According to a Neptune representative with whom I spoke, this is only partially true. When TC factor is non-zero and the salinity probe is configured, the current water temperature is stored inside the salinity calibration point that Apex records. This temperature is used to apply temperature compensation going forward. If the TC factor is zero when the salinity probe is configured, then the current water temperature is not stored, making future temperature compensation impossible. Therefor, in order for TC factor to take effect, it must be configured with a non-zero value at the time that the salinity probe is calibrated.

Fear not, however, in your search for the correct TC factor value. The Neptune representative informed me that once the salinity probe calibration point includes temperature data, the TC factor value can be changed at will without requiring re-calibration. Changes to the TC factor will be applied to all salinity readings going forward - readings that were taken in the past will not change.

How to Configure TC Factor for the First Time
  1. Ensure that a temperature probe is plugged into the same module as your salinity probe
  2. Navigate to the `Advanced` tab of the salinity probe configuration
  3. Change the TC factor value to any non-zero positive number (2.1 is a good starting point)
  4. Click "Update Apex" in the top right corner to upload the new configuration
  5. Perform a salinity calibration
  6. Salinity readings going forward will now have temperature compensation applied accordingly
Calibrating Salinity
There are a number of factors that require attention when calibrating the salinity probe. If any one of them are ignored, the resulting readings will be incorrect. Some of these are described in the documentation, and some are not.

We have learned that salinity readings are affected by temperature. As a result, temperature plays a very important part in the calibration process. When temperature compensation is disabled, and the Apex salinity calibration solution is used (the grey pouch, 53,000 microsiemens), the temperature during calibration is assumed to be 77.0 degrees fahrenheit. This is rarely actually the case, and folks calibrating their salinity probes who aren't aware of this detail will almost certainly experience inaccurate readings.

If however, you have a temperature probe, you do not need to warm the calibration solution to precisely 77.0 degrees. Apex will instead calibrate the salinity using the temperature reading provided by the probe. Note that this still means that the solution must be at precisely the same temperature as the probe! To accomplish this, you can simply float the calibration solution pouch in water near the temperature probe location for 15 minutes. Note that the calibration solution temperature can change rapidly - for instance, holding the pouch in your hand while performing the calibration will affect the accuracy. Be sure to keep the pouch in the water during the entire calibration process to ensure that the solution temperature doesn't deviate.

How to Calibrate Salinity
  1. Ensure that your TC factor has already been configured with a non-zero value
  2. Float the calibration solution pouch near your temperature probe for 15 minutes
  3. Navigate to the `Basic` tab of the salinity probe configuration
  4. Click `Automatic Calibration`
  5. When asked to dry the salinity probe, remove it from the water and rinse with RO/DI. Then gently shake the probe, and pat the bottom dry with a paper towel
  6. When asked to insert the salinity probe into the calibration solution, open the pouch and insert the probe. Be sure to keep the pouch submerged in water during this process so the temperature remains stable
  7. Gently shake the probe in the calibration solution to remove any bubbles that might be trapped. This step is VERY important - shake for a little longer than you might otherwise
  8. Wait for the readings to stabilize, and complete the calibration. You're all done!
Note that the manual calibration process might work here as well, though I've never done that and so have no idea what it does or doesn't involve :)

Finding More Information
I spent several days searching for information on this topic. This post is, as far as I know, the most complete source of information on Apex salinity calibration and temperature compensation currently available. As a new Apex owner, I was shocked to find that none of this information was included in any of the documentation I could find. When asking Neptune customer support about this, they referred me to the PM2 documentation. There is definitely more information there than the "regular" Apex docs, however it is still not complete and does not fully apply to the use of the base module for salinity readings.

If you find that any of the information in this post is inaccurate, or if you have information that is not included here but should be, please post a comment on this thread!!

