Side note because this might cause confusion otherwise:
Short Version first:
I think the Sera Hydrometer uses a different underlying calibration standard than most devices for the hobby. Don't be confused when you use another device and get different readings. Both devices are working perfectly fine, but the numbers have different interpretations. Each device has an underlying conversion table so your readings can be translated to salinity in ppt. And this conversion table is a bit different for the Sera (and some other hydrometers like JBL, AquaMedic DensiMeter, Tropic Marin hydrometer sold in Europe) than it is for most refractometers or US hydrometers.
Long/Detailed Version:
I think the Sera hydrometer is similar to the AquaMedic DensiMeter, JBL Hydrometer, and Tropic Marin hydrometer sold in Europe.
Most devices (especially in the US) measure the density of the saltwater
relative to the density of freshwater at 20°C (≈ 0.998207g/mL). Pretty much all refractometers for the hobby work this way. Looking through a hydrometer you often see a text "d20/20" which specifies this. Sometimes it's also written on the neck of a hydrometer.
But the listed hydrometers above measure density relative to freshwater at roughly 4°C (≈ 0.999975g/mL) which is pretty dang close to the actual density so they are often called "Densimeters" as well.
This means the readings on the listed hydrometers will differ from refractometers for the hobby or US hydrometers.
For example:
1.025 on a d25/4 calibrated device is roughly
37.2ppt where 25°C being the temperature of your sample to measure and 4°C the reference freshwater temperature the density is measured relative to.
1.025 on a d20/20 calibrated to device is roughly
32.9ppt where 20°C being the sample temperature and also the reference freshwater temperature.
You can double check using this calculator:
https://www.capitalcityaquatics.com/salinity
As you can see the difference is 4.3ppt. Small but at least visible. This can cause confusion as to why two devices show different readings.
I had a small discussion about this topic with Randy Holmes-Farley and he said he wouldn't recommend 25/4 devices for most people. While I personally prefer 25/4 because it's a scientific standard due to measuring the actual density, I have to agree with him. Most people are confused when two of their measuring devices don't show the same readings. This doesn't mean you need to throw the Sera hydrometer away. Just be aware that if you see articles online talking about specific gravity/relative density "1.025" means something different for them, than it does for you.