PART 4 "GETTING IT RIGHT" COLORIMETRIC INSTRUMENTAL TESTING METHODS (DIGITAL)

As we turn our attention to the use of digital testers (Colorimeters) such as the HANNA CHECKER, most of the same elements necessary for good precision and accuracy that exist in Visual Testing also exist in Digital Testing. Digital Testing methods have the same two steps as the Visual method: Evaluation (The Chemistry) and Assessment. Evaluation is the test procedure itself and Assessment is the determination of the results.

TESTING PROCEDURES (EVALUATION)

Most test kits consist of a series of reagents added to the sample according to a prescribed procedure. The procedure generally requires that the reagents be measured out by “drops”, ml or some type of measuring spoon. They are added to the sample being tested in prescribed order. There is usually a time element associated with the procedure…Mix for 1 min…wait for 30 seconds…etc. Part of the procedure using digital tests most often requires the measurement of a “Blank”. The Blank is the initial measurement that is taken in order to compensate for the “color” of the test sample itself (Water Color). After the last step in the procedure the sample is placed into the device for measurement. The device measures the light absorbed by the sample across a defined spectral range (generally defined by the device supplier). That data is then run through an algorithm to calculate the sample value which is displayed on the device. . . It is important to note that the algorithm calculates the correct value only for clear solutions, with no light scattering from suspended particles (Turbidity). The calculation is based on the Beer’s Lamberts Law which states “The concentration is directly proportional to the absorption of light at a given wave length. Light scattering by particles or bubbles mimics light absorbance, resulting in an incorrect value.

Here is a link to an article written by Kevin Costa of Hanna Instruments that is a good overview of how the Hanna Checkers (Colorimeter) work https://www.reef2reef.com/ams/understanding-colorimetry-and-your-hanna-checker®-hc.341/

ASSESSMENT
Assessment using a digital measuring device is much easier than visual assessment… No color evaluation necessary…Just read the digital output to obtain your value.

SOURCES OF ERROR IN THE TESTING (EVALUATION) PROCEDURE

The sources of error in the testing procedure for digital testing are much the same as the visual testing methods with exception of the Instrument.

1) Inaccurate reagent measurement
a. Inaccurate measuring devices
i. Droppers
ii. Syringes/Pipettes
iii. Spoons
iv. Other
b. Procedural error in reagent measurement
i. Not reading the measurement meniscus (A Parallax Error-See Part 2)
ii. Not holding droppers vertical for dispensing
iii. Air bubbles in drops when dispensing
iv. Measuring Spoon over/under fill

(For Examples See Table 1-6 and Fig 9-10 in PART 2 of this article)

2) Instrument (Meter---Hanna Checkers or other ) Errors
a. Procedural errors with instrument
i.Vials not clean on exterior; finger prints water marks & or scratched
ii. Not positioning the vials the same way every time
iii. Using the same vial for multiple tests. (Label the Vials to prevent this See Fig 21)
iv. Not covering the top of the meter to keep out stray light
v. Inside of vials and caps contaminated and not properly cleaned & rinsed
vi. Micro-Bubbles in test vile causing light scattering
vii. Suspended material in vile causing light scattering
b. Instrument Errors
i. Does not meet the calibration test standard—Out of Calibration
ii. Damaged Instrument
iii. Test Meter not intended for testing salt water

The tables and illustrations below are some examples of how some of the above errors can affect the results. The precision statement is the Mean ± the Range (High Minus Low)

EXTERIOR OF VIAL DIRTY


TEST #12345PRECISION
ppb915217812 ± 14
TABLE 9


EXTERIOR OF VIAL CLEAN


TEST #12345PRECISION
ppb10129101210.6 ± 3
TABLE 9A

RANDOM POSITIONING OF VIAL


TEST #12345PRECISION
ppb9131281010.4 ± 5
TABLE 10

CONSISTENT POSITIONING OF VIAL


TEST #12345PRECISION
ppb910812910.2 ± 2
TABLE 10A

As you can see from Tables 9 to 10A a clean vial and consistent positioning increases the precision .To help with positioning of the vials I placed a mark on the vial and the Checker. I line them up when I place the vial in the checker. (See Fig 20) I also label the vials to prevent cross contamination. (See Fig 21)