Summary
I have found the Apex salinity sensor to be relatively accurate when correctly configured. Of particular importance is configuring temperature compensation, ensuring stable temperature conditions during calibration, and ensuring the correct order of operations. I hope that others who may be struggling with their Apex salinity sensor are able to use this thread to fix things up. Happy reefing!
What a wonderful and informational post! I too have been having issues with my Apex reading salinity all over the place over the last year and a half since I've started reefing. My Milwaukee and refractometer always read the same when refractometer is calibrated, but salinity probe always reading extremely high. i.e, Milwaukee and refractometer would read 35ppt or 1.026, but Apex would be reading 36 up to 37.8ppt. Now that I've done this TC calibration starting from 2.1% my Apex reads salinity same the Milwaukee, at least for right now. Still a little confused as far as how to further adjust and dial in.
 

powers2001

started reefing 1999
View Badges
Joined
May 2, 2013
Messages
5,812
Reaction score
5,508
Location
LIVING WELLS MONTANA
Rating - 0%
0   0   0
Meet the SPS:

S=salinity
P=probe
S=shield

5C0726DC-EAF4-44B8-B726-39BC2D16A7ED.png

8C323C42-D1D5-41E4-AFF9-379B8E5D802D.png

538FE21D-019E-4D53-8239-18DC6FAF5090.png

If you saw the last episode of Neptune Systems: Let’s Talk Reef with Terence Fuggazi on YouTube, you would have caught this exciting new invention, but I’ll fill you in if you didn’t see it. Antonio, the owner of Vivid Creative Aquatics came up with a shield that slips over the end of a NS salinity probe and has a John Guest fitting attached to the SPS to provide a water supply through a 1/4” OD hose. The water coming through the device pushes water out the end of the shield where the tip of the probe is inside. The water supply provides the probe with fresh system water for an accurate parameter reading and at the same time pushes water out and away from the tip of the salinity probe containing all micro air bubbles. Genius
 
Last edited:

DeputyDog95

Active Member
View Badges
Joined
May 12, 2019
Messages
406
Reaction score
106
Rating - 0%
0   0   0
Great post!

Is there a way to force a particular reading?

I calibrated my handheld Hanna salinity checker to 35 PPT. I then set the tank water (through minor water adjustments) to precisely 35 PPT.

Following that, I calibrated the Apex temp probe to the Hanna temp reading for consistency.

Lastly, I dried the probe (wipe down and then some canned air) and calibrated the apex salinity probe using the 35 PPT tank water.

It's stable so far at 34.6 PPT. However, it's bugging me to have gone through all that work and it not match the Hanna. 4 tenths of a point is not the end of the world, but if I can make it match....

So, can I "force" the Apex to read 35 PPT to match my Hanna?
 

ReefersLove

New Member
View Badges
Joined
May 11, 2021
Messages
24
Reaction score
13
Location
Petaluma
Rating - 0%
0   0   0
I was trying to calibrate my new salinity probe a few weeks ago, and ran into all sorts of problems. Calibrations that were way off the mark, big fluctuations in readings that couldn't possibly be correct (and were verifiably not), etc. After searching around the internet, it was clear that a lot of folks have problems with this sensor. Some have written it off completely, and others insist that such problems are the result of calibration error. Information on correct calibration and configuration was present but scattered. Some of it was incomplete or incorrect. After having read many forum threads, spoken to Apex customer service, and experimented with different approaches, I am here to tell you that the sensor does work reliably when correctly configured and calibrated. This post is intended to capture everything I've learned in a single place, hopefully making this easier for someone else in the future.

Apex Configurables
There are only two configurable aspects of the Apex salinity sensor that can affect the reading. The first is, of course, the calibration point. The second is the temperature compensation factor (or TC factor). The salinity probe's readings are significantly affected by temperature - the TC factor setting allows you to compensate for this if you also have a temperature probe configured. The rest of this post captures the concepts and steps necessary to configure these inputs correctly.

TC Factor and Correctness
The TC factor configurable can be found on the advanced tab of the conductivity probe configuration page. It is a percentage, set manually - apparently different tanks land on different values. The percentage controls exactly how much compensation should be applied (i.e. x% compensation per degree).

One way to see if your TC factor is correct is by looking at your salinity graph with your temperature readings overlaid. If the TC factor needs to be adjusted, you will be able to see that the salinity readings track fluctuations in temperature. If the TC factor is correct(-ish), this correlation will not be present.

