I am in the process of moving to Hydros and set up only the temp and ph probes for now while the Apex is still running. I purchased the Hydros second hand and wanted to ensure there were no issues before I moved everything to the Hydros. I placed the magnetic hydros probe into the same area as the Neptune and it read 76.0F while the Neptune stated 78.2F. I thought, well that is strange and to try to rule out the magnetic Hydros probe, I tossed in the original temp probe that comes with the Hydros. This thermometer read 76.5F. I figured I’d give it a few days to see if they needed to have the temp normalize throughout the probe. There was no change, so I ordered a NIST calibrated thermometer and checked the temp this evening. The NIST thermometer read 76.5F with a deviation of .5F per the NIST certification document. For over 2 years I thought the Apex had the right temp but I was way wrong. Consider this a friendly reminder to calibrate your thermometers from time to time.