I wanted to see if my HI782 was accurate, so I conducted a small experiment, and the results were quite interesting.
I used 9.5 mL of the ASTM D1141 standard solution and added 0.5 mL of a 100 mg/L NO3-N standard solution; the theoretical NO3 concentration of the resulting mixture should be 22.13 ppm.However, the HI782 yielded a reading of only 16.3. Is this reading normal? It looks like it far exceeds the margin of error allowed for the HI782.
I used 9.5 mL of the ASTM D1141 standard solution and added 0.5 mL of a 100 mg/L NO3-N standard solution; the theoretical NO3 concentration of the resulting mixture should be 22.13 ppm.However, the HI782 yielded a reading of only 16.3. Is this reading normal? It looks like it far exceeds the margin of error allowed for the HI782.