And here's my point....
Two years waiting for the IOND.
People getting it and having issues setting it up.
But where is GHL's support?
Sitting at the ticket system waiting for tickets while people help on here doing GHL's support for them.
If this is what you all are satisfied with and willing to put up with?
God bless and good luck. Y'all getting what you deserve.
One person is having a problem.
Many people have not.
So you are wishing that people fail with their IOND purchases? Nice.
Have you managed to get through the ‘prepare measurement cell’ phase?
And all of the pumps are working correctly and priming/filling the measurement cell? (i.e. the probe has had the 12 hours to stabilise in the fluid).
And after that, you get the ‘bad condition’ error when you try to run a test?
On this video, you can briefly see the voltage page (albeit this is on GHL connect, but presumably you should also be able to see that on GCC somewhere).
There is also a troubleshooting page here: https://www.aquariumcomputer.com/knowledge-base/ion-director-tips-maintenance-and-troubleshooting/
It doesn’t specifically mention your error but it does mention poor probe health indications caused by air in the system. I have read that other people have had this problem.
It could be that the sensor is faulty, but I’m guessing GHL will want to try to rule out everything else first.
And here's my point....
Two years waiting for the IOND.
People getting it and having issues setting it up.
But where is GHL's support?
Sitting at the ticket system waiting for tickets while people help on here doing GHL's support for them.
If this is what you all are satisfied with and willing to put up with?
God bless and good luck. Y'all getting what you deserve.
"Yes" to all your questions. I've seen the trouble-shooting
WHAT??!! I'm only saying this for anybody new on this GHL forum (because all the "regulars" already know this):yea i block both him and lasse, theyre like antimatter to me
Perfect....just the way I want it. I'm happy with how it turned out.How is your IOND going?
Nema problema Dr.Jim - I´m only proud to be seen as an antimatter to the unsensitive persons - unsensitive is antimatter to sensitive as you knowWHAT??!! I'm only saying this for anybody new on this GHL forum (because all the "regulars" already know this):
Lasse is one of the most knowledgeable and helpful contributors you will encounter. He is always quick and willing to help others. (It is hard to imagine anybody "bad-mouthing" him!)
@Dr. Jim Did you solve your issue? I have an IOND working perfectly fine, and can help if needed.DAY 6 and STILL NO SOLUTION: I thought for sure that after waiting all weekend for today (Monday), I would have gotten this problem resolved. I am a fairly calm, reasonable and patient individual but I am getting a bit frustrated now.
I had a few quick emails back and forth this morning with Vinny, but that abruptly ended. At this point, I need more than people telling me to "read the manual." I know I made a mistake by starting the "Prime Measurement Cell" with the Sample Tube empty (which led to the error and red bar) but nothing I've done since has worked. I've read the Manual several times and am confident that I'm doing everything right now. But, I might be missing something because I don't understand 3 things Vinny asked me for and perhaps somebody can help me with these questions:
1) Did you provide the voltages of your sensor?
2) Did you run a new test afterwards and provide the updated voltages? (I ran several tests but none registered anything)
What is meant by these questions regarding"voltages" and where/how do I do this?
Is this something that has to be done with the App? (I've been doing everything on GCC)
3) Press (i) icon beside the sensor health and attach the info to your ticket. Where do I find this "(i) icon beside the sensor health?"
------------
--When I click on the box next to the Sensor Performance, I get the message: "Bad condition; no more measurements possible; replace sensor" That doesn't sound too good to me! (Is this the icon Vinny mentioned?)
------------
I apologize for taking up so many posts on this thread...... Thank you for your help. Jim
I just received the replacement probe yesterday and am running tests now, but thank you for your offer to help.@Dr. Jim Did you solve your issue? I have an IOND working perfectly fine, and can help if needed.
I always run high precision, the difference in reference solution usage is so little, that it doesn't make sense to run normal tests. I ran 3 multi reference test in a row, to make sure results were appropriate, the only parameter that was slightly deranged was nitrate 3 below (reference is 11, reading at 8). the rest was pretty spot on. Don't bother with salifert, the precision of an ionic probe vs. a reagent test is on 2 different planets.I just received the replacement probe yesterday and am running tests now, but thank you for your offer to help.
It seems to be running fine. The problem was obviously a bad probe. I suspect there was something wrong with the plug because it didn't insert as easily as I think it should have, and had a slight wiggle afterwards. The new plug pushed in firmly like I suspect it should.
I will make a table and post my results (hopefully tomorrow). I am performing 5 runs: Two "regular"; two "high precision" and one with the Multi Reference. I also ran Salifert tests at the same time with the Ca, Mg, NO3 and K kits, (and will comment on recent ICP tests for Na).
