Product Review: AquaMaxx Puratek Reverse Osmosis/Deionization Filtration Unit

mell0w

Community Member
View Badges
Joined
Jul 15, 2020
Messages
48
Reaction score
7
Location
Henderson
Rating - 0%
0   0   0
I love that the screen is actually LED lit, mine is hard to see sometimes!
 

Lil' RegalReefster

Active Member
View Badges
Joined
Oct 31, 2019
Messages
132
Reaction score
39
Rating - 0%
0   0   0
I love the booster pump because I have low water pressure. The microprocessor unit also adds extra functionality that is rarely seen in RODI units for such an amazing price.
 

Ross Petersen

Well-Known Member
View Badges
Joined
May 25, 2019
Messages
543
Reaction score
311
Location
Vancouver BC
Rating - 0%
0   0   0
Lots of positive vibes on this one. Not much user experience yet though? Seems like this was just launched? It's in stores in Canada though, can't be that new :p
 

TheWB

Valuable Member
View Badges
Joined
Sep 22, 2017
Messages
2,329
Reaction score
3,601
Location
Puyallup,WA
Rating - 0%
0   0   0
Lots of positive vibes on this one. Not much user experience yet though? Seems like this was just launched? It's in stores in Canada though, can't be that new :p
There’s a few of us that have them. Mine is an older version from back when it was branded as Puritek, not Aquamaxx. It works great. I expanded mine with two additional stages so I can have a cation, an anion and a mixed bed of DI resin. If you need a new system this one is pretty darn good.
 

Ross Petersen

Well-Known Member
View Badges
Joined
May 25, 2019
Messages
543
Reaction score
311
Location
Vancouver BC
Rating - 0%
0   0   0
Thanks. I’ve actually got a Spectrapure CSPDI 90 (great system) new in box but this new Aquamaxx has all the built in functional bells and whistles that I’ll have to otherwise chase down or install / plumb In. It just makes sense. I might sell the Spectrapure because I don’t love the need to upgrade and take up more real estate with separate booster pumps etc.

There’s a few of us that have them. Mine is an older version from back when it was branded as Puritek, not Aquamaxx. It works great. I expanded mine with two additional stages so I can have a cation, an anion and a mixed bed of DI resin. If you need a new system this one is pretty darn good.
 

Zadok Jollie

Active Member
View Badges
Joined
Jul 10, 2019
Messages
178
Reaction score
142
Rating - 0%
0   0   0
This system is so much better than other systems I've seen. Thanks for such a comprehensive review!
 

Drasi

Community Member
View Badges
Joined
Jul 3, 2020
Messages
31
Reaction score
6
Location
Grand Rapids
Rating - 0%
0   0   0
Interesting for sure.. I like the features! I could probably get away with the lower model at $160 vs this $300 one, but the features of the tds meter and the booster pump and all the other bells and whistles.. hmmm :)
 

RyanW250

Active Member
View Badges
Joined
May 15, 2019
Messages
211
Reaction score
252
Rating - 0%
0   0   0
Product Review: AquaMaxx Puratek Reverse Osmosis/Deionization Filtration Unit
By Dana Riddle

1.jpg


Water evaporates, salts do not – this is one of the first things a marine hobbyist learns. Hence fresh water must be added to keep the aquarium water from becoming too salty. If you’re fortunate, tap water in your area might be suitable for topping off the water level. In many cases, this is not the case and some sort of pretreatment is required. If you’re unsure of your local water quality, an internet search for ‘map USA groundwater hardness’ will give you an idea.

One of the most common methods of pretreating top off water is a reverse osmosis (RO) filter, often followed by de-ionization. We’ll take a look at one of the better RO units – the AquaMaxReverse Puratek Deluxe Osmosis/Deionization unit. This system has a prefilter (rated at 5 microns) to remove silt, clay, rust, etc., a carbon filter for some organic and total and residual chlorine, a thin film composite filter reverse osmosis filter, and finally a deionization cartridge containing anionic and cationic resins. Also included is a small pump boosts feed water pressure to increase permeate production, a conductivity meter to monitor permeate quality (reported as parts per million total dissolved solids (TDS) and a wrench for removal of housings when filter replacement is necessary. A 9-page instruction manual is also provided. The only things not included are two lag bolts for mounting to wall, if desired.