1577816221630.png

FIG 20

1577816288939.png
1577816304333.png

FIG 21


VIAL RINSING & CONTAMINATION

1577816479412.png

FIG 22

Fig 22 Explanation

Sample vile was emptied then placed on an absorbent material (actually toilet paper) and allowed to drain. It was then rinsed and emptied and again allowed to drain. This was repeated 3 times. At the end of the second rinsing a small amount of contamination is still visible and a third rinse was required. This would indicate a good “best” practice would be 3 rinses. It should also be noted that the cap of the vile needs also to be rinsed because it can be a source of contamination. (Note on contamination: 1ppm is roughly 7 drops in 60 gal so that means a very-- very small amount is required to contaminate a 10 ml test vial--- .000001mL)


CONCLUSIONS ON DIGITAL TESTING METHODS


Many of the same conclusions that we saw in the Visual Testing can be applied to Instrumental (Digital) Testing. The key elements are the same. These are covered in the section, “Establish Good Laboratory Practices (EVALUATION)” in the Visual Testing section Good laboratory “best practices” apply here also. (See PART 3--CONCLUSION ON VISUAL TESTING METHODS) The main difference is the instrument itself.

KNOW YOUR INSTRUMENT:

It is important to understand the performance statistics of the instrument (meter) you are using as well as the instrument care, calibration and correct usage. Here are some things to consider when using an instrument that can help in utilizing “best practices”:

1) What is the stated accuracy?

This important to know the significance of the value you get as a result of a test. Knowing the significance of the value helps to make decisions on any actions that might be necessary. For example the Hanna Calcium Checker has a reported accuracy of ±6% of the reading. If the reading is 400ppm this could actually be 424ppm or 376ppm. This would mean that the “actual” value could be between these two values. One way of better quantifying this value for an individual instrument is to run tests on known solutions or standards. By doing this the accuracy can be better defined. Once this value is know as long as the instrument has good repeatability a correction factor can be applied to get more accurate results. For example, my Calcium Checker consistently measurers’ 15-20 ppm lower when compared to a known standards of 350 & 400 ppm…(4-5%) The key word here is “consistently”. When measuring a sample I simply multiply the value obtained by 1.05 and get my results. Using this method I am consistently within ± 10ppm of external testing sources results.

All of the Hanna Checkers have reported accuracy values. They are stated either a % of the meter reading or as a ± ppm or ppb value plus a % of the meter reading. For example the HI 736 Phosphorous Ultra Low Range has a reported accuracy of ±5ppb ±5% of the reading. A meter reading of 10 would then have a range of 10ppb ±5ppb plus 5% of 10ppb = .5ppb…Meaning the “actual” value would be between 4.5ppb and 15.5ppb. Again by “calibrating” the meter with known solutions the “true” value can better be determined for that particular meter.

It is important to note, accuracy must be determined by measuring against a known sample. Obtaining known standards is a challenge. Reference Calibration Standards for our hobby (seawater) are not easy to come by…Especially certified ones…and they are expensive!...Here is an example https://highpuritystandards.com/elements-in-seawater/ (This is the one used in the article by Rich Ross & Dr. Chris Maupin in Skeptical Reefkeeping). There are other ones that you can get, they are not certified but verified via ICP testing…Like this one https://www.bulkreefsupply.com/fauna-marin-multi-reference-solution.html I generally prepare my own standards. The point is you have to have an good target in order to know if your test is giving accurate results. So the precision can be high (good repeatability and reproducibility) but it is precisely incorrect.

2) What is the precision of the instrument?


Precision is how close two or more measurements are to each other regardless of the accuracy of the measurement. It is important that an instrument be able to measure the same sample multiple times and get the same (or very close) results. If this is not the case it becomes very difficult to establish accuracy.

