http://wamas.org/forums/topic/41663-diy-10-dkh-alkalinity-standard/
500g rodi, 7.5g baking soda
then using 10g of that solution mixed with 490g rodi
Tested with the Hanna to verify. The one I mixed up came back 183ppm or 10.248dKH.
Within the 5% error, but 5% of the measurement is 9.15ppm or .51 dkh
I ran just the 10dKH standard on the Guardian after I purged the sw line and the chamber and it ran 8 tests before running out of standard
1-8.00
2-8.00
3-8.00
before the 4th test I changed the KH auto correction to 8, was previously set at 6.5 from running on tank. and had gotten
KH > ERR for test 1-3. I also got KH>ERR on test 4. I went back and manually changed the last test number to 8.0. I think that sometimes it doesn't update this if there is Any error during the test. even though KH is most likely due to being greater than 1.5 from the auto correction number.
4- 9.00
5-9.20
6-9.20
7-9.40
8-9.40
So I placed Guardian back on tank and let it run every 30 min for 7 hours. Giving me a range 6.60-6.70. This morning I +1.4 ADJ KH.
Which brought it up to 8.00 the same numbers I am getting with Hanna 142/143ppm 7.952 &8.008dKH. Lights are off during the day I am going to continue to allow Guardian to run every 30 minutes. I will come back and test with Hanna every couple hours to verify the numbers. I would expect the numbers to climb since whole system in on reverse cycle. Running 60ml /min from masterflex pump on Reef Octopus CR5000D reactor set at 6.60 pH.