Can pH buffer solution used for calibrating pH Probes be stored and reused without negatively affecting accuracy?

Indoor Reef

New Member
View Badges
Joined
Aug 22, 2021
Messages
20
Reaction score
15
Location
Uk
What state or country do you live in
United Kingdom
Rating - 0%
0   0   0
After mixing pH buffer solution from the little sachets (6.86, 9.81) what would be the impact on accuracy if these buffer solutions were stored in sealed glass jars and re-used used for calibration at a later date?
 
The pH 9.81 will drift down as it absorbs CO2. I've used buffer solutions for years, just keeping them in the plastic containers they came in, but i did remeasure the pH of them occasionally. The pH 10 solutions drift into the 9.7-9.9 range.

If your meter allows you to input the pH of the buffers, that works fine. if it assumes a certain value, that's not fine.
 
The pH 9.81 will drift down as it absorbs CO2. I've used buffer solutions for years, just keeping them in the plastic containers they came in, but i did remeasure the pH of them occasionally. The pH 10 solutions drift into the 9.7-9.9 range.

If your meter allows you to input the pH of the buffers, that works fine. if it assumes a certain value, that's not fine.
sorry, i think i misunderstood question.

i never have, but thought question was ,use solution, then store and reuse.
figured it would dilute solution.
correct?? Thanks
 
sorry, i think i misunderstood question.

i never have, but thought question was ,use solution, then store and reuse.
figured it would dilute solution.
correct?? Thanks

If the probe has just a bit of RO/DI left on it from rinsing before you put it in (which is appropriate), then the dilution is minor and in general, the pH of buffers is resistant to the impact of dilution.

The pH of a buffer solution is determined by the ratio of concentration of the two ions creating the buffer (HA and A-, as an example). The governing equation is shown below.

pH= pKa+ log( [A−]/ [HA] )


where "pKa" is the pKa of the acid HA.

Thus, since pH is determined by the ratio of HA and A-, diluting them both equally has no substantial impact on pH.
 
After mixing pH buffer solution from the little sachets (6.86, 9.81) what would be the impact on accuracy if these buffer solutions were stored in sealed glass jars and re-used used for calibration at a later date?
I do this now and BBC am able to successfully calibrate. I have now gone to the packet solutions
 
Thanks everyone! I don't have any plastic containers so I'll make up the buffers and store them in glass jam jars. When I come to the next calibration with fresh buffer sachets then I'll test the old solution and see how much its changed. This will maybe help me understand the resuability over time.
 

TOP 10 Trending Threads

WHAT DO YOU ALWAYS RUN OUT OF AT THE WORST TIME?

  • Salt mix

    Votes: 17 17.7%
  • Fish or coral food

    Votes: 5 5.2%
  • Test kits or reagents

    Votes: 10 10.4%
  • Filter media or replacement fleece

    Votes: 5 5.2%
  • RO/DI filters or supplies

    Votes: 15 15.6%
  • Fragging and repair supplies, like glue or epoxy

    Votes: 12 12.5%
  • I keep a backup of everything

    Votes: 25 26.0%
  • Other (Post in the comments)

    Votes: 7 7.3%
Back
Top
Home
Post thread…
Market
What's new