Errors Testing PO4 and Nitrate with RSPro

Ghxst

Well-Known Member
View Badges
Joined
Dec 6, 2014
Messages
586
Reaction score
415
Location
Colorado
Rating - 0%
0   0   0
When testing NO3 the results are clear with a metal grey residue, when mixed it separates and clumps back. See pic

The PO4 appears to work and says I have 0.01. Would you agree with the That from the pic?


What can cause the metal grey, does it mean I have no nitrates, too high or just a bad test?

90D5AE85-5EE3-4EE6-937F-83D3240BD21A.jpeg


BD6D0453-6B05-4C53-B5C0-3D7194218373.jpeg
 

2una

Active Member
View Badges
Joined
Dec 27, 2016
Messages
451
Reaction score
403
Rating - 0%
0   0   0
I use the same kit & never seen that before - if i'm measuring my denitrator output that is also zero & don't do that.
Po4 looks about right but hard call working off a photo
 
OP
OP
Ghxst

Ghxst

Well-Known Member
View Badges
Joined
Dec 6, 2014
Messages
586
Reaction score
415
Location
Colorado
Rating - 0%
0   0   0
Thank you ! Good to know it’s not normal.

The grey was also fizzing.
 

XNavyDiver

Insightful answer loading... please wait.
View Badges
Joined
Jun 3, 2017
Messages
1,392
Reaction score
2,753
Location
Sierra Vista, AZ
Rating - 0%
0   0   0
When testing NO3 the results are clear with a metal grey residue, when mixed it separates and clumps back. See pic

The PO4 appears to work and says I have 0.01. Would you agree with the That from the pic?


What can cause the metal grey, does it mean I have no nitrates, too high or just a bad test?

90D5AE85-5EE3-4EE6-937F-83D3240BD21A.jpeg


BD6D0453-6B05-4C53-B5C0-3D7194218373.jpeg
I get that same grey metal looking specks with the RSPro NO3 test. I have no idea what they are.
 

Lasse

10K Club member
View Badges
Joined
Mar 20, 2016
Messages
10,887
Reaction score
29,890
Location
Källarliden 14 D Bohus, Sweden
Rating - 0%
0   0   0
I have seen that too - both in coloured samples (nitrate) and not coloured (no nitrate). It comes and goes - even with the same batch of test

Sincerely Lasse
 

2una

Active Member
View Badges
Joined
Dec 27, 2016
Messages
451
Reaction score
403
Rating - 0%
0   0   0
You guys seeing this in the older greyish colour box or the newer more black colour box?
I'm black box (algae control pro which has both No3 + Po4) - never seen that happening
 

XNavyDiver

Insightful answer loading... please wait.
View Badges
Joined
Jun 3, 2017
Messages
1,392
Reaction score
2,753
Location
Sierra Vista, AZ
Rating - 0%
0   0   0
You guys seeing this in the older greyish colour box or the newer more black colour box?
I'm black box (algae control pro which has both No3 + Po4) - never seen that happening
Black box.
 

roberthu526

Valuable Member
View Badges
Joined
Jan 11, 2016
Messages
1,663
Reaction score
1,221
Rating - 0%
0   0   0
I get the same residue when testing for nitrate. It dissolves after adding C if you have nitrate in the water. If you truely don’t have nitrate you will see this residue sitting there after 9 minutes.
 

roberthu526

Valuable Member
View Badges
Joined
Jan 11, 2016
Messages
1,663
Reaction score
1,221
Rating - 0%
0   0   0
What I have noticed is B dissolves in test water and releases a lot of bubble. The residue forms after adding C and it releases bubble again which makes the water cloudy. I think this is why Red Sea says wait for 9 minutes then swirl again then read the result so the color is not impacted by the bubbles.
 
OP
OP
Ghxst

Ghxst

Well-Known Member
View Badges
Joined
Dec 6, 2014
Messages
586
Reaction score
415
Location
Colorado
Rating - 0%
0   0   0
Thanks everyone for the heads up and tips. I tried again and it did the same thing. I also have the 2 set kit in a black box. I tried with the lights out and on all day to see if something was using the NO3 and still clear with grey chunks. When I test again I’ll wait 9 extra mins but my instructions say the color is only stable 5 mins after the wait time.
 
OP
OP
Ghxst

Ghxst

Well-Known Member
View Badges
Joined
Dec 6, 2014
Messages
586
Reaction score
415
Location
Colorado
Rating - 0%
0   0   0
I get that also, it's from the step b powder.


Do you have anything kit to verify zero nitrates? Or a LFS to check it maybe

I can bring it to a lab for ICP and field testing using a hach. I’ll test simultaneously to verify
 

2una

Active Member
View Badges
Joined
Dec 27, 2016
Messages
451
Reaction score
403
Rating - 0%
0   0   0
When I test again I’ll wait 9 extra mins but my instructions say the color is only stable 5 mins after the wait time.

Just checking

16ml water x 2
5 drop "A" - shake 15 sec
1 spoon "B" - shake 60 sec
1 spoon "C" - shake 15 sec
Wait 9min for reaction to do its thing.
Shake again - take reading
 

Luno

Well-Known Member
View Badges
Joined
Nov 16, 2017
Messages
536
Reaction score
591
Location
Australia
Rating - 0%
0   0   0
Just checking

16ml water x 2
5 drop "A" - shake 15 sec
1 spoon "B" - shake 60 sec
1 spoon "C" - shake 15 sec
Wait 9min for reaction to do its thing.
Shake again - take reading

That's correct also try shaking immediately like immediately immediately after using B and C. I've noticed if i put the spoon in cap and shake immediately the particles are less to none. If I let it sit for say 15 seconds while I put the cap on the powder bottle than put that cap on the vial and shake it it leaves more residue
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
67,349
Reaction score
63,689
Location
Arlington, Massachusetts, United States
Rating - 0%
0   0   0
I'm not sure exactly which method Red Sea uses, or what might cause this, but many kits use a cadmium metal (a grey metal) reduction method, and Hach says:

https://www.google.com/url?sa=t&rct=j&q=&esrc=s&source=web&cd=3&cad=rja&uact=8&ved=0ahUKEwjrlbvLmc3XAhUM54MKHccnAfEQFggyMAI&url=https://www.hach.com/asset-get.download.jsa?id=7639983736&usg=AOvVaw1bU9pEKOoqloCwjXBB1SCj

"
The reagents that are used in this test contain cadmium. Rinse the sample cell immediately after use to remove all cadmium
particles. Collect the reacted samples for proper disposal.

A deposit of unoxidized metal will remain at the bottom of the sample cell after the reagent dissolves. The deposit will not
affect results."
 
OP
OP
Ghxst

Ghxst

Well-Known Member
View Badges
Joined
Dec 6, 2014
Messages
586
Reaction score
415
Location
Colorado
Rating - 0%
0   0   0
Thanks Dr.! Exactly what I couldn’t find. I’ll use a magnetic stirer next time incase it’s an issue. I really wish the test kits had an option for weighing the regents too.
 

Hans-Werner

Valuable Member
View Badges
Joined
Aug 24, 2016
Messages
1,501
Reaction score
2,295
Rating - 0%
0   0   0
If Red Sea still uses cadmium for reduction of nitrate this has to be stated on the package. If there is no warning on the package they likely use zinc for nitrate reduction and it is zinc that forms the particles.
 