PH on Apex bumps up every time chiller runs

schuby

Valuable Member
View Badges
Joined
Jun 24, 2019
Messages
1,011
Reaction score
841
Location
Orange County, CA
Rating - 0%
0   0   0
1594246538117.png


I've tried calibrating my PH probe with Temperature Compensation enabled and disabled. Both ways I get the same result you see in my screen-shot: PH goes up temporarily when the chiller drops the temp. When the chiller turns off, PH goes back to about where it was before.

Does this happen for anybody else?
 

Breadman03

Valuable Member
View Badges
Joined
Jun 30, 2015
Messages
2,249
Reaction score
2,022
Location
Luzerne County, PA
Rating - 0%
0   0   0
That looks like electrical interference. Try cycling it on/off a few times manually. And note that the pH probably “switches” by the same amount each time.
 