KH Director NOT taking Measurement

Users Who Are Viewing This Thread (Total: 1, Members: 0, Guests: 1)

buck27

Well-Known Member
View Badges
Joined
Jul 9, 2011
Messages
632
Reaction score
156
Location
Huntsville AL
Rating - 0%
0   0   0
Just got this thing all set up and calibrated, put the KH Director on the dashboard with my other stuff and then told it to test, three times, shows no results but gives me an alarm and 2 warnings.

Alarm: Measured value has fallen below or exceeded the alarm threshold
Warning: check filling level pump 3
Warning: check filling level pump 4

I have them set up just like the GHL Video, pump 1 is the sample water, pump 2 is the reagent, pump 3 is the waste water, pump 4 is for my Alk dosing.

I am lost as to why I am not getting a reading. it says it fell below the threshold but I all I see is zero.
 

esther

Valuable Member
View Badges
Joined
Oct 12, 2019
Messages
1,849
Reaction score
3,213
Location
Long Beach, CA
Rating - 0%
0   0   0
Just got this thing all set up and calibrated, put the KH Director on the dashboard with my other stuff and then told it to test, three times, shows no results but gives me an alarm and 2 warnings.

Alarm: Measured value has fallen below or exceeded the alarm threshold
Warning: check filling level pump 3
Warning: check filling level pump 4

I have them set up just like the GHL Video, pump 1 is the sample water, pump 2 is the reagent, pump 3 is the waste water, pump 4 is for my Alk dosing.

I am lost as to why I am not getting a reading. it says it fell below the threshold but I all I see is zero.
So, you have nothing where it says current value?
1F88292B-3C40-470B-8B42-1CFA04F857D5.png
 
OP
OP
buck27

buck27

Well-Known Member
View Badges
Joined
Jul 9, 2011
Messages
632
Reaction score
156
Location
Huntsville AL
Rating - 0%
0   0   0
If anyone is running one and got it going good, shoot me a PM and I will give you my phone number and we can text or talk me through it.
 

esther

Valuable Member
View Badges
Joined
Oct 12, 2019
Messages
1,849
Reaction score
3,213
Location
Long Beach, CA
Rating - 0%
0   0   0
Hmm. Maybe try removing all the alarms and flush the measurement cell and try again? Ive never had that issue. Maybe @Ditto can help.
 

AZMSGT

2500 Club Member
View Badges
Joined
May 6, 2019
Messages
3,988
Reaction score
7,895
Rating - 0%
0   0   0
You need to go back and review the videos and have the manual open and read everything. My guess is your sample line measurement is off and something else might not be right, perhaps probe calibration.

Manual
 

Ditto

Valuable Member
View Badges
Joined
Apr 25, 2017
Messages
2,229
Reaction score
2,972
Location
Albany, NY
Rating - 0%
0   0   0
When you ask it to do the test manually does the unit look like it doing the tes?

is it drawing water?
Are the vent tube and drain tube nit in any water?

Can you screen shot the diagnostics?
 

Attachments

  • 1EB73C69-F595-44F2-ACA4-5A1A512C11C4.png
    1EB73C69-F595-44F2-ACA4-5A1A512C11C4.png
    244 KB · Views: 70

Ditto

Valuable Member
View Badges
Joined
Apr 25, 2017
Messages
2,229
Reaction score
2,972
Location
Albany, NY
Rating - 0%
0   0   0
I still see red alarms in that screen shot is it your two dosing pumps ?

If it is pump 3 turn off level monitoring unless you want to monitor the amount of water and placing it into a container and want Ronnie when it is filled.
0566B12E-CC9E-4055-95ED-66A65F54F9F3.png

pump 4 if your also doing level monitoring your going to want to fill a virtual container
B7AF156B-284D-43E3-B271-EC55B0550A4F.png
 
OP
OP
buck27

buck27

Well-Known Member
View Badges
Joined
Jul 9, 2011
Messages
632
Reaction score
156
Location
Huntsville AL
Rating - 0%
0   0   0
I'm going to try and re calibrate the problem right now.

They're is definitely water going in, going out, etc. I can see it doing whatever it's supposed to be doing with the exception of giving a value on the screen.

Give be about 10 to 15 minutes. Be back shortly.
 
OP
OP
buck27

buck27

Well-Known Member
View Badges
Joined
Jul 9, 2011
Messages
632
Reaction score
156
Location
Huntsville AL
Rating - 0%
0   0   0
Alright! Got it going now. When I went back to double check all the lines and the PH Probe I found the probe connector not seated all the way *facepalm* I swear I had it on there! So that would mean to me the calibration was not correct because of the connection. Recalibrated the probe and as I suspected my ALK is low. I do top off with Kalk so now I need to get some soda ash mixed up and get this baby up to 8 at least. Screenshot_20201213-090621_GHL Connect.jpg Screenshot_20201213-090629_GHL Connect.jpg
 

paul01609

Active Member
View Badges
Joined
May 20, 2020
Messages
476
Reaction score
439
Location
U.k
Rating - 0%
0   0   0
after calibration I let mine settle for a day befor dosing
 