Trident Inconsistent Reading At End Of Reagent Pack?

  • Thread starter Thread starter ShaggyRS6
  • Start date Start date
  • Tagged users Tagged users None

ShaggyRS6

Valuable Member
View Badges
Joined
Nov 6, 2017
Messages
1,101
Reaction score
1,272
Location
MT Juliet, Tennessee
Rating - 0%
0   0   0
Has anyone noticed that when you get toward the end of a Reagent Pack that your Trident starts reporting DKH in correctly? I have noticed this on my last two refills. When I get to around 30 tests of B and C the DKH starts to report low.
 
That happened on one of my Tridents briefly, but I adjusted the pickup tube to make sure it was drawing from the bottom of the container, and now its perfect.
 
I have heard others report this but I have not experienced it myself. I get consistent readings on all tests up until Apex tells me to replace them.
 
When the reagents get low the pickup tube may not be getting the correct amount causing it to continue “adding” reagent to the sample which will make the Trident think it used more reagent than it actually did and give an abnormally high result. My B and C normally need replacing when they report 30% left and my A usually around 20% left.
 
When the reagents get low the pickup tube may not be getting the correct amount causing it to continue “adding” reagent to the sample which will make the Trident think it used more reagent than it actually did and give an abnormally high result. My B and C normally need replacing when they report 30% left and my A usually around 20% left.
I could totally see that happening if the Trident was sitting on an unlevel surface. If it was pitched enough in a certain way.
 

TOP 10 Trending Threads

ARE YOU READY TO CONFESS TO CRAZIEST, DUMBEST, FUNNIEST THING YOU’VE EVER DONE IN REEFING?

  • Yeah, I'll confess! (Share your story in the comments!)

    Votes: 68 56.7%
  • Nah, I'll keep mine a secret...(Don't be like that, share with the class!)

    Votes: 52 43.3%
Back
Top
Home
Post thread…
Market
What's new