I'm not trying to chase per say but I do have a question.

My Apex controller pH is registering my pH at a really low 7.35 (did a water change a few days ago about 10% volume) though when I test with my api test kit im getting 8.4...what gives?

I did calibrate the probe when I got it and it's only maybe 3/4 months old...do they really lose calibration that quick or do I have another issue?
 
I'm not trying to chase per say but I do have a question.

My Apex controller pH is registering my pH at a really low 7.35 (did a water change a few days ago about 10% volume) though when I test with my api test kit im getting 8.4...what gives?

I did calibrate the probe when I got it and it's only maybe 3/4 months old...do they really lose calibration that quick or do I have another issue?

I would verify that calibration but checking it against the calibration solution in regular mode. Did you notice if the pH probe drifted over time, of just suddenly started reading low? If it drifted, it could be the probe and/or calibration. If it was a sudden change, you likely have electrical interference. Check the probe wire routing to make sure its not tangled or coiled with other wires or parallel to electrical wires. Could also be stray voltage, which you can verify with a volt meter and/or eliminate with a ground probe. Ultimately, you will want to identify that source and eliminate it, aka replace the faulty equipment.
 
I would verify that calibration but checking it against the calibration solution in regular mode. Did you notice if the pH probe drifted over time, of just suddenly started reading low? If it drifted, it could be the probe and/or calibration. If it was a sudden change, you likely have electrical interference. Check the probe wire routing to make sure its not tangled or coiled with other wires or parallel to electrical wires. Could also be stray voltage, which you can verify with a volt meter and/or eliminate with a ground probe. Ultimately, you will want to identify that source and eliminate it, aka replace the faulty equipment.
I'm gonna get some calibration fluid in the next few days as well as have another test kit test my water (more points, should help narrow it down a little), I have my pH and temp probe in the Apex holder in my sump in the return pump area. And for the most part my pH has been steady between 7.8 and 8.2 over the last few weeks.
 
So, looks like my PH is finally heading back into normal ranges, moved some cabling around (thank you velcro ties) to make sure i dont get any interference from a/c
 

TOP 10 Trending Threads

HOW WOULD YOU FEEL ABOUT RECEIVING A FISH OR CORAL AS A HOLIDAY GIFT?

  • I’d love one, as long as it’s a species I chose

    Votes: 31 24.0%
  • I’d trust someone who knows my tank to choose

    Votes: 11 8.5%
  • I’d rather get a gift card and pick it myself

    Votes: 47 36.4%
  • I’d prefer gear or supplies instead

    Votes: 6 4.7%
  • Please don’t surprise me with livestock

    Votes: 34 26.4%
  • Other (Post in the comments)

    Votes: 0 0.0%
Back
Top
Home
Post thread…
Market
What's new