Hi,

I have the latest Apex model with an EB832 and a base unit, nothing else. I am wondering why there is so much fluctuation in my salinity readings, which is directly correlated to the tank temperature? The salinity readings rise and drop with my tank temperature all day long. I am also fighting an issue with the 832 tripping (with under 7A combined load, separate issue.....) and my temperature got down to 70 degrees last night due to this, and my salinity dropped substantially with the temperature.

I have a hard time understanding and trusting my salinity probe. I'm also measuring 1.25 on my refractometer while getting a 31ppt on my probe readings....

Any help is appreciated!

Mike
 

Terence

Valuable Member
View Badges
Joined
Mar 14, 2010
Messages
1,838
Reaction score
3,482
Location
Gilroy, CA
Rating - 0%
0   0   0
There are a few things going on here.

First, salinity readings are absolutely tied to temperature when derived from conductivity (how seawater is electrically measured). To adjust for this you can enable temperature compensation in the settings for that probe.

As far as accuracy is concerned, a properly located and calibrated conductivity probe will always be more accurate than the average hobby refractometer. Also, many in the hobby use salinity reference refractometers instead of seawater (not the same thing). Usually also those refractometer are not calibrated properly before the measurement. And, for instance, among other things Slight jolts to a refractometer like when it is laid on a hard surface, will throw it off. Even the light used as a backlight can change it.

Finally, your 31ppt is only about 5% off the refractometer value you have. Given many of the variables I would never expect multiple disparate salinity measurement methods like these to accurately match up with much any better than +/-3%. Finding consistency in that 33-35ppt range (as measured) has been the key thing for me.

As far as your load issue, be sure to contact our support team if you cannot work it out.
 
OP
OP
M

michaelr7170

Community Member
View Badges
Joined
Jun 17, 2016
Messages
49
Reaction score
6
Rating - 0%
0   0   0
Thank you for the reply and info provided.

So to clarify, how do I know what my salinity actually is with the probe? Is it only accurate when the temperature of my tank is the same as it was when I calibrated in the solution?

Could you please tell me how to enable temperature compensation?
 

Terence

Valuable Member
View Badges
Joined
Mar 14, 2010
Messages
1,838
Reaction score
3,482
Location
Gilroy, CA
Rating - 0%
0   0   0
Select the gear icon, then the probes icon (looks like a probe going in water), select your Cond probe from the list, click advanced, enter 2.2 in the temp compensation field, click the send to apex button in the upper right corner.
IMG_8188.JPG
 

Terence

Valuable Member
View Badges
Joined
Mar 14, 2010
Messages
1,838
Reaction score
3,482
Location
Gilroy, CA
Rating - 0%
0   0   0
Oh, and follow the guidelines on our site for calibrating the temp probe. There are some nuances such as getting the solution temp near the same temp as the tank water by submersing it to equalize the temp when calibrating.
 
OP
OP
M

michaelr7170

Community Member
View Badges
Joined
Jun 17, 2016
Messages
49
Reaction score
6
Rating - 0%
0   0   0
Terence, I left the solution in the tank water for 20 minutes prior to calibrating when I installed a couple of months ago. I would do it again but don't have any solution unfortunately.

Is it ok to enable the temperature compensation now or do I need to wait until the temperature is where I want it to be? Ie currently it's heating up from the ev832 trip, at 72 degrees, but normally I maintain 77-78
 

jonmos75

Active Member
View Badges
Joined
Sep 23, 2015
Messages
225
Reaction score
100
Rating - 0%
0   0   0
Just make sure if you calibrated your probes without using temp compensation and then you choose to use temp compensation that you will need to re-calibrate the conductivity probe, make sure that you float the solution in the same location as the temp probe and enable Temp compensation and follow the calibration process.
 
OP
OP
M

michaelr7170

Community Member
View Badges
Joined
Jun 17, 2016
Messages
49
Reaction score
6
Rating - 0%
0   0   0
Just make sure if you calibrated your probes without using temp compensation and then you choose to use temp compensation that you will need to re-calibrate the conductivity probe, make sure that you float the solution in the same location as the temp probe and enable Temp compensation and follow the calibration process.

I know when I calibrated, I didn't use temperature compensation.... Unfortunately i don't have any solution left..
 

PlanB

Active Member
View Badges
Joined
Oct 1, 2016
Messages
165
Reaction score
169
Location
Jacksonville - FL
Rating - 0%
0   0   0
I too have problems with my salinity probe. I followed the Apex startup (control freak newbie), and it didn't say anything about acclimating the calibration fluid. Unfortunately, I too don't have any calibration fluid left, so up a creek I guess.. :mad:
 
OP
OP
M

michaelr7170

Community Member
View Badges
Joined
Jun 17, 2016
Messages
49
Reaction score
6
Rating - 0%
0   0   0
So I re-calibrated again tonight, both my refractometer (using two different solutions of 1.026 fluid, both align) and my salinity probe.