Here is what this looks like when TC is disabled (i.e. 0%):
Screenshot 2020-05-10 at 15.17.10.png


Here is the same system with TC configured:
Screenshot 2020-05-10 at 15.18.09.png


You can see from the first screenshot a clear correlation between temperature and salinity. Compare to the second screenshot with TC applied, and the salinity is relatively stable. I'm still not quite sure about those little bumps though. I continue to play around fine tuning the TC factor to see how they react (if at all - it's possible that the fluctuation is related to some funny business with my overflow).

Configuring the TC Factor
The TC factor defaults to 0. This is confusing for new Apex owners, as the salinity probe is all but worthless without applying temperature compensation. After speaking with customer service, and reading through some documentation, it appears that the default is 0 because setting it to anything other than zero causes problems in systems that don't have a temperature probe installed (the documentation includes a large WARNING section about this). Since Neptune cannot know or guarantee that any given system will have its temperature probe installed, the default is 0 to be safe.

If you have a temperature probe installed, you will definitely want to adjust the default TC factor. It is important to note that the temperature probe must be plugged into the same module as the salinity probe. In other words, if your salinity probe is plugged into your base unit, then you must also have a temperature probe plugged into the base unit. Similarly, if your salinity probe is plugged into a PM2 salinity module, then the temperature probe must be plugged into the same PM2.

TC Factor Value Range
It may take some experimentation to find the correct TC factor for your system. The Apex documentation states that seawater normally has a TC factor between 2.1% and 2.3%, thus recommending 2.2% as a good starting point. By comparison, a Neptune representative told me that in their experience, correct values are often found in the 1.9% to 2.1% range. Your mileage may vary.

The Apex documentation states that the salinity probe must be re-calibrated when changing the TC factor. According to a Neptune representative with whom I spoke, this is only partially true. When TC factor is non-zero and the salinity probe is configured, the current water temperature is stored inside the salinity calibration point that Apex records. This temperature is used to apply temperature compensation going forward. If the TC factor is zero when the salinity probe is configured, then the current water temperature is not stored, making future temperature compensation impossible. Therefor, in order for TC factor to take effect, it must be configured with a non-zero value at the time that the salinity probe is calibrated.

Fear not, however, in your search for the correct TC factor value. The Neptune representative informed me that once the salinity probe calibration point includes temperature data, the TC factor value can be changed at will without requiring re-calibration. Changes to the TC factor will be applied to all salinity readings going forward - readings that were taken in the past will not change.

How to Configure TC Factor for the First Time
  1. Ensure that a temperature probe is plugged into the same module as your salinity probe
  2. Navigate to the `Advanced` tab of the salinity probe configuration
  3. Change the TC factor value to any non-zero positive number (2.1 is a good starting point)
  4. Click "Update Apex" in the top right corner to upload the new configuration
  5. Perform a salinity calibration
  6. Salinity readings going forward will now have temperature compensation applied accordingly
Calibrating Salinity
There are a number of factors that require attention when calibrating the salinity probe. If any one of them are ignored, the resulting readings will be incorrect. Some of these are described in the documentation, and some are not.

We have learned that salinity readings are affected by temperature. As a result, temperature plays a very important part in the calibration process. When temperature compensation is disabled, and the Apex salinity calibration solution is used (the grey pouch, 53,000 microsiemens), the temperature during calibration is assumed to be 77.0 degrees fahrenheit. This is rarely actually the case, and folks calibrating their salinity probes who aren't aware of this detail will almost certainly experience inaccurate readings.

If however, you have a temperature probe, you do not need to warm the calibration solution to precisely 77.0 degrees. Apex will instead calibrate the salinity using the temperature reading provided by the probe. Note that this still means that the solution must be at precisely the same temperature as the probe! To accomplish this, you can simply float the calibration solution pouch in water near the temperature probe location for 15 minutes. Note that the calibration solution temperature can change rapidly - for instance, holding the pouch in your hand while performing the calibration will affect the accuracy. Be sure to keep the pouch in the water during the entire calibration process to ensure that the solution temperature doesn't deviate.

