Using colored pH Buffer solution for probe calibration?

JordanM

Active Member
View Badges
Joined
Mar 26, 2018
Messages
471
Reaction score
163
Rating - 0%
0   0   0

redfishbluefish

Stay Positive, Stay Productive
View Badges
Joined
Mar 22, 2012
Messages
11,713
Reaction score
25,768
Location
Sayreville, NJ
Rating - 0%
0   0   0
Yes, fine to use....and you only use two calibration solutions to calibrate.

For measuring acids.....use 4 and 7
For measuring bases....use 7 and 10
 
OP
OP
J

JordanM

Active Member
View Badges
Joined
Mar 26, 2018
Messages
471
Reaction score
163
Rating - 0%
0   0   0
So i just checked it against the Neptune calibration pH and in the neptune solution it reads 10.0, in the fisher buffer it reads 10.1. In the neptune solution I get 6.98 and in the fisher buffer i get a reading of 7.03. Are these values within tolerance to use the Fisher buffer and what one do you think is actually the better one to calibrate against? I am guessing the Fisher might actually be a little more accurate as this is meant to calibrate lab instruments.
 

Reef of Fillory

Valuable Member
View Badges
Joined
May 19, 2017
Messages
1,133
Reaction score
637
Location
Tampa
Rating - 0%
0   0   0
So i just checked it against the Neptune calibration pH and in the neptune solution it reads 10.0, in the fisher buffer it reads 10.1. In the neptune solution I get 6.98 and in the fisher buffer i get a reading of 7.03. Are these values within tolerance to use the Fisher buffer and what one do you think is actually the better one to calibrate against? I am guessing the Fisher might actually be a little more accurate as this is meant to calibrate lab instruments.
I don't see any issues with that range, pH monitoring should be used as an indicator of something going wrong rather than a magic number anyway. As long as things are consistent and in an acceptable range, there shouldn't be any issues
 