After calibration my refractometer is showing a reading of 35 ppt in my aquarium.

My salinity probe is giving a reading of 32.6-33.0, which it was also giving before calibration as well.

SO - I want my salinity at 35ppt, should I trust my refractometer and assume my salinity is spot on?! Or should I assume my refractometer is reading high, and up my salinity until my probe reads 35.
 

Bob Escher

Welcome to Saltwater
View Badges
Joined
Aug 8, 2015
Messages
2,918
Reaction score
1,675
Location
Nashotah Wisconsin
Rating - 0%
0   0   0
I'm trusting my refractometer and not my probe myself. I've read too many stories about probes not calibrating correctly ( not implying they don't at all just saying and mine didn't either both of them
 

Addicted2ACRO's

Valuable Member
View Badges
Joined
Sep 4, 2013
Messages
1,158
Reaction score
316
Location
Miami, Florida
Rating - 0%
0   0   0
Hello, sorry I'm late to the party.
If your Salinity measures 1.024, 25 or 26 on a refractometer...what should the reading of your salinity be on your Apex display?
Mine displays 29.6, would you consider that to be accurate? If not, what should it read?
Thank you
 

revhtree

Owner Administrator
View Badges
Joined
May 8, 2006
Messages
47,909
Reaction score
88,471
Rating - 100%
1   0   0
Just looking for a thread like this! How long does it take for temperature compensation to take effect as far as logging is concerned once you edit it?
 

Leishman

143gadgets' Mentor
View Badges
Joined
May 18, 2007
Messages
2,062
Reaction score
1,626
Location
Frederick, MD
Rating - 0%
0   0   0
Just got my 1st Apex and the salinity calibration has ruined what I thought would be a fun weekend setting it up. Took several attempts to get to the second step, kept getting a fail on the 'acceptable range'. The two times I got to step two (after waiting twelve hours for the probe to dry) it would again fail on the, 'acceptable range', with a big red error box stating, "Calibration failed. Probe was not able to be calibrated".

Now what do I do?
 

jonmos75

Active Member
View Badges
Joined
Sep 23, 2015
Messages
225
Reaction score
100
Rating - 0%
0   0   0
You do not need to wait 12 hours for the probe to dry. When I pull my probe out of the water, and give it a good onetime shake to get off the large water out of the probe and will blow on the probe or use canned air to dry the probe in 30 seconds to a minute.

Instead of using the Automatic calibration choose the Advanced Tab and at the bottom of the page you will see Manual calibration. When looking at the Dry settling numbers you should see anything from 60-80, and when in the calibration solution you can see the settling number range from 200-800.

If the Dry settling numbers are in the 100's then you will need to reset the Salinity probe port back to factory settings, by manually calibrating the probe with NO PROBE connected to the base unit.
 

Leishman

143gadgets' Mentor
View Badges
Joined
May 18, 2007
Messages
2,062
Reaction score
1,626
Location
Frederick, MD
Rating - 0%
0   0   0
You do not need to wait 12 hours for the probe to dry. When I pull my probe out of the water, and give it a good onetime shake to get off the large water out of the probe and will blow on the probe or use canned air to dry the probe in 30 seconds to a minute.

Instead of using the Automatic calibration choose the Advanced Tab and at the bottom of the page you will see Manual calibration. When looking at the Dry settling numbers you should see anything from 60-80, and when in the calibration solution you can see the settling number range from 200-800.

If the Dry settling numbers are in the 100's then you will need to reset the Salinity probe port back to factory settings, by manually calibrating the probe with NO PROBE connected to the base unit.

Will give it a try tonight. I hoping it's user error, not a faulty probe.

Thanks.
 

Axel Juarez

New Member
View Badges
Joined
Dec 5, 2018
Messages
1
Reaction score
0
Rating - 0%
0   0   0
Are ya'll still there i need help So i am doing the calibration manually for the salt and it says 776 for the dry step and it wont go down what do i do ?
 

Looking back to your reefing roots: Did you start with Instant Ocean salt?

  • I started with Instant Ocean salt.

    Votes: 195 71.4%
  • I did not start with Instant Ocean salt, but I have used it at some point.

    Votes: 21 7.7%
  • I did not start with Instant Ocean salt and have not used it.

    Votes: 50 18.3%
  • Other.

    Votes: 7 2.6%

New Posts

Back
Top