How to Calibrate Salinity
  1. Ensure that your TC factor has already been configured with a non-zero value
  2. Float the calibration solution pouch near your temperature probe for 15 minutes
  3. Navigate to the `Basic` tab of the salinity probe configuration
  4. Click `Automatic Calibration`
  5. When asked to dry the salinity probe, remove it from the water and rinse with RO/DI. Then gently shake the probe, and pat the bottom dry with a paper towel
  6. When asked to insert the salinity probe into the calibration solution, open the pouch and insert the probe. Be sure to keep the pouch submerged in water during this process so the temperature remains stable
  7. Gently shake the probe in the calibration solution to remove any bubbles that might be trapped. This step is VERY important - shake for a little longer than you might otherwise
  8. Wait for the readings to stabilize, and complete the calibration. You're all done!
Note that the manual calibration process might work here as well, though I've never done that and so have no idea what it does or doesn't involve :)

Finding More Information
I spent several days searching for information on this topic. This post is, as far as I know, the most complete source of information on Apex salinity calibration and temperature compensation currently available. As a new Apex owner, I was shocked to find that none of this information was included in any of the documentation I could find. When asking Neptune customer support about this, they referred me to the PM2 documentation. There is definitely more information there than the "regular" Apex docs, however it is still not complete and does not fully apply to the use of the base module for salinity readings.

If you find that any of the information in this post is inaccurate, or if you have information that is not included here but should be, please post a comment on this thread!!

Summary
I have found the Apex salinity sensor to be relatively accurate when correctly configured. Of particular importance is configuring temperature compensation, ensuring stable temperature conditions during calibration, and ensuring the correct order of operations. I hope that others who may be struggling with their Apex salinity sensor are able to use this thread to fix things up. Happy reefing!
 

ReefersLove

New Member
View Badges
Joined
May 11, 2021
Messages
24
Reaction score
13
Location
Petaluma
Rating - 0%
0   0   0
I was trying to calibrate my new salinity probe a few weeks ago, and ran into all sorts of problems. Calibrations that were way off the mark, big fluctuations in readings that couldn't possibly be correct (and were verifiably not), etc. After searching around the internet, it was clear that a lot of folks have problems with this sensor. Some have written it off completely, and others insist that such problems are the result of calibration error. Information on correct calibration and configuration was present but scattered. Some of it was incomplete or incorrect. After having read many forum threads, spoken to Apex customer service, and experimented with different approaches, I am here to tell you that the sensor does work reliably when correctly configured and calibrated. This post is intended to capture everything I've learned in a single place, hopefully making this easier for someone else in the future.

Apex Configurables
There are only two configurable aspects of the Apex salinity sensor that can affect the reading. The first is, of course, the calibration point. The second is the temperature compensation factor (or TC factor). The salinity probe's readings are significantly affected by temperature - the TC factor setting allows you to compensate for this if you also have a temperature probe configured. The rest of this post captures the concepts and steps necessary to configure these inputs correctly.

TC Factor and Correctness
The TC factor configurable can be found on the advanced tab of the conductivity probe configuration page. It is a percentage, set manually - apparently different tanks land on different values. The percentage controls exactly how much compensation should be applied (i.e. x% compensation per degree).

One way to see if your TC factor is correct is by looking at your salinity graph with your temperature readings overlaid. If the TC factor needs to be adjusted, you will be able to see that the salinity readings track fluctuations in temperature. If the TC factor is correct(-ish), this correlation will not be present.

Here is what this looks like when TC is disabled (i.e. 0%):
Screenshot 2020-05-10 at 15.17.10.png


Here is the same system with TC configured:
Screenshot 2020-05-10 at 15.18.09.png


You can see from the first screenshot a clear correlation between temperature and salinity. Compare to the second screenshot with TC applied, and the salinity is relatively stable. I'm still not quite sure about those little bumps though. I continue to play around fine tuning the TC factor to see how they react (if at all - it's possible that the fluctuation is related to some funny business with my overflow).