One method to determine the Precision of an instrument is by measuring the same sample with the instrument multiple times. The values are then added together and then divided by the total number of samples. This gives the Mean (Avg.) of the sample set. The lowest value in the set is subtracted from the highest value in the set to get the “range” The precision is then given as Mean ± Range value. (See Example &Table 11)

Example: HI 736 Phosphorous Ultra Low Range Precision Test

MEASUREMENT 123456MEAN
VALUE ppb10510299101100102101.5
TABLE 11

RANGE = 105-99= 6 PRECISION STATEMENT 101.5 ± 6ppb


The test was conducted using the Hanna Calibration Check Sample Set as the sample to be measured. These sets are available for all of the Hanna Checkers and are useful for precision testing as well as checking to see if the instrument is within its calibration range.

There are other methods to evaluate the precision and if you are interested here is a link to see how they are done. https://www.wikihow.com/Calculate-Precision

3) Is the instrument subject to EMC Deviation?

EMC Deviation is the deviation in accuracy due to Electro-magnetic interference. Using the instrument near an Electro-magnet source (unshielded electric motor, computer screen, some fluorescent light fixtures etc.) can cause errors in the readings. The specifications sheet for the instrument will let you know if this is an issue. For example the HI-736 Phosphorous Checker has an EMC Deviation of ±5ppb.

4) Run periodic calibration checks on the instruments.

Checking to see if the instrument is measuring in an acceptable range will help to insure the dependability of the results. The Hanna Calibration Check Samples are designed for this purpose. I check mine monthly or if I suspect there might be an issue with erratic measurements. Keep a control chart (record) of your calibration results over time. This will help you to track the instruments performance over time and help to spot any trends.

5) Read the operating manual of the Instrument.

Carefully read the operating manual of the instrument. Take note of things like operating environment, calibration requirements, error codes and other operational instructions.

All of these and most likely more are steps to take to insure the measurement is as Accurate & Precise as possible. Having measurements be as Accurate and Precise as possible coupled with best laboratory practices provides reliable information about the water quality of an aquarium. This will give a higher degree of confidence in taking any corrective action required.

BRS did a video on along these lines (Link Below) that also illustrates this point. The one issue I have with their approach is that they defined “Accuracy” as the average of the reading subtracting out the highest and lowest 10% of reading to get the target value. I do not think this defines accuracy. A valid reference standard would have been a better approach, but none the less it is valuable to illustrate the point. https://www.bulkreefsupply.com/video/view/can-you-trust-your-saltwater-test-kit-or-your-testing-skills-on-your-reef-tank-brstv-investigates/?browse_eid=b3e52ff6-9fa6-460d-900f-e636559973d2&utm_source=bronto&utm_medium=email&utm_term=WATCH+VIDEO&utm_content=[VIDEO]+Can+you+trust+your+test+kit?&utm_campaign=Video+071919&_bta_tid=00891137605476391548835156685805154972018283493119219376896555356694754057504034082108396405036106573321

ICP TESTING

ICP testing of salt water has become very popular in the last few years. That being the case a few things to think about with regards to the results obtained from this method. ICP (Inductively Coupled Plasma) Spectroscopy is an analytical method used to detect and measure elements to analyze chemical samples. The process is based on the ionization of a sample by extremely hot plasma, usually made from argon gas. ICP comes in different “flavors”.

ICP AES
Inductively Coupled Plasma Atomic Emission Spectroscopy (ICP-AES) or ICP Atomic Emission Spectroscopy is a technique that can determine concentrations of trace to major elements and can detect most elements in the periodic table. Reliable results can be obtained for about 70 elements with detection limits in the parts per billion ranges.

ICP MS
Inductively Coupled Plasma Mass Spectrometry (ICP-MS) or ICP Mass Spectrometry is highly sensitive and capable of multi-element trace analysis and ultra trace analysis, often at the parts-per-trillion level. Testing for trace elements can be performed on a range of materials from super alloys to high purity materials.

ICP OES (This is the method most widely used for testing salt water samples)

ICP-OES (Inductively coupled plasma - optical emission spectrometry) is a technique in which the composition of elements in (mostly water-dissolved) samples can be determined using plasma and a spectrometer.

