Hopefully this helps someone. I was calibrating my probe the other night and through the automatic process, was able to calibrate to the 7 pouch just fine, however the 10 never went green acceptable range.
My temp workaround was manual calibration but that resulted in a lower than expected reading, falling outside the range triggering every built-in apex alarm.
So, I can't remember if I read this somewhere and if so...apologies for a double post, or whether I had some crazy dream, but after trying to recalibrate locally (http://apex.local) vs through fusion...it worked just fine almost instantly. Obviously only works if you're on your network.
Apologies again if this is widely known...I'll pull down.
My temp workaround was manual calibration but that resulted in a lower than expected reading, falling outside the range triggering every built-in apex alarm.
So, I can't remember if I read this somewhere and if so...apologies for a double post, or whether I had some crazy dream, but after trying to recalibrate locally (http://apex.local) vs through fusion...it worked just fine almost instantly. Obviously only works if you're on your network.