Configuring the TC Factor
The TC factor defaults to 0. This is confusing for new Apex owners, as the salinity probe is all but worthless without applying temperature compensation. After speaking with customer service, and reading through some documentation, it appears that the default is 0 because setting it to anything other than zero causes problems in systems that don't have a temperature probe installed (the documentation includes a large WARNING section about this). Since Neptune cannot know or guarantee that any given system will have its temperature probe installed, the default is 0 to be safe.

If you have a temperature probe installed, you will definitely want to adjust the default TC factor. It is important to note that the temperature probe must be plugged into the same module as the salinity probe. In other words, if your salinity probe is plugged into your base unit, then you must also have a temperature probe plugged into the base unit. Similarly, if your salinity probe is plugged into a PM2 salinity module, then the temperature probe must be plugged into the same PM2.

TC Factor Value Range
It may take some experimentation to find the correct TC factor for your system. The Apex documentation states that seawater normally has a TC factor between 2.1% and 2.3%, thus recommending 2.2% as a good starting point. By comparison, a Neptune representative told me that in their experience, correct values are often found in the 1.9% to 2.1% range. Your mileage may vary.

The Apex documentation states that the salinity probe must be re-calibrated when changing the TC factor. According to a Neptune representative with whom I spoke, this is only partially true. When TC factor is non-zero and the salinity probe is configured, the current water temperature is stored inside the salinity calibration point that Apex records. This temperature is used to apply temperature compensation going forward. If the TC factor is zero when the salinity probe is configured, then the current water temperature is not stored, making future temperature compensation impossible. Therefor, in order for TC factor to take effect, it must be configured with a non-zero value at the time that the salinity probe is calibrated.

Fear not, however, in your search for the correct TC factor value. The Neptune representative informed me that once the salinity probe calibration point includes temperature data, the TC factor value can be changed at will without requiring re-calibration. Changes to the TC factor will be applied to all salinity readings going forward - readings that were taken in the past will not change.

How to Configure TC Factor for the First Time
  1. Ensure that a temperature probe is plugged into the same module as your salinity probe
  2. Navigate to the `Advanced` tab of the salinity probe configuration
  3. Change the TC factor value to any non-zero positive number (2.1 is a good starting point)
  4. Click "Update Apex" in the top right corner to upload the new configuration
  5. Perform a salinity calibration
  6. Salinity readings going forward will now have temperature compensation applied accordingly
Calibrating Salinity
There are a number of factors that require attention when calibrating the salinity probe. If any one of them are ignored, the resulting readings will be incorrect. Some of these are described in the documentation, and some are not.

We have learned that salinity readings are affected by temperature. As a result, temperature plays a very important part in the calibration process. When temperature compensation is disabled, and the Apex salinity calibration solution is used (the grey pouch, 53,000 microsiemens), the temperature during calibration is assumed to be 77.0 degrees fahrenheit. This is rarely actually the case, and folks calibrating their salinity probes who aren't aware of this detail will almost certainly experience inaccurate readings.

If however, you have a temperature probe, you do not need to warm the calibration solution to precisely 77.0 degrees. Apex will instead calibrate the salinity using the temperature reading provided by the probe. Note that this still means that the solution must be at precisely the same temperature as the probe! To accomplish this, you can simply float the calibration solution pouch in water near the temperature probe location for 15 minutes. Note that the calibration solution temperature can change rapidly - for instance, holding the pouch in your hand while performing the calibration will affect the accuracy. Be sure to keep the pouch in the water during the entire calibration process to ensure that the solution temperature doesn't deviate.