The important factor to consider with ICP testing is that it is a Method. That means just like the Methods covered above (Visual Testing & Instrumental Testing) it is subject to many of the same errors…Errors in sample preparation…Errors in instrument calibration…Procedural Errors..ETC. Doing an ICP test is not a trivial matter. It is not just a complex instrument that you put your sample in and out pops the answer. It consists of procedures and methods that need to be followed precisely in order to get accurate results. I say this to point out that although ICP tests produce large amounts of numerical data, reporting in ppm and ppb it is subject to errors and incorrect methods just like our Hobby Grade Kits. This mean
that ICP testing is subject to precision and accuracy ranges the same as the previous test discussed. Rich Ross and Dr. Chris Maupin published a great article in Reefs Magazine titled “Skeptical Reefkeeping 12: Triton Lab ICP-OES Testing of a Certified Artificial Saltwater Standard”. This article illustrates this point. ---Here is a link to the article
http://packedhead.net/2015/triton-lab-icp-oes-testing-of-a-certified-artificial-saltwater-standard/

The key points to remember is the ICP testing are subject to variations and errors and therefore are not “absolute” values.

For the last several months I have been sending in my water samples to 3 different ICP test facilities and recording the data for comparison. All 3 facilities state that they have high “accurate” and “precision”. As the article by Ross & Maupin illustrates this is not totally the case in all the areas of analysis. By looking at 3 different ICP services (Identified as ICP 1-3) it becomes apparent that not all ICP testing is created equal…at least from my point of view. The table (Table 12) below is a comparison of 3 different ICP testing facilities testing the same sample. 3 samples were taken on each of the indicated dates and shipped to the ICP test facility on the same day. Many of the values are close, but some are not. The question then becomes which one is correct. This becomes a difficult question to answer because the data does not include any specific accuracy statements for each of the elements tested. From a Quality System perspective this presents a problem. If the reported value is let’s say .1 ppm what is the possible error in this value based on the accuracy statement is it ± .01 ppm? or ± .05 ppm? These two could have completely different responses to the reported value if let’s say we are dealing with Iodine. The Yellow cells in Table 12 are the elements most commonly monitored and dosed for adjustments.

I want to be clear. I am not negative on ICP testing. I think it brings a new level of testing to our hobby. That being said I think a clear statement to as to the expected accuracy for each of the elements would be very helpful in making decisions as to any actions we should be taking based on the results.

1577818031231.png

TABLE 12

FINAL CONCLUSIONS


We have covered a great deal in this 4 part series and there is a lot more that could be said. Hopefully you have found some valuable tips to help improve the Accuracy & Precision of your water testing.

As was mentioned in the introduction; Aquarium water testing is an analytical method. Just like any other analytical method it shares the challenge of getting good Accuracy and Precision. By approaching our testing from a Quality Systems mind set, we can improve our accuracy and precision and have confidence in our results to provide our tank critters with the best environment possible.


1577818155473.png
 
Last edited by a moderator:

Dan_P

5000 Club Member
View Badges
Joined
Sep 21, 2018
Messages
6,675
Reaction score
7,170
Rating - 0%
0   0   0
This article or rather series of articles was quite a generous gift to us. Thanks!

After reading the series, I almost feel as though you are looking over my shoulder when I run a test, and when I start to consider taking a short cut, I can hear you TSK TSK’ing behind me :)
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
This article or rather series of articles was quite a generous gift to us. Thanks!

After reading the series, I almost feel as though you are looking over my shoulder when I run a test, and when I start to consider taking a short cut, I can hear you TSK TSK’ing behind me :)

You are most kind Sir...Actually that is not me looking over your shoulder...It's Scientist conscience coming back to haunt you :cool:
 

ingchr1

Valuable Member
View Badges
Joined
Sep 9, 2018
Messages
1,498
Reaction score
1,127
Rating - 0%
0   0   0
Nice article, a lot of good info on test practices.

I've been rinsing my vials twice after use, looks like I need to add a third. I also rinse with sample water for each test.

I also need to run my checkers through the Hanna standards that I picked up. Thanks for the link to the Fauna Marin reference, may look into that the next time I place an order with BRS.
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Nice article, a lot of good info on test practices.

I've been rinsing my vials twice after use, looks like I need to add a third. I also rinse with sample water for each test.

I also need to run my checkers through the Hanna standards that I picked up. Thanks for the link to the Fauna Marin reference, may look into that the next time I place an order with BRS.

