Despite API reputation, the Chemicals in the test kits are actually good

brandon429

why did you put a reef in that
View Badges
Joined
Dec 9, 2014
Messages
29,778
Reaction score
23,748
Location
tejas
Rating - 0%
0   0   0
Based on your work, if we update the procedure and method the actual kit is strong, agreed. Api why have you delayed even one microsecond implementing a new instruction kit with a usage and context addendum for the readout

An updated usage packet with context info and confounds to avoid could streamline results much stronger + increases buyer brand quality association. Moving past 'must be zero to be safe' is the number one stricken clause in the new set

Considering the degree of doubt api has created in the ammonia test will the same instruction set since 1982, the undoing is simple, sell this to them as an addendum to or a new instruction set for the kit

they owe you for showing a method to streamline results/ social media big money/ they owe you a chunk. This thread shows its potential when the right order of ops plus reader ruleset is applied
 
Last edited:
OP
OP
taricha

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,565
Reaction score
10,146
Rating - 0%
0   0   0
I’d be much more interested in alkalinity and calcium though. In nearly 18 years and more than a few dozen tanks setup, I’ve never once tested ammonia or nitrite.
Here's a look at API alkalinity.

I did two API kits and my Red sea kit and ran them each as though a 10mL sample (so 2x the API drops of a 5mL sample).
In each case I added (to distilled water) the amount of those alk reagents that would give you a certain dKH and then titrated with 0.100 Molar NaOH until I got the color change endpoint (a green like on the Red Sea color card).

The two API kits were essentially identical and about 10% off from my Red Sea kit.
The black line represents the theoretical correct value of 1 meq/L = 2.8 dKH.
(I don't know why all the kits were that far off from theory, but I'll check my NaOH stock later.)

API _RS alk.png


Anyway, looks like both the API kits give me something quite close to my Red Sea kit.
 
Last edited:

PeterErc

Well-Known Member
View Badges
Joined
Aug 15, 2014
Messages
653
Reaction score
916
Location
S Fl.
Rating - 0%
0   0   0
Thanks for doing this, I have used API for years. A 5ml syringe for samples and all is good.
Keep it Simple. I have one of those checkers for PO4. If ALK and CA are the same process then f that. API all day.
The Nitrate and I believe Ammonia kit can be used with 2.5ml sample water and half the reagents. Maybe some others but I mostly use alkalinity and calcium. Would be nice for a magnesium kit from API, although I don’t test too often.

For me I just have to be close, if alkalinity is low calcium probably is too. An extra scoop of kalkwasser in the top off bucket. If nitrate is creeping up, some vinegar in the top off. If phosphate creeps up some lanthanum in the overflow. If nutrients are too low (never a problem) feed fish more. Every blue moon I bump the magnesium, use NSW so that’s not usually an issue.
 
OP
OP
taricha

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,565
Reaction score
10,146
Rating - 0%
0   0   0
you're right, @PeterErc stability is about tracking the trend - day to day and week to week. All you need is a kit that's decently close and will be very repeatable. API alk can do that, and you're also right that it's easier to catch a change in calcification by looking at Alk trend and not Ca.
 

MnFish1

10K Club member
View Badges
Joined
Dec 28, 2016
Messages
22,842
Reaction score
21,974
Rating - 100%
1   0   0
Here's a look at API alkalinity.

I did two API kits and my Red sea kit and ran them each as though a 10mL sample (so 2x the API drops of a 5mL sample).
In each case I added (to distilled water) the amount of those alk reagents that would give you a certain dKH and then titrated with 0.100 Molar NaOH until I got the color change endpoint (a green like on the Red Sea color card).

The two API kits were essentially identical and about 10% off from my Red Sea kit.
The black line represents the theoretical correct value of 1 meq/L = 2.8 dKH.
(I don't know why all the kits were that far off from theory, but I'll check my NaOH stock later.)

API _RS alk.png


Anyway, looks like both the API kits give me something quite close to my Red Sea kit.
Curious - did you do more than 1 RedSea test?
 

MnFish1

10K Club member
View Badges
Joined
Dec 28, 2016
Messages
22,842
Reaction score
21,974
Rating - 100%
1   0   0
The big problem I think people are having with API, though some readings are sky high (of ammonia) - I have not heard reports of 'low ammonia' when tested. From my readings MOST of the API tests are called out for reading a 0.25 level of total ammonia. IMHO - the reason for this is people don't realize that with these kits (ammonia) - you're not supposed to average visually the results. You're supposed to pick the color closest to the chart - and thats the numbers. i.e. a slightly green ammonia is read as 0. A green closer to .25 is read as .25.

