When trying to calibrate the PH probe. Some of the fields are blanked out. Is this a problem ? Or is it my build ? Low and High.......? No drop down.
PH.jpg
 
When trying to calibrate the PH probe. Some of the fields are blanked out. Is this a problem ? Or is it my build ? Low and High.......? No drop down.
PH.jpg
I think you have to run it in that order ...but let me check , the ui has been updated recently
 
ph calibration has to be run in order. That was a requirement for the Atlas scientific pH board. Once you run for ph 7, then the next field is available.
 
Makes sense, because in my reading, the calibration at 7 gives you the offset from zero, then the calibration at 4 give you the slope at low ph(4 to 7) and then the calibration at 10 give you a slope at high ph(7 to 10)
 