Thank you very much. I'm happy that you found it useful. Have you had a chance to read Parts one two and three they have some additional suggestion that might be useful
 

kecked

Active Member
View Badges
Joined
Dec 9, 2017
Messages
380
Reaction score
218
Rating - 0%
0   0   0
Nice job that’s a mini book! Being a chemist I look at icp3 and say they need to find the contamination in their setup. Something is wrong. Finding all those trace metals tells me likely they have a calibration issue. They need to run spikes dups splits and blanks and provide qc data along with the sample data before I’d trust those results. For most analytics being off by 1ppm is fine like waste water or hazardous waste. For fish tanks 1ppm might be the desired limit. That phosphate is majorly off.

anyway great job. I was hoping for one thing more. How to adapt the existing calorimeters to measuring nitrate with a visual kit. It must be possible to make a calibration and run samples. why does no one offer one? What is the issue? I did do it on an old beckman du with no issue.
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Nice job that’s a mini book! Being a chemist I look at icp3 and say they need to find the contamination in their setup. Something is wrong. Finding all those trace metals tells me likely they have a calibration issue. They need to run spikes dups splits and blanks and provide qc data along with the sample data before I’d trust those results. For most analytics being off by 1ppm is fine like waste water or hazardous waste. For fish tanks 1ppm might be the desired limit. That phosphate is majorly off.

anyway great job. I was hoping for one thing more. How to adapt the existing calorimeters to measuring nitrate with a visual kit. It must be possible to make a calibration and run samples. why does no one offer one? What is the issue? I did do it on an old beckman du with no issue.


Thank you very much...Coming from an industrial R&D background my passion is for continuous improvement...Continuously working for more accurate and precise results...

Here is a link to an article on nitrate measurement using an existing Hanna Colorimeter HI-764 this should do what you are asking for if I understand your question...https://www.reef2reef.com/ams/test-meter-for-testing-nitrates.599/

There is also one that was developed for iodine... https://www.reef2reef.com/ams/using-hanna-checker-hi-707-to-test-for-iodine.743/
 

kecked

Active Member
View Badges
Joined
Dec 9, 2017
Messages
380
Reaction score
218
Rating - 0%
0   0   0
Perfect. Exactly what I had in mind. You know it’s stupid easy to put an rgb led inline with a 3d printed cuvette holder. Pop a solar cell on the other side and add a volt meter to it. Sure you need to make cal curves but that’s all there is to making a basic spectrophometer. I think I will use your method but instead use a Haloven light through a 1cm quartz cuvette and run that into an ocean optic usb2000 spec. I can see the whole spectrum. Take a blank before adding reagents and then run a set of calibrations then sample. Lot of work for a nitrate test. Do you think the colors are stable to store the calibration samples? I could save the curves from the spec. Ocean view let’s you setup a method.

my issue is I’m trying to determine 40ppm vs 80ppm. I cant see it by eye. I run the tank at 20 usually as I had a Dino infestation and now run high nutrients. Works well. Ps vibrant kills monitporia even 1ml in 55gal....just saying.

i see why nitrite tester...that’s what you are actually measuring.
 

revhtree

Owner Administrator
View Badges
Joined
May 8, 2006
Messages
47,787
Reaction score
87,409
Rating - 100%
1   0   0
Thank you for taking the time and effort in putting these together!
 

WallysWorld

Active Member
View Badges
Joined
May 7, 2018
Messages
223
Reaction score
110
Location
Edmonds, WA
Rating - 0%
0   0   0
I'm a metrologist (no not a weatherman!) and it's very nice to see the basic principals of metrology being applied to our hobby measurements
 

WallyB

REEF Techno-Geek
View Badges
Joined
Jan 24, 2017
Messages
3,127
Reaction score
8,094
Location
GTA Toronto, CANADA
Rating - 0%
0   0   0
Just stumbled on this article.

Well written. Covered off all the key points very well! (With Supportive Data)

Your Rinse # test was most valable to me, since I autmoted my Hanna Egg for Alk Testing and programmed two rinses thinking it would be enough. Just changed setting to 3 Rinses after seeing your Paper Towel Test.

Thanks.
 
