I switched out Reagents A, B, C this morning for the time and also calibrated it for the first time since the initial start up a couple months ago. I'm a little confused by the readings I got. Prior to the calibration, my readings were pretty much spot-on to my settings...8.75 alk, 450 calcium and 1250 mag. It calibrated fine this morning but when I ran the first test afterwards, it read 9.25 alk, 390 calcium and 1250 mag. I'm trying to understand why my alk and calcium readings would be so different after the calibration?
 

SuncrestReef

That Apex guy
View Badges
Joined
Jan 18, 2018
Messages
4,214
Reaction score
9,224
Location
Oregon
Rating - 0%
0   0   0
Calibration adjusts how the Trident interprets its readings. Prior to calibration, your Trident was artificially reporting lower alk than what you actually had, and now it's reporting closer to reality. Also, keep in mind the Trident's margin of error is +/- 0.05 dKH for alk, and +/- 15 ppm for cal & mag, so there will always be some fluctuations between readings.

What were you using to test your alk before the Trident? Did you try using that test kit to test the Trident's calibration solution to see how far off your other test kit is?
 
OP
OP
D

denverjon

Active Member
View Badges
Joined
Aug 16, 2009
Messages
445
Reaction score
52
Location
denver
Rating - 0%
0   0   0
Thanks for the feedback. I previously used Salifert but I didn't test the calibration solution with it today. I actually had a couple extra bottles of calibration solution so I calibrated the trident twice. The first test after each calibration produced similar readings.
 

Mal11224

Valuable Member
View Badges
Joined
Mar 8, 2017
Messages
1,747
Reaction score
3,249
Location
Long Island, NY
Rating - 0%
0   0   0
I never really found the trident to be very accurate even after using the calibration fluid. A suggestion would be to test your parameters with trustworthy kits and use your tank water with those readings to calibrate your trident. That’s what I do and it works well. I mean… why can’t you use your tank water to calibrate? :rolleyes:
Just a suggestion.
 

xabo

Valuable Member
View Badges
Joined
Aug 21, 2009
Messages
1,035
Reaction score
514
Location
Wash. D.C.
Rating - 100%
1   0   0
I never really found the trident to be very accurate even after using the calibration fluid. A suggestion would be to test your parameters with trustworthy kits and use your tank water with those readings to calibrate your trident. That’s what I do and it works well. I mean… why can’t you use your tank water to calibrate? :rolleyes:
Just a suggestion.
This it what I've started doing. Seem to remember a member here who calibrates his Trident this way saying Neptune recommended this process to him.
 

areefer01

2500 Club Member
View Badges
Joined
Jun 28, 2021
Messages
2,696
Reaction score
2,759
Location
Ca
Rating - 0%
0   0   0
why can’t you use your tank water to calibrate? :rolleyes:

Question: Why can't you use your tank water to calibrate?

Answer: Because it is not a reference solution.

Longer answer: Un-opened calibration fluid has a "known" alk, ca, and mg number. The calibration task on the Apex will ask you to enter these values before it runs the tests. You can use manual test kits, the left over calibration fluid, and compare results to the trident if you want but should do so within 24 hours.
 

xabo

Valuable Member
View Badges
Joined
Aug 21, 2009
Messages
1,035
Reaction score
514
Location
Wash. D.C.
Rating - 100%
1   0   0
Question: Why can't you use your tank water to calibrate?

Answer: Because it is not a reference solution.

Longer answer: Un-opened calibration fluid has a "known" alk, ca, and mg number. The calibration task on the Apex will ask you to enter these values before it runs the tests. You can use manual test kits, the left over calibration fluid, and compare results to the trident if you want but should do so within 24 hours.
If I do this and the numbers either don't match or fall within the margin of error, where do I go from there?
 

Mal11224

Valuable Member
View Badges
Joined
Mar 8, 2017
Messages
1,747
Reaction score
3,249
Location
Long Island, NY
Rating - 0%
0   0   0
Question: Why can't you use your tank water to calibrate?

Answer: Because it is not a reference solution.

Longer answer: Un-opened calibration fluid has a "known" alk, ca, and mg number. The calibration task on the Apex will ask you to enter these values before it runs the tests. You can use manual test kits, the left over calibration fluid, and compare results to the trident if you want but should do so within 24 hours.
If you test the calibration fluid and it is not correct, according to what is on the bottle, then it is not a sufficient reference solution. I use Hanna to check my trident and tests are pretty much the same.
 

areefer01

2500 Club Member
View Badges
Joined
Jun 28, 2021
Messages
2,696
Reaction score
2,759
Location
Ca
Rating - 0%
0   0   0
If I do this and the numbers either don't match or fall within the margin of error, where do I go from there?

I'm not sure I understand the question. Sorry.

I replied why you do not use the display's water for reference when calibrating the Trident. The provided calibration fluid is a "known" value and marked on the bottle. Once opened it should be used right way using the calibration task found in the Apex portal. You run through a series of steps one of which will ask you for the values on the bottle.

If you use the displays values via a manual test kit then those numbers are only as good as your manual test, titration, or auto tool like a bot or Hanna). I don't know about you but I've run back to back tests and my numbers are not the same although close. That is why that task exists and the bottles included.

Lastly don't chase numbers and pick a record of source. In my case I use the Trident. I do send off ATI ICP tests 2 or 3 times a year. When I do this I will grab a sample of water before the Trident runs, run phosphate, nitrate, and alk tests (hanna kits), then pull ATI's sample, and record everything down. Once ATI's results come back I can do a comparison. Each time I've done this the numbers are within their respected margin of error.

Not sure if this helps and apologies if not. About the only other thing that I do is set the calibration level as close as I can to the normal trident sample line. Outside of that I'm not sure other than I tend to trust the calibration more. If in doubt fire off a ticket later today to Neptune and see what they recommend if it is that far off.

