My KH director, i dont know what happened on it. Everything was ok, after a day, it was wrong, my KH is 7.7 and now it is 0.4 0.5 and 0.6 each check.


Does anyone have the same problem before? I tried to reset my device and do a test with ph 4 and 7, but when i put probe to ph4, the number is still 0.


So is there any problem with it? But it is strange, when i put back the probe, i see in KHDirector , ph still change, 2.8 ~ 3.5.
 

wmb0003

Active Member
View Badges
Joined
Sep 3, 2020
Messages
347
Reaction score
458
Location
Auburn AL
Rating - 100%
1   0   0
Sounds to me like your pH probe is no good. How old is the unit or pH probe?
 

Lasse

10K Club member
View Badges
Joined
Mar 20, 2016
Messages
12,736
Reaction score
31,462
Location
Källarliden 14 D Bohus, Sweden
Rating - 0%
0   0   0
I´ll think you need to recalibrate the electrode - start with that and see whats happens. Check your callibration data - they should somewhere like these (blue markings)

1667405582833.png


Sincerely Lasse
 
OP
OP
N

norival1992

New Member
View Badges
Joined
Nov 1, 2022
Messages
3
Reaction score
0
Location
Vietnam
Rating - 0%
0   0   0
Hi my PH value atm, looks like it is not good true. I heared some one, they said good value is 4.5 right ?

1667405890675.png
 

Lasse

10K Club member
View Badges
Joined
Mar 20, 2016
Messages
12,736
Reaction score
31,462
Location
Källarliden 14 D Bohus, Sweden
Rating - 0%
0   0   0
It should stop titrating at 4.3 (pretty close at least). Star the troubling shooting with a new calibration of both pumps and electrode

Sincerely Lasse
 

wmb0003

Active Member
View Badges
Joined
Sep 3, 2020
Messages
347
Reaction score
458
Location
Auburn AL
Rating - 100%
1   0   0
but when i put probe to ph4, the number is still 0.
This statement tells me something is not right with the pH probe, as Lasse said I would recalibrate the probe. If you can not get it to read pH of 4 and 7 when in the solutions, something is wrong with the probe and that should be investigated before you look at the calibration of the dosing heads.
 