Last edited:
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Just stumbled on this article.

Well written. Covered off all the key points very well! (With Supportive Data)

Your Rinse # test was most valable to me, since I autmoted my Hanna Egg for Alk Testing and programmed two rinses thinking it would be enough. Just changed setting to 3 Rinses after seeing your Paper Towel Test.

Thanks.

Thank you....appreciate it...
 

WallyB

REEF Techno-Geek
View Badges
Joined
Jan 24, 2017
Messages
3,127
Reaction score
8,094
Location
GTA Toronto, CANADA
Rating - 0%
0   0   0
Thank you....appreciate it...
My Pleasure.

I do have one question since you commented and made comparsions on percision of the Hanna Testers.

I tested my Hanna Alk Tester to death. Reason is it's the Core Component in my DIY Automated Alk Monitor.
I didn't have a good reference solution when I finished it and began the testing/validataion.
Made some DIY rough reference Solution using Salefert Test Kit..... asked Fellow Reefers to give me a water Samples and their Trident Results, Mixed up Salt with known approx Alk (with dead on Salinity), etc.

Those tests prove the Hanna Alk Tester is Consistent. It also proved my Alk monitor design is reliable since it's test results are good as the Hanna (test by Hand).

The Hanna Alk testers as you yourself mentioned are consistent.
That's good enough and more important. Stability is what we persue in our tank. Accuracy can be a bit off.

Yet, I've used Salifert Alk Test Kit for years. Trusted it, still do.
So I compared the Hanna to the Salifert during my testing. To test different levels of Alk. (That my tank may experience. Ranges between 6.5 and 9.0 dkh.
Both are consistent. BUT!!! About 0.5 dkh off (Salifert being the higher result).

I even bought a 2nd Hanna Alk tester. It matches the first Hanna.

What perplexes me is sometimes they (Salifert & Hanna) match.
It's like the Hanna has a sweet spot. I don't remember the value, but it around 7.x dkh. Maybe it's related to it's central calibration point. Or a the Hanna Regeant. Or PH of (reference, or Tank water) since the Hanna is PH senstive. No idea.

I even thought my Salifert Testing method was off. The Titration Color change point from Blue to Pink.
There is a thread on this forum that illustrates the transition isn't the pink point, it's just before. (can't find that thread right now)
That testing method change got the Hanna and Salifert match, at certain Alk levels, but again there is a shift away when you get around 8.0dkH.

With the Fauna Multi Reference Solution that is 6.6 dkH. Hanna is dead on. So is Salifert.

My querk is my SPS tank runs around Hanna 8.0dkH (Target), and The Salifert gives me 8.5 dkh. ALWAYS!!
I would like to run my tank close to 7.5dkh, then then what if the Hanna's 7.5, is truely 7.0 dkH.
Does it really matter? No.... but having my tank running below 7.0 during a spike in Alk consuption isn't my confort zone. Thus I'm stuck running my tank at 8.0 Hanna, and 8.5 Salifert.

If the multi reference solution from Fauna was 7.5, I'd be set to go with the Hanna trusted result.
My Hanna based Alk Monitor, which about to be Married to a DIY Smart Doser (controlled by the Monitor).....would be perfectly (dead on).

I wonder why Fauna pick 6.6 dkh for a Reference solution.

Hope I explained my (oddity) observation well...

Have you done any testing comparisons and noticed that.? Specifically with the HANNA Alk tester.
 
Last edited:
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
My Pleasure.

I do have one question since you commented and made comparsions on percision of the Hanna Testers.

I tested my Hanna Alk Tester to death. Reason is it's the Core Component in my DIY Automated Alk Monitor.
I didn't have a good reference solution when I finished it and began the testing/validataion.
Made some DIY rough reference Solution using Salefert Test Kit..... asked Fellow Reefers to give me a water Samples and their Trident Results, Mixed up Salt with known approx Alk (with dead on Salinity), etc.

Those tests prove the Hanna Alk Tester is Consistent. It also proved my Alk monitor design is reliable since it's test results are good as the Hanna (test by Hand).

The Hanna Alk testers as you yourself mentioned are consistent.
That's good enough and more important. Stability is what we persue in our tank. Accuracy can be a bit off.

