Hi,
i am trying to setup a new tank (so for now no 'my tank' thread, as it's still empty) , and i am quite a newbie, so first things first i am trying to check salinity :)
now, something strange happen when using digital refractometer, that is labeled with ATC
if i put a drop of water, and press read, it shows 32ppt.
Water T is 26*C, room T is 19*C
now if i leave the glass of water used to pick a sample (again 32ppt) for a while till it cool down from 26 to 19, i get a new sample from the glass, and try again, the salinity raise of 2% (34ppt).
few questions now
1) is ATC working at all? shouldn't the ppt be the same?
2) what is the 'best' way to do use it? leave the sample reach room temperature and then trust this value?
3) anything i am missing
for the record:
- i used bi-distilled water to calibrate, at room temperature, and subsequent read report 0% correctly
- the refractometer is a green milwaukee
thanks for helping (and sorry for my terrible english)
i am trying to setup a new tank (so for now no 'my tank' thread, as it's still empty) , and i am quite a newbie, so first things first i am trying to check salinity :)
now, something strange happen when using digital refractometer, that is labeled with ATC
if i put a drop of water, and press read, it shows 32ppt.
Water T is 26*C, room T is 19*C
now if i leave the glass of water used to pick a sample (again 32ppt) for a while till it cool down from 26 to 19, i get a new sample from the glass, and try again, the salinity raise of 2% (34ppt).
few questions now
1) is ATC working at all? shouldn't the ppt be the same?
2) what is the 'best' way to do use it? leave the sample reach room temperature and then trust this value?
3) anything i am missing
for the record:
- i used bi-distilled water to calibrate, at room temperature, and subsequent read report 0% correctly
- the refractometer is a green milwaukee
thanks for helping (and sorry for my terrible english)