From API: "
  1. Read the test results by comparing the color of
    the solution to the appropriate Ammonia Color Card (choose either Freshwater or Saltwater. For ponds, use the freshwater color card). The tube should be viewed in a well-lit area against the white area of the card. The closest match indicates the ppm (mg/L) of ammonia in the water sample. Rinse the test tube with clean water after use. Note: Do not pour test tube contents back into the aquarium/pond."
 
OP
OP
taricha

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,565
Reaction score
10,146
Rating - 0%
0   0   0
I re-checked this:
The two API kits were essentially identical and about 10% off from my Red Sea kit.
The black line represents the theoretical correct value of 1 meq/L = 2.8 dKH.
(I don't know why all the kits were that far off from theory, but I'll check my NaOH stock later.)

API _RS alk.png


My Hach 1.00N NaOH exactly balances an equal volume of my Hach 1.0N HCl, so the standards are likely fine.
And using the NaOH standard I confirmed that a redo gave me the same result as in the above chart.

Here's the same data but flipped around and done in terms of just dKH to make it a little easier to interpret.
API _red sea Alk test.png

So an actual alkalinity of 8 kH would be measured at about 7 on my Red Sea kit, and 6-6.5 on my two API kits.

Some people may decide that's close enough. Others may not like that performance.
(in Randy's alkalinity standard thread, Red Sea was also measured there to give a result of 7 for an alk standard of 8 dKH.)
 

MnFish1

10K Club member
View Badges
Joined
Dec 28, 2016
Messages
22,842
Reaction score
21,974
Rating - 100%
1   0   0
I re-checked this:



My Hach 1.00N NaOH exactly balances an equal volume of my Hach 1.0N HCl, so the standards are likely fine.
And using the NaOH standard I confirmed that a redo gave me the same result as in the above chart.

Here's the same data but flipped around and done in terms of just dKH to make it a little easier to interpret.
API _red sea Alk test.png

So an actual alkalinity of 8 kH would be measured at about 7 on my Red Sea kit, and 6-6.5 on my two API kits.

Some people may decide that's close enough. Others may not like that performance.
(in Randy's alkalinity standard thread, Red Sea was also measured there to give a result of 7 for an alk standard of 8 dKH.)
I personally would not think it's close enough. - however I would also be doing multiple measures with mean and standard deviations - and then measuring whether the differences were within statistical significance? The interesting thing is that neither test matches well, the stock - but unknown whether its significantly different
 

Dan_P

5000 Club Member
View Badges
Joined
Sep 21, 2018
Messages
6,701
Reaction score
7,184
Rating - 0%
0   0   0
I re-checked this:



My Hach 1.00N NaOH exactly balances an equal volume of my Hach 1.0N HCl, so the standards are likely fine.
And using the NaOH standard I confirmed that a redo gave me the same result as in the above chart.

Here's the same data but flipped around and done in terms of just dKH to make it a little easier to interpret.
API _red sea Alk test.png

So an actual alkalinity of 8 kH would be measured at about 7 on my Red Sea kit, and 6-6.5 on my two API kits.

Some people may decide that's close enough. Others may not like that performance.
(in Randy's alkalinity standard thread, Red Sea was also measured there to give a result of 7 for an alk standard of 8 dKH.)
This work leads to question that haunts (?) the hobby. How close is close enough, and how much variation in the measurement is acceptable? The answer should lead us to keep livestock healthy and at a reasonable cost, i.e., just enough accuracy and precision. Is there a scientifically based answer? My guess at the answer, based on the fact most aquarist do not test replicate samples, use a chemical standard to validate the test method each time, nor know the accuracy and precision of the analytical method, is that quite a bit of leeway exists for the consistency of the analyte level and the true concentration. Thoughts?
 

Seansea

Valuable Member
View Badges
Joined
Dec 20, 2022
Messages
1,127
Reaction score
981
Location
Flat Rock
Rating - 0%
0   0   0
I use api for nitrate. Took a reading on the day i sent my icp test in. I had it at 2. Icp came back 1.9. Is that close enough?
 

calhoun3186

Community Member
View Badges
Joined
May 13, 2023
Messages
36
Reaction score
58
Location
NE TX
Rating - 0%
0   0   0
Very good research and write up. I really appreciate your use of applicable and appropriately applied statistical methods. Impressive.
 
OP
OP
taricha

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,565
Reaction score
10,146
Rating - 0%
0   0   0
This work leads to question that haunts (?) the hobby. How close is close enough, and how much variation in the measurement is acceptable? The answer should lead us to keep livestock healthy and at a reasonable cost, i.e., just enough accuracy and precision. Is there a scientifically based answer? My guess at the answer, based on the fact most aquarist do not test replicate samples, use a chemical standard to validate the test method each time, nor know the accuracy and precision of the analytical method, is that quite a bit of leeway exists for the consistency of the analyte level and the true concentration. Thoughts?
I've thought about this occasionally. When you ask large groups of R2R people about their alkalinity, you get these really cleanly normal distribution curves like this....
Screen Shot 2023-07-15 at 8.40.19 AM.png


And this one's even better...
Screen Shot 2023-07-15 at 8.38.52 AM.png


That's an amazingly smooth bell curve! Why?

