I don’t ever remember having a result as highest .9, but the color gets darker the higher the reading, so if you’re getting a high reading and a clear vial that would indicate something blocking the sensor or a bad checker.Would you take a look at Shirak’s comment? He made a comment about the .9 would have been a dark blue. It was in fact not blue when I pulled it out. Would that be a problem with the checker then?I clean every vial with rubbing alcohol and a microfiber cloth, if you’re dunking the whole vial in the tank, be careful about water sitting underneath the cap, the Hannah readers are not water resistant.
Also, I found the biggest discrepancy to be little micro bubbles inside the vile, after my two minutes of shaking I slowly rock the bottle back-and-forth to clear up any bubbles inside.
If the checker is used and new reagents, don’t fix the problem, it might make sense to try to take it apart and clean it.
You can also have issues with defects in the glass vial, so maybe something to take a close look at too.