If you’re new to the hobby, some of these terms might not be familiar, so before continuing, let’s define some of the terminology.

Glossary
Reverse Osmosis (RO)– A water treatment filter where pressurized water is forced through a filter than has tiny pores to remove contaminants. Nanofiltration (NF)– Similar to reverse osmosis, but the filter membrane has larger filter pores and thus can operate at a lower water pressure. Removal of many substances can be as high as 98%.
Feed – Water to be treated by a RO or NF filter.
Permeate – Clean water produced by RO or NF filtration.
Concentrate – Water containing the substances that are removed by a RO or NF. Also called ‘reject’ this water goes down the drain. Micron – One thousandths of a millimeter.
Ion – An atom or molecule demonstrating a positive (cation) or negative (anion) charge.
Electrical Conductivity (EC) – The ability of substances dissolved in water to conduct an electrical charge. Also called Conductivity for short. Reported in units of microSiemens per square centimeter or milliSiemens per square centimeter. The lower the EC, the less dissolved salts the water contains. As a footnote, silica does not conduct an electrical charge. Total Dissolved Solids – The amount of organic and inorganic substances dissolved in water. Abbreviated as TDS, it can be monitored by a electrical conductivity meter. Deionization – Removal of charged ions through use of cationic and/or anionic resins.
GPD – Gallons per day.
PSI – Pounds per square inch, a measure of pressure.

What is Reverse Osmosis?
Reverse osmosis is a treatment process based on pressure driving water through a semi-permeable polymer membrane. This membrane resembles a roll of paper towels and has multiple pores through which the water molecules pass while many contaminants cannot. The size of the pores determines what contaminants will be removed. Technically, units sold for household use are almost always ‘nanofiltration’ devices. Nanofiltration membranes have larger pore size and thus require less feed water pressure. Since an average household water pressure is about 50 psi, a true reverse osmosis membrane (with a smaller pore size) would perform poorly in that situation. True RO filters require pressures of 250 to 1,200 psi, while nanofiltration membranes are effective at lower pressure (75 to 150 psi). Nanofiltration devices can remove large particles (sand, silt, clays although these are usually removed by a prefilter rated at 5 microns), bacteria, high molecular weight proteins, organics including humic acids, pyrogens, viruses and divalent ions, larger monovalent ions, color and odor. Since the term ‘reverse osmosis’ is entrenched in the vernacular among hobbyists, it will be the term used in this article.

Cycles of Concentration
This is a term normally used in the cooling tower water treatment field but it applies to aquarium as well. If no means are taken to remove phosphorus through biological, chemical, or dilution (water changes), the addition of tap water containing phosphorus will increase that concentration. This could be especially important if your water utility adds sodium hexametaphosphate to control corrosion in water mains and household potable water plumbing. Of course, cycle of concentrations apply to any chemical we might consider to be a contaminant.

Rejection Rate
This is the ratio of ‘clean’ water produced (permeate) and the ‘dirty’ water that goes down the drain (the ‘concentrate’ or ‘reject’.) This depends upon many factors, including feed water pressure and chemical composition, temperature, filter pore size, and so on. I checked and found permeate production was 36% of feed water. In other words, 1 gallon of pure water is produced for every 1.77 gallons of concentrate. Again, your ratio could be different.

Pressure Booster Pump
The AquaMaxx unit uses a small diaphragm pump to boost (roughly double) household water pressure. A small water pressure gauge shows pressure to be about 100 psi and this increased psi increases permeate production rate (while reducing concentrate flow.) See Figure One.

2.jpg

Figure One. The booster pump.


Power Consumption

Using a watt meter, I found the booster pump consumes 35 watts and 0.34 amps. The pump operates at 24v DC.

Performance
This RO unit is rated at 100 GPD, but I found it to be ~128 GPD. Bear in mind that feed water quality and pressure, temperature, etc. all play important parts in permeate production – your results could be different. These were my feed water parameters: Temperature = 78°F (25.6°C); Conductivity = 165 µS∙cm; Water pressure = 100 psi.

Increased temperature (up to 100°F) and pressure will increase permeate flow (but be aware of pressure ratings of filter vessels.)

TDS Meter
A digital meter reports total dissolved solids (TDS) in parts per million. It is possible to convert TDS to Electrical Conductivity (EC), but at such low permeate TDS, this is not necessary. See Figure Two.