My guess is that it means that in the context of hobby reefkeeping, the actual alkalinity doesn't really matter - within the broad agreement that 7 to 9.5 is the target. Also that so many people have a clear answer means that they have a number they like and they aim for it consistently. And their tank likes this consistent target more than experimentation.
I've never heard a reefer say "my tank really loves a swing up and down from 7 dKH to 10."

We could wonder to ourselves if everyone used a kit calibrated against lab standards, and charted their rate of coral growth by mass or length/month maybe the normal distribution wouldn't look like that - maybe it'd be skewed narrowly toward an actual "best" value.
But since that's not the hobby situation, my interpretation of this chart is that it's just a vote for consistency aiming at a widely agreed target giving good results.

(feel free to make up your own stories about these charts :) )
 
OP
OP
taricha

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,565
Reaction score
10,146
Rating - 0%
0   0   0
I use api for nitrate. Took a reading on the day i sent my icp test in. I had it at 2. Icp came back 1.9. Is that close enough?
I'll still use my hanna or Red Sea as go-to NO3 but I'd be happy with those numbers.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
67,480
Reaction score
63,877
Location
Arlington, Massachusetts, United States
Rating - 0%
0   0   0

Seansea

Valuable Member
View Badges
Joined
Dec 20, 2022
Messages
1,127
Reaction score
981
Location
Flat Rock
Rating - 0%
0   0   0
Wont argue hannah is easier and the digital is nice. I use the alk checker which i won on the cherry coral sale. But at 10 bucks i can buy six api nitrate test kits for 1 hannah checker. Thats 6 years worth of api nitrate tests.
 

Dan_P

5000 Club Member
View Badges
Joined
Sep 21, 2018
Messages
6,701
Reaction score
7,184
Rating - 0%
0   0   0
I've thought about this occasionally. When you ask large groups of R2R people about their alkalinity, you get these really cleanly normal distribution curves like this....
Screen Shot 2023-07-15 at 8.40.19 AM.png


And this one's even better...
Screen Shot 2023-07-15 at 8.38.52 AM.png


That's an amazingly smooth bell curve! Why?

My guess is that it means that in the context of hobby reefkeeping, the actual alkalinity doesn't really matter - within the broad agreement that 7 to 9.5 is the target. Also that so many people have a clear answer means that they have a number they like and they aim for it consistently. And their tank likes this consistent target more than experimentation.
I've never heard a reefer say "my tank really loves a swing up and down from 7 dKH to 10."

We could wonder to ourselves if everyone used a kit calibrated against lab standards, and charted their rate of coral growth by mass or length/month maybe the normal distribution wouldn't look like that - maybe it'd be skewed narrowly toward an actual "best" value.
But since that's not the hobby situation, my interpretation of this chart is that it's just a vote for consistency aiming at a widely agreed target giving good results.

(feel free to make up your own stories about these charts :) )
I like the story.
 

Garf

5000 Club Member
View Badges
Joined
Oct 23, 2020
Messages
5,162
Reaction score
5,990
Location
BEEFINGHAM
Rating - 0%
0   0   0
I bought another API phosphate kit to use in the Hanna yesterday. I noticed 3 things.

1) the zero phos calibration on new salt water was a lot higher than the previous test kit. Old test kit read 0.51 on Hanna, new test kit reads 1.10. (There May be a little phos in the salt I suppose), but that’s a big jump.

2) the slope of colour change seems to have changed dramatically. Previously there was a 1:1 relationship between API increase and theoretical increase. For example, simply subtracting the 0.51 starting point from any reading with API gave a good reading throughout the entire range. This appears now not to be the case with the new API kit. it’s actually a much shallower slope, almost 1:2. For example, a theoretical increase of 0.3 is resulting in an API/Hanna increase of 0.16.

3) initially I was getting extremely wobbly results, with numbers wobbling around by 0.14. Strangely this happened with the original test kit to begin with, but settled down after a while. I shook the living daylights out of both API bottles in the new kit (not in the instructions as far as I can see), and it settled down.

Waiting for Hanna reagents to arrive to get my sanity back.
 

Keeping it clean: Have you used a filter roller?

  • I currently use a filter roller.

    Votes: 26 28.0%
  • I don’t currently use a filter roller, but I have in the past.

    Votes: 3 3.2%
  • I have never used a filter roller, but I plan to in the future.

    Votes: 26 28.0%
  • I have never used a filter roller and have no plans to in the future.

    Votes: 34 36.6%
  • Other.

    Votes: 4 4.3%
Back
Top