pH probe not calibrating

zbrusko

Active Member
View Badges
Joined
Mar 14, 2023
Messages
437
Reaction score
235
Location
CT
Rating - 0%
0   0   0
I’m attempting to calibrate my Apex pH probe. I follow the steps. Works fine at 7.0. Near the end of the 10.0 calibration, it says it cannot calibrate, and tells me to make sure AOS is up to date, which it is. Not sure how to proceed. Any thoughts?
 

BryanM

7500 Club Member
View Badges
Joined
Jun 27, 2024
Messages
7,884
Reaction score
9,689
Location
Morgan Hill
Rating - 100%
1   0   0
I’m attempting to calibrate my Apex pH probe. I follow the steps. Works fine at 7.0. Near the end of the 10.0 calibration, it says it cannot calibrate, and tells me to make sure AOS is up to date, which it is. Not sure how to proceed. Any thoughts?
This was always a problem for me as well. sometimes after 4 or 5 tries it would actually work. Super frustrating
 
OP
OP
zbrusko

zbrusko

Active Member
View Badges
Joined
Mar 14, 2023
Messages
437
Reaction score
235
Location
CT
Rating - 0%
0   0   0
I’m attempting to calibrate my Apex pH probe. I follow the steps. Works fine at 7.0. Near the end of the 10.0 calibration, it says it cannot calibrate, and tells me to make sure AOS is up to date, which it is. Not sure how to proceed. Any thoughts?
This was always a problem for me as well. sometimes after 4 or 5 tries it would actually work. Super frustrating
So I’ll just keep trying then. Thanks.
 

Tangdora

Well-Known Member
View Badges
Joined
Feb 8, 2020
Messages
724
Reaction score
394
Rating - 0%
0   0   0
If you have not gotten this to work yet Would try resetting the calibration back to default, rebooting and then trying the calibration.
 

Thrium

New Member
View Badges
Joined
May 15, 2026
Messages
24
Reaction score
4
Location
Texas, USA
Rating - 0%
0   0   0
Classic "probe failing at the upper end" pattern — usually one of three things:

1. Buffer solution age. pH 10 absorbs CO2 from air much faster than 7.0 and goes off in months once opened. If your 10.0 bottle has been open a while, try a fresh single-use sachet — that fixes it more often than people expect.

2. Probe age. Apex pH probes are usually good for 1-2 years before slope drift makes 10.0 calibration unreliable. The probe still reads fine at pH 7 (middle range) but the millivolt response gets compressed at the extremes, so 10.0 fails the slope check.

3. Probe glass cleanliness. Salt creep, biofilm, or even fingerprints on the bulb can throw off the upper end more than the lower. Soak the tip in pH 4 buffer or a commercial probe cleaner for 30 min, rinse with RO, then try again.

Order I'd try: fresh 10.0 buffer first (cheap, fast), then clean the probe, then assume the probe needs replacement.

How old is the probe?
 

TOP 10 Trending Threads

HOW DO YOU ADJUST YOUR CUC AS ALGAE DISAPPEARS?

  • Capture and re-home CUC

    Votes: 7 7.1%
  • Increase white light/hours in tank to spur algae growth to feed CUC

    Votes: 6 6.1%
  • Feed nori to support CUC

    Votes: 35 35.4%
  • Feed herbivore pellets to support CUC

    Votes: 33 33.3%
  • Allow attrition to balance CUC and algae

    Votes: 43 43.4%
  • Provide macro algae to feed CUC

    Votes: 6 6.1%
  • Introduce CUC predators

    Votes: 1 1.0%
  • Other (please explain)

    Votes: 11 11.1%
Back
Top