pH probe not calibrating

  • Thread starter Thread starter zbrusko
  • Start date Start date
  • Tagged users Tagged users None

zbrusko

Active Member
View Badges
Joined
Mar 14, 2023
Messages
460
Reaction score
245
Location
CT
Rating - 0%
0   0   0
I’m attempting to calibrate my Apex pH probe. I follow the steps. Works fine at 7.0. Near the end of the 10.0 calibration, it says it cannot calibrate, and tells me to make sure AOS is up to date, which it is. Not sure how to proceed. Any thoughts?
 
I’m attempting to calibrate my Apex pH probe. I follow the steps. Works fine at 7.0. Near the end of the 10.0 calibration, it says it cannot calibrate, and tells me to make sure AOS is up to date, which it is. Not sure how to proceed. Any thoughts?
This was always a problem for me as well. sometimes after 4 or 5 tries it would actually work. Super frustrating
 
I’m attempting to calibrate my Apex pH probe. I follow the steps. Works fine at 7.0. Near the end of the 10.0 calibration, it says it cannot calibrate, and tells me to make sure AOS is up to date, which it is. Not sure how to proceed. Any thoughts?
This was always a problem for me as well. sometimes after 4 or 5 tries it would actually work. Super frustrating
So I’ll just keep trying then. Thanks.
 
If you have not gotten this to work yet Would try resetting the calibration back to default, rebooting and then trying the calibration.
 
Classic "probe failing at the upper end" pattern — usually one of three things:

1. Buffer solution age. pH 10 absorbs CO2 from air much faster than 7.0 and goes off in months once opened. If your 10.0 bottle has been open a while, try a fresh single-use sachet — that fixes it more often than people expect.

2. Probe age. Apex pH probes are usually good for 1-2 years before slope drift makes 10.0 calibration unreliable. The probe still reads fine at pH 7 (middle range) but the millivolt response gets compressed at the extremes, so 10.0 fails the slope check.

3. Probe glass cleanliness. Salt creep, biofilm, or even fingerprints on the bulb can throw off the upper end more than the lower. Soak the tip in pH 4 buffer or a commercial probe cleaner for 30 min, rinse with RO, then try again.

Order I'd try: fresh 10.0 buffer first (cheap, fast), then clean the probe, then assume the probe needs replacement.

How old is the probe?
 

TOP 10 Trending Threads

ARE YOU READY TO CONFESS TO CRAZIEST, DUMBEST, FUNNIEST THING YOU’VE EVER DONE IN REEFING?

  • Yeah, I'll confess! (Share your story in the comments!)

    Votes: 41 53.9%
  • Nah, I'll keep mine a secret...(Don't be like that, share with the class!)

    Votes: 35 46.1%
Back
Top
Home
Post thread…
Market
What's new