PARAMETER | CA | MG | K | Na | NO3 |
1st Run (Reg) | 454 | 1370 | 435 | 10,754 | 6 |
2nd Run (Reg) | 454 | 1367 | 435 | 10,754 | 3 |
3d Run (High Precision) | 441 | 1311 | 429 | 10,605 | 12 |
4th Run (High Precision) | 445 | 1317 | 430 | 10,621 | 5 |
SALIFERT/ICP* | 440 | 1335 | 390 | 10,269* | 1 |
MULTI-REF (TEST #1) | 411 | 1317 | 382 | 10,610 | 11 |
MULTI-REF (TEST #2) | 427 | 1384 | 388 | 10,729 | 4 |
MULTI-REF (REFERENCE VALUES) | 420 | 1300 | 390 | 10,700 | 11 |
I think the IOND nitrate measures definitely higher than the Red Sea Pro nitrate test kit. If we all adjusted our carbon dosing (or whatever nutrient export method) based on the colorimetric test kits, and our tanks are doing fine, I am not sure it's wise to modify our approach. It would be interesting to see @Vinny@GHLUSA chime in, maybe they know why nitrate on IOND seems to read higher than on different test kits/ICP? I don't think it's an issue per se, it's more getting used to a different way to look at it.For what it is worth, I made a table of results using these tests, each one being performed immediately after the prior one:
RUN #1: Tested sample with "Regular" Precision
RUN #2: Tested sample with "Regular Precision
RUN #3: Tested sample with "High Precision"
RUN #4: Tested sample with "High Precision"
Next Row in Table shows Salfert test results on same sample (and *ICP result from 5 weeks ago for Na, just out of curiosity)
RUN # 5: Multi-Reference Solution (High Precision)
RUN #6: Multi-Reference Solution (High Precision)
Multi-Reference Solution's Reference numbers.
PARAMETER CA MG K Na NO3 1st Run (Reg) 454 1370 435 10,754 6 2nd Run (Reg) 454 1367 435 10,754 3 3d Run (High Precision)
441
1311
429
10,605
124th Run (High Precision)
445
1317
430
10,621
5SALIFERT/ICP* 440 1335 390 10,269* 1 MULTI-REF
(TEST #1)
411
1317
382
10,610
11MULTI-REF
(TEST #2)
427
1384
388
10,729
4MULTI-REF
(REFERENCE VALUES)
420
1300
390
10,700
11
I know that there aren't enough tests performed to make any of this statistically significant, but some may find some of this interesting. My thoughts on these limited tests:
1) The first two runs ("Low Precision) yielded excellent results.
2) The next two "High Precision" RUNS seem acceptable, but the high NO3 of 12 is a little disappointing.
3) I know we aren't supposed to compare these readings to test kits but it has been test kits that I have been relying on for many years to target levels that have given me much success. It doesn't hurt to compare them and keep in mind the differences, going forward, at least for me. (I have compared many brands of test kits with many ICP tests over the years and have found the Salifert kits to correlate fairly well with them).
4) I know that comparing Na to a ICP test from 5 weeks ago is of minimal significance, but I will say that my ICP Na results have historically always been a little low.
4) I'm a little disappointed that the two tests on the Multi-Reference solution are not closer to being the same for Ca, Mg, and NO3; and, that some of the parameters aren't closer to the Reference Values used to make the Multi-Reference Solution. The fact that there are these differences makes me wonder if it can be said that the IOND is more accurate than ICP testing.
5) For me, the IOND is an interesting "toy" to play with but I would never use it to control dosing (nor would I do that with the KH Director). I will probably use the IOND once a week and continue comparing it to my test kits until I am comfortable with a trend, if a steady one develops. Like with any instrument, one must always be leery of it's accuracy. And, as @CEReefer suggested, I see no reason to NOT always run the tests on the High Precision mode.
6) When I purchased the IOND, I was mostly interested in the NO3 monitoring, and this is the one parameter that I have the least confidence in after these limited trials which is a little disappointing. But, if the IOND NO3 values are accurate, then they are much higher than the Salifert NO3 results that I have been monitoring for years! (But the big question: which values are the accurate ones considering the range I got?!)
All in all, I'm glad that I have this instrument. I will continue to monitor it and repeat some of these "comparison tests" again in the near future. I would guess that it might take a little time for the unit to "break in" and possibly rid itself of some unseen microbubbles that may be in the tubing of a new set-up.
Hope this is of some interest to any of you!