I have been running the Trident for a couple weeks now. Did the calibration when installed.

Only since a couple days I started dosing and noticed my ALK is going up which is good.
I was just curious about the accuracy and started comparing the measurements with the Hanna ALK checker.
I noticed the Hanna always shows around 0.7 dKH higher as the Trident.

Anybody can give some input what the issue could be and how to fix it?
Thanks
 

TheHarold

5000 Club Member
View Badges
Joined
Oct 3, 2015
Messages
5,148
Reaction score
8,760
Rating - 0%
0   0   0
Did you calibrate the Trident? If that was done properly, it is much more accurate than hanna checkers. The hanna checker is advertised as +-.3 dkh, while the trident is around +-.05 dkh.

The batches of Hanna reagent are not known to be particularly consistent. When I change to a new bottle, the readings can change by easily .4 dkh.

One way to verify that the hanna checker is working properly is to use a calibration standard. For example after you calibrate the Trident next time, use the remaining calibration fluid with your hanna checker (As if the calibration fluid was the tank water to be tested) and see what results the Hanna checker provides. That will show you how high, or low it is reading compared to a precision standard.

Or you can buy their calibration check set, though that will again only verify it is reading within +- .3dkh of the real value.

 
OP
OP
MAXreefer

MAXreefer

Active Member
View Badges
Joined
Jan 9, 2009
Messages
434
Reaction score
135
Location
Las Vegas
Rating - 0%
0   0   0
Did you calibrate the Trident? If that was done properly, it is much more accurate than hanna checkers. The hanna checker is advertised as +-.3 dkh, while the trident is around +-.05 dkh.

The batches of Hanna reagent are not known to be particularly consistent. When I change to a new bottle, the readings can change by easily .4 dkh.

One way to verify that the hanna checker is working properly is to use a calibration standard. For example after you calibrate the Trident next time, use the remaining calibration fluid with your hanna checker (As if the calibration fluid was the tank water to be tested) and see what results the Hanna checker provides. That will show you how high, or low it is reading compared to a precision standard.

Or you can buy their calibration check set, though that will again only verify it is reading within +- .3dkh of the real value.


Thanks for the info. Will follow your suggestions
 

Going off the ledge: Would you be interested in a drop off aquarium?

  • I currently have a drop off style aquarium

    Votes: 2 1.4%
  • I don’t currently have a drop off style aquarium, but I have in the past.

    Votes: 2 1.4%
  • I haven’t had a drop off style aquarium, but I plan to in the future.

    Votes: 22 15.4%
  • I am interested in a drop off style aquarium, but have no plans to add one in the future.

    Votes: 66 46.2%
  • I am not interested in a drop off style aquarium.

    Votes: 48 33.6%
  • Other.

    Votes: 3 2.1%
Back
Top