Microprocessor Controller
The microprocessor is the brain of the unit and controls these functions:

  • Flush Function: RO and NF membranes tend to foul with silt, organics, and bacteria hence all household RO filters utilize micron filtration and activated carbon as pretreatments. Drinking water usually maintains a chlorine residual of 0.1 mg/L hence bacteria are keep in check. However, the AquaMaxx RO unit has an automatic flush function to remove membrane-fouling substances, hence the life of the membrane is extended.​
  • Source Water Monitoring: If the source (feed) water supply is interrupted, the unit will shut down.​
  • Reservoir Full: If you’re using a permeate storage reservoir (not included) and it becomes full, the microprocessor will shut the unit down.​
  • Permeate Production: The microprocessor can sense production of permeate and an indicator light turns on, and the TDS meter reports water quality after a few seconds.​

See Figure Two.

3.jpg

Figure Two. The microprocessor and TDS meter.


Discussion

I found the AquaMaxx nanofiltration unit to exceed the advertised production of ‘pure’ water. Power consumption by the booster pump is low (35 watts) while almost doubling feed water pressure. The booster pump reduces the amount of concentrate that goes down the drain. The unit’s microprocessor automatically flushes the membrane thus extending the filter’s useful life. These factors reduce operating and maintenance costs. In short, this filtration system is effective and efficient.

There are some things the owner should know. This filter can remove things we would prefer to not enter the aquarium, but it also removes some good things (such as calcium and magnesium) – it is not selective. If your water is hard (contains a lot of calcium and magnesium), water straight from the tap could raise these elements, perhaps leading to imbalances in the aquarium water’s chemistry. It is best to add ‘pure’ water as a top off and supplement elements as necessary. The same is true for mixing artificial seawater.

Effect on pH
An effective nanofilter will reduce compounds that buffer against drops in pH, this is compounded by the fact that carbon dioxide gas will pass through the membrane, thus lowering pH. I do not have a pH probe designed for use with low-conductivity water, but I did add the bromcresol green/methyl red reagent and observed an immediate pinkish color, indicating a pH somewhere near 4.5 standard units. See Figure Three.

4 copy.jpg

Figure Three. Most buffering agent are removed by the membrane while carbon dioxide passes, resulting in depressed pH.

Effect on Alkalinity

Alkalinity can also be reduced. This was determined by titrating samples of feed and permeate samples with a sulfuric acid (of known strength) to an endpoint pH of ~4.5 (using bromcresol green/methyl red as the endpoint indicator.) See Figure Four.

5.jpg

Figure Four. The feed water has a low alkalinity, and the permeate contains less.

Effect on Dissolved Oxygen

Dissolved oxygen (DO) can be reduced by some reverse osmosis units. Using a Hach luminescent DO probe connected to a HQ40d multi-meter, I found impact on DO to be minimal. See Figure Five.

6.jpg

Figure Five. Removal of dissolved oxygen is minimal.

Effect on Electrical Conductivity

As mentioned earlier, substances dissolved in water can conduct an electrical current (pure water does not) and in many cases conductivity can be used to indicate purity. EC was measured by a Corning conductivity meter (model 311). See Figure Six.

7.jpg

Figure Six. Almost all substances that can conduct an electrical current have been removed.

Effect on Phosphorus and Nitrate

I am fortunate that my tap water contains little phosphorus (ortho-phosphate as P) and nitrate (as NO3-N). The P content at the tap was below the detection limit of the analytical device (a Hach DR 890 colorimeter), hence it would be an exercise in futility to test further. Nitrate in the tap water showed a reduction (71%) but the permeate result was below the detection limit as well, therefore I do not consider the result as reliable, but will say there was nitrate removal.

I’ve owned many RO units over the years and the AquaMaxx unit is by far the nicest, and I look forward to many years of service.

Reference
Reverse Osmosis and Nanofiltration, Manual of Water Supply Practices –M46, 1999. American Water Works Association. Denver, Colorado. 173 pp.