Yet, I've used Salifert Alk Test Kit for years. Trusted it, still do.
So I compared the Hanna to the Salifert during my testing. To test different levels of Alk. (That my tank may experience. Ranges between 6.5 and 9.0 dkh.
Both are consistent. BUT!!! About 0.5 dkh off (Salifert being the higher result).

I even bought a 2nd Hanna Alk tester. It matches the first Hanna.

What perplexes me is sometimes they (Salifert & Hanna) match.
It's like the Hanna has a sweet spot. I don't remember the value, but it around 7.x dkh. Maybe it's related to it's central calibration point. Or a the Hanna Regeant. Or PH of (reference, or Tank water) since the Hanna is PH senstive. No idea.

I even thought my Salifert Testing method was off. The Titration Color change point from Blue to Pink.
There is a thread on this forum that illustrates the transition isn't the pink point, it's just before. (can't find that thread right now)
That testing method change got the Hanna and Salifert match, at certain Alk levels, but again there is a shift away when you get around 8.0dkH.

With the Fauna Multi Reference Solution that is 6.6 dkH. Hanna is dead on. So is Salifert.

My querk is my SPS tank runs around Hanna 8.0dkH (Target), and The Salifert gives me 8.5 dkh. ALWAYS!!
I would like to run my tank close to 7.5dkh, then then what if the Hanna's 7.5, is truely 7.0 dkH.
Does it really matter? No.... but having my tank running below 7.0 during a spike in Alk consuption isn't my confort zone. Thus I'm stuck running my tank at 8.0 Hanna, and 8.5 Salifert.

If the multi reference solution from Fauna was 7.5, I'd be set to go with the Hanna trusted result.
My Hanna based Alk Monitor, which about to be Married to a DIY Smart Doser (controlled by the Monitor).....would be perfectly (dead on).

I wonder why Fauna pick 6.6 dkh for a Reference solution.

Hope I explained my (oddity) observation well...

Have you done any testing comparisons and noticed that.? Specifically with the HANNA Alk tester.

I have not done any comparison with the Alk Tester...But have done it with the calcium tester....If I understand your question correctly, you are wondering why on some occasions the two tests match and why on other occasions they do not. I would guess it has to do with the variability (error) range of each tests. They most likely overlap and just by "chance" they land on the same value. This will only occur if the 2 tests yield results in the same testing range...I observed this in my work with the Calcium Tester and a test that required titration...I was working to get the test protocol of the Calcium tester to give me more accurate and precise results....Although they did not agree every time there were occasions where they did.....hope this makes sense


rick

PS...I totally love your DIY Alk Monitor!!....You are a very talented reefer !!
 

WallyB

REEF Techno-Geek
View Badges
Joined
Jan 24, 2017
Messages
3,127
Reaction score
8,094
Location
GTA Toronto, CANADA
Rating - 0%
0   0   0
I have not done any comparison with the Alk Tester...But have done it with the calcium tester....If I understand your question correctly, you are wondering why on some occasions the two tests match and why on other occasions they do not. I would guess it has to do with the variability (error) range of each tests. They most likely overlap and just by "chance" they land on the same value. This will only occur if the 2 tests yield results in the same testing range...I observed this in my work with the Calcium Tester and a test that required titration...I was working to get the test protocol of the Calcium tester to give me more accurate and precise results....Although they did not agree every time there were occasions where they did.....hope this makes sense


rick

PS...I totally love your DIY Alk Monitor!!....You are a very talented reefer !!
Thanks.

Just to close on this topic.

The Match isn't by chance (ie sometimes). The Match is consistant.

ALK(HANNA) = 8.0, ALK(Salifert) =8.5
ALK(HANNA)= 6.6 ALK(Salifert) =6.8 (Against Fauna Reference Sample=6.6)

Every test. Various water samples.

My alk MOnitor can do the offset (conversion) in the Report, but I would have to develop a table as things shift. I actually think the Salifert is the correct ALK (WHen there is a mismatch).

It's hair splitting...... CONSISTANCY is most imporant.
YOUR ARTICLE TACKLES THAT (BETTER CONSISTANCY) !! So thanks again.
 