How to Calibrate Salinity
  1. Ensure that your TC factor has already been configured with a non-zero value
  2. Float the calibration solution pouch near your temperature probe for 15 minutes
  3. Navigate to the `Basic` tab of the salinity probe configuration
  4. Click `Automatic Calibration`
  5. When asked to dry the salinity probe, remove it from the water and rinse with RO/DI. Then gently shake the probe, and pat the bottom dry with a paper towel
  6. When asked to insert the salinity probe into the calibration solution, open the pouch and insert the probe. Be sure to keep the pouch submerged in water during this process so the temperature remains stable
  7. Gently shake the probe in the calibration solution to remove any bubbles that might be trapped. This step is VERY important - shake for a little longer than you might otherwise
  8. Wait for the readings to stabilize, and complete the calibration. You're all done!
Note that the manual calibration process might work here as well, though I've never done that and so have no idea what it does or doesn't involve :)

Finding More Information
I spent several days searching for information on this topic. This post is, as far as I know, the most complete source of information on Apex salinity calibration and temperature compensation currently available. As a new Apex owner, I was shocked to find that none of this information was included in any of the documentation I could find. When asking Neptune customer support about this, they referred me to the PM2 documentation. There is definitely more information there than the "regular" Apex docs, however it is still not complete and does not fully apply to the use of the base module for salinity readings.

If you find that any of the information in this post is inaccurate, or if you have information that is not included here but should be, please post a comment on this thread!!

Summary
I have found the Apex salinity sensor to be relatively accurate when correctly configured. Of particular importance is configuring temperature compensation, ensuring stable temperature conditions during calibration, and ensuring the correct order of operations. I hope that others who may be struggling with their Apex salinity sensor are able to use this thread to fix things up. Happy reefing!
Thanks for the detailed information! I’m in the process of setting up a new Apex, so perfect timing for me.
 

Daniel@R2R

Living the Reef Life
View Badges
Joined
Nov 18, 2012
Messages
37,488
Reaction score
63,896
Location
Fontana, California
Rating - 100%
1   0   0
Meet the SPS:

S=salinity
P=probe
S=shield

5C0726DC-EAF4-44B8-B726-39BC2D16A7ED.png

8C323C42-D1D5-41E4-AFF9-379B8E5D802D.png

538FE21D-019E-4D53-8239-18DC6FAF5090.png

If you saw the last episode of Neptune Systems: Let’s Talk Reef with Terence Fuggazi on YouTube, you would have caught this exciting new invention, but I’ll fill you in if you didn’t see it. Antonio, the owner of Vivid Creative Aquatics came up with a shield that slips over the end of a NS salinity probe and has a John Guest fitting attached to the SPS to provide a water supply through a 1/4” OD hose. The water coming through the device pushes water out the end of the shield where the tip of the probe is inside. The water supply provides the probe with fresh system water for an accurate parameter reading and at the same time pushes water out and away from the tip of the salinity probe containing all micro air bubbles. Genius
This is cool! Hadn't seen it before today
 

chicago

Valuable Member
View Badges
Joined
Apr 28, 2014
Messages
1,602
Reaction score
548
Location
chicago
Rating - 0%
0   0   0
Ok.. i have a large tank.. and large sump... so i actually run two salinity probes.. just cleaned them and one is now reading 1 and the other is now reading. 53. Crazy.
 

chicago

Valuable Member
View Badges
Joined
Apr 28, 2014
Messages
1,602
Reaction score
548
Location
chicago
Rating - 0%
0   0   0
ok so now I calibrated one and in the calibration fluid it reads spot on 35.. i pull it out put in the tank it reads.. 57.. put in back in the solution and back to 35.. what gives.... hand held says my tank is 35 ...
 

DWill

Active Member
View Badges
Joined
Jan 11, 2020
Messages
444
Reaction score
374
Rating - 0%
0   0   0
ok so now I calibrated one and in the calibration fluid it reads spot on 35.. i pull it out put in the tank it reads.. 57.. put in back in the solution and back to 35.. what gives.... hand held says my tank is 35 ...
First make absolutely sure there’s no bubbles on it and none can get to it. Second if there are no bubbles let it settle in the tank for a while.
 

chicago

Valuable Member
View Badges
Joined
Apr 28, 2014
Messages
1,602
Reaction score
548
Location
chicago
Rating - 0%
0   0   0
Ok. So my issue I have no idea what was causing it. I ended up taking the probe out. And disconnecting it. Running the wire away from all other wires. Same with temp probe. And now reads solid tank water at 34.9.
 
Back
Top