Disclaimer: This RO (nanofiltration) unit was donated to my laboratory by Marine Depot (www.marinedepot.com and Reef2Reef (www.reef2reef.com) in return for an objective review. All claims made in this review are true to the best of my knowledge. Various parameters (temperature, feed water quality, and others) could affect the performance of this unit and your results could be different.
The auto shut off function for water supply error is pretty sweet!
 

reefingBeAmazin22

Active Member
View Badges
Joined
Sep 20, 2020
Messages
127
Reaction score
105
Location
pecatonica
Rating - 0%
0   0   0
I think it sounds like an amazing unit, that has large water capacity, I’m currently running well water so it’s not ideal, wish I could have that one
 

Greenreef75

Active Member
View Badges
Joined
Oct 12, 2017
Messages
436
Reaction score
2,074
Rating - 0%
0   0   0
Great review that has information for both beginners and more advanced reefers.
 

Water Dog

2500 Club Member
View Badges
Joined
Mar 28, 2016
Messages
4,219
Reaction score
4,892
Location
Fairfield, CT
Rating - 0%
0   0   0
I’m curious as to how the “auto flush” mechanism works... My understanding of a regular manual RO flush valve is that it allows you to redirect and pass water over the exterior of the waste side of the membrane under no pressure... no water is actually pushed through the membrane. Since it’s under no pressure, I’ve read that flush valves are debatable as to whether or not they really do anything to actually flush out the RO membrane. Do the electronic auto flush valves differ in how they flush membranes from their manual counterparts?

The way I understand it, a properly flushed membrane is due to having a waste to product water ratio of anywhere between 3 to 1 and 4 to 1. With those ratios, the membrane gets flushed under pressure by pushing a given amount of water through membrane and its waste water side for a given amount of water pushed through to the product water side. Without the proper 3 or 4 to 1 waste to product water ratio, you get scale buildup within the laters of the RO membrane, thereby reducing its useful life.

I know Spectrapure also has an auto flush system on their UHE 2:1 (Ultra High Effepifiency) RODI units and now, we see them on the Aquamaxx Puratek units. I guess my question is, does an auto-flush mechanism actually extend the life of the RO membrane? If so, how does it do it? Or is it the proper waste to product water ratio that flushes the membrane under pressure during operation that maintains the RO membrane life...

@Dana Riddle @Buckeye Hydro any thoughts on this? How does that auto flush system work? Am I mistaken as to my understanding of how an RO membrane is properly flushed?
 
Last edited:

Citrus

Community Member
View Badges
Joined
Jul 24, 2019
Messages
59
Reaction score
12
Location
Naples
Rating - 0%
0   0   0
Love that it has a minimal water waste. Great article to help new users become familiar with all of the functions the unit has and can do
 

madisonmajhen

New Member
View Badges
Joined
Oct 18, 2020
Messages
23
Reaction score
66
Location
Naples
Rating - 0%
0   0   0
Looks like a decent RO system! I haven’t owned one yet; still carrying around 5g buckets from the LFS unfortunately. Low waste, affordable and convenient it looks awesome. I desperately need it for my freshwater shrimp tanks also. Saving my pennies! :)
 

MasterDinh

Community Member
View Badges
Joined
May 10, 2019
Messages
37
Reaction score
48
Rating - 0%
0   0   0
This is a product that is worth taking a strong look at! I like how it has a microprocessor controller. I also like the booster pump. Thanks for such an awesome review! Crossing my finger to win this unit and it will be my first one. Have been doing a lot of research on a RODI unit.
 

Forklifthazzerd

New Member
View Badges
Joined
Jul 16, 2020
Messages
12
Reaction score
11
Location
US
Rating - 0%
0   0   0
I could use a Dana Riddle's Product Review of the Puratek Deluxe 100 GPD RO/DI Filter System because I have ben buying RO/DI from the local pet store . It cheap for a my 65g and 16g , but if I upgrade I will be haling a lot of water. This wood be a grate Update.
 

High pressure shells: Do you look for signs of stress in the invertebrates in your reef tank?

  • I regularly look for signs of invertebrate stress in my reef tank.

    Votes: 42 32.6%
  • I occasionally look for signs of invertebrate stress in my reef tank.

    Votes: 29 22.5%
  • I rarely look for signs of invertebrate stress in my reef tank.

    Votes: 25 19.4%
  • I never look for signs of invertebrate stress in my reef tank.

    Votes: 33 25.6%
  • Other.

    Votes: 0 0.0%
Back
Top