Hope you are having a great day!
 

Sean Clark

7500 Club Member
View Badges
Joined
May 16, 2019
Messages
8,055
Reaction score
31,580
Location
Michigan
Rating - 0%
0   0   0
Trident Calcium numbers are notorious for being way off. Alk is solid, but that is it. "Known values" should always match up and they do not even after calibration to the "known solution". Consistency is much more important than precision. Trident offers none of these.
 

xabo

Valuable Member
View Badges
Joined
Aug 21, 2009
Messages
1,035
Reaction score
514
Location
Wash. D.C.
Rating - 100%
1   0   0
I'm not sure I understand the question. Sorry.

I replied why you do not use the display's water for reference when calibrating the Trident. The provided calibration fluid is a "known" value and marked on the bottle. Once opened it should be used right way using the calibration task found in the Apex portal. You run through a series of steps one of which will ask you for the values on the bottle.

If you use the displays values via a manual test kit then those numbers are only as good as your manual test, titration, or auto tool like a bot or Hanna). I don't know about you but I've run back to back tests and my numbers are not the same although close. That is why that task exists and the bottles included.

Lastly don't chase numbers and pick a record of source. In my case I use the Trident. I do send off ATI ICP tests 2 or 3 times a year. When I do this I will grab a sample of water before the Trident runs, run phosphate, nitrate, and alk tests (hanna kits), then pull ATI's sample, and record everything down. Once ATI's results come back I can do a comparison. Each time I've done this the numbers are within their respected margin of error.

Not sure if this helps and apologies if not. About the only other thing that I do is set the calibration level as close as I can to the normal trident sample line. Outside of that I'm not sure other than I tend to trust the calibration more. If in doubt fire off a ticket later today to Neptune and see what they recommend if it is that far off.

Hope you are having a great day!
Will give it a try...................Thanks for detailed information.
 

J1a

Well-Known Member
View Badges
Joined
Nov 14, 2021
Messages
666
Reaction score
946
Location
Singapore
Rating - 0%
0   0   0
I opened a new calibration solution and tested multiple times with 2 seperate test kits and all tests matched (within reason), but didn't match the amount on the bottle. That is when I quit using it.
There is always a possibility that both test kits are wrong.
 

Gtinnel

10K Club member
View Badges
Joined
Mar 20, 2020
Messages
21,401
Reaction score
29,931
Location
Charleston, WV
Rating - 0%
0   0   0
There is always a possibility that both test kits are wrong.
I agree there is, but there is also a chance that the amount listed on the calibration solution bottle is wrong. When multiple tests from two seperate test kits all match then it seems fairly unlikely they're both wrong by about the same amount. Besides even if they are all wrong, they are what I've used to set my tank to 8.5dkh so even if they're wrong at least using them to set my trident will keep my parameters where I've always kept them.

As we all know it's more about stability than the actual values anyway.
 

Jon's Reef

Valuable Member
View Badges
Joined
Jun 5, 2020
Messages
1,100
Reaction score
987
Location
Madison, WI
Rating - 100%
2   0   0
Trident Calcium numbers are notorious for being way off. Alk is solid, but that is it. "Known values" should always match up and they do not even after calibration to the "known solution". Consistency is much more important than precision. Trident offers none of these.
This is also my experience. Alk and Mag close. Ca consistently +50 to +100.

When I swap the 2 month kit, my Ca will be reading 440-450 consistent. Swap reagents, run 3 tests then calibrate. Ca will show 500ish. Check with Hanna... 440-450... just calibrate to the tank water.

I have tried digging deeper with checking reference solutions, multiple tests on multiple samples, ICP taken at the same time. I just come back to doing it this way.
 

Coolcasino

Well-Known Member
View Badges
Joined
Apr 29, 2020
Messages
713
Reaction score
348
Rating - 0%
0   0   0
I gave up on the Neptune calibration solutions a long time ago. You can test with the water after you have done manual tests. The solutions are way off. They aren't even sealed, just a closed cap. Who knows if the values drift. When I use to use the calibration solution the trident wouldn't even end up with those numbers. They would be off... all 3. Calibrated with my water and at the end all the numbers matched.
 

xabo

Valuable Member
View Badges
Joined
Aug 21, 2009
Messages
1,035
Reaction score
514
Location
Wash. D.C.
Rating - 100%
1   0   0
Question: Why can't you use your tank water to calibrate?

Answer: Because it is not a reference solution.

Longer answer: Un-opened calibration fluid has a "known" alk, ca, and mg number. The calibration task on the Apex will ask you to enter these values before it runs the tests. You can use manual test kits, the left over calibration fluid, and compare results to the trident if you want but should do so within 24 hours.
Tried this...........just for fun......
Calibration Fluid Numbers:
Alk 8.55
Ca 475
Mg 1285

Calibration Fluid Tested w/Salifert:
Alk 9.3
Ca 400
Mg 1350

Trident Reading after calibration:
Alk 8.68
Ca 532
Mg 1548

Salifert Reading After calibration/Trident test:
Alk 9.6
Ca 500
Mg 1650

"The Trident tests consistently with a precision of +/- 0.05 dKh for Alkalinity and 15ppm for Calcium and Magnesium. Using any hobby-grade test kit on the market, even humans cannot perform the tests that consistently."
 

Keeping it clean: Have you used a filter roller?

  • I currently use a filter roller.

    Votes: 68 35.1%
  • I don’t currently use a filter roller, but I have in the past.

    Votes: 7 3.6%
  • I have never used a filter roller, but I plan to in the future.

    Votes: 49 25.3%
  • I have never used a filter roller and have no plans to in the future.

    Votes: 62 32.0%
  • Other.

    Votes: 8 4.1%
Back
Top