Neptune pH Calibration Help

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
Hello,

I am trying to calibrate my apex pH probe. After calibrating in pH 10 solution I keep the probe in the pH 10 and apex reads between 9.0 - 9.7 after calibration.

I calibrate pH meters as part of my job, why is it saying the pH 10 is in acceptable range if the pH10 solution is actually 9.0?

I suspect the buffer is bad, but if the buffer is bad the apex should fail calibration no?

Picture is of after successful calibration with probe still sitting in 10 solution, it should read 10.0 because that is what I told the machine it is.

help?

51ED0985-D2E9-4DE2-AD62-052DCC95F275.png
 

radfly

Active Member
View Badges
Joined
Apr 29, 2022
Messages
209
Reaction score
147
Location
Charleston
Rating - 0%
0   0   0
Your solutions are 4 and 10? I would suspect a faulty probe. Do Manual calibration, do not use calibration Task...
 

Attachments

  • BA818F62-282C-4524-AD78-1E951ADF8F0F.png
    BA818F62-282C-4524-AD78-1E951ADF8F0F.png
    78.6 KB · Views: 40
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
Your solutions are 4 and 10? I would suspect a faulty probe. Do Manual calibration, do not use calibration Task...


Buffers 7 and 10, using a brand new
probe.

what do the values mean when manual calibrating? Does for example , 765 mean 7.65 pH?
 

radfly

Active Member
View Badges
Joined
Apr 29, 2022
Messages
209
Reaction score
147
Location
Charleston
Rating - 0%
0   0   0
Buffers 7 and 10, using a brand new
probe.

what do the values mean when manual calibrating? Does for example , 765 mean 7.65 pH?
I think so. Just wait until it reads 100% on the meter, and the number stays stable for a few seconds.

Not sure why your graph indicates 5.77 when you used a 7 reference?
 
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
I think so. Just wait until it reads 100% on the meter, and the number stays stable for a few seconds.

Not sure why your graph indicates 5.77 when you used a 7 reference?

Yes I did that, and it’s not close to the buffer solution, but it should calibrate to the buffer.

After calibration at 7 it thought the 7 was 5.77
 
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
The 7.0 buffer “value” is 577 and 10.0 is 907. Which I’m seeing now relates to the pH, however after calibration it should make the pH 7.0 and 10.0 but it’s staying at 907 as the probe in pH 10 is 9.07 after calibration
 

Attachments

  • F412F8A5-FD7E-4FE9-BB02-525E9B3C80A2.png
    F412F8A5-FD7E-4FE9-BB02-525E9B3C80A2.png
    77 KB · Views: 29
  • E03A5774-3A04-4844-B5E2-3BB10B787BF6.png
    E03A5774-3A04-4844-B5E2-3BB10B787BF6.png
    120.3 KB · Views: 27
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
Ph 7 buffer
 

Attachments

  • A7863FB3-4C44-4671-9020-95BDD0D4BFD0.png
    A7863FB3-4C44-4671-9020-95BDD0D4BFD0.png
    118.3 KB · Views: 32

spsick

Valuable Member
View Badges
Joined
Feb 17, 2018
Messages
1,426
Reaction score
1,669
Location
Mpls, MN
Rating - 0%
0   0   0
It’s part of the calibration process. I’ve never had it show 7 in the 7 or 10 in the 10 but probe works properly after calibration.
 
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
It’s part of the calibration process. I’ve never had it show 7 in the 7 or 10 in the 10 but probe works properly after calibration.
Yes correct, before calibration it should not read 7 or 10 and that is why you calibrate.

After calibration though your probe should read 10 in the 10, if it doesn’t it didn’t calibrate
 
Last edited:

radfly

Active Member
View Badges
Joined
Apr 29, 2022
Messages
209
Reaction score
147
Location
Charleston
Rating - 0%
0   0   0
During Manual calibration you are to select your reference pair, did you select the 7 10 pair, or the 4 10 pair?
 

Attachments

  • 811B9F4F-7FDB-453D-AB1F-14EDDBEC5633.png
    811B9F4F-7FDB-453D-AB1F-14EDDBEC5633.png
    103.3 KB · Views: 40
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
During Manual calibration you are to select your reference pair, did you select the 7 10 pair, or the 4 10 pair?

7 and 10, there is no 4 and 10


Has your buffer expired?

No, but I agree this could be it, however what is happening does not explain bad buffer. Even with bad buffer I should
Be able to calibrate but have a “bad” calibration
 

radfly

Active Member
View Badges
Joined
Apr 29, 2022
Messages
209
Reaction score
147
Location
Charleston
Rating - 0%
0   0   0
If there is no 4 10 option, your Apex software may need an update. Here is my current version...
 

Attachments

  • 0618A684-BD65-4EF5-AE4B-20EE0B8E51DA.png
    0618A684-BD65-4EF5-AE4B-20EE0B8E51DA.png
    121 KB · Views: 67
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
If there is no 4 10 option, your Apex software may need an update. Here is my current version...

I think you mean 4 and 7

there is 4 and 7……and 7 and 10 options

I’m on the latest version
 
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
These steps fixed the issue. Reading 696 in ph7 and 996 in pH10.

For the ones saying that it should not read the buffer after calibration I recommend these steps because your ph is off. After calibrating it should read exactly or very close to 10.0 in pH10, that is how calibrating works :)
 

Attachments

  • 6AFC0399-AEEF-4F67-81E6-060C340A3CC4.png
    6AFC0399-AEEF-4F67-81E6-060C340A3CC4.png
    240.9 KB · Views: 75
  • 84FC7308-75BA-4429-B6DF-AD0AAC96532F.png
    84FC7308-75BA-4429-B6DF-AD0AAC96532F.png
    139.8 KB · Views: 67

spsick

Valuable Member
View Badges
Joined
Feb 17, 2018
Messages
1,426
Reaction score
1,669
Location
Mpls, MN
Rating - 0%
0   0   0
These steps fixed the issue. Reading 696 in ph7 and 996 in pH10.

For the ones saying that it should not read the buffer after calibration I recommend these steps because your ph is off. After calibrating it should read exactly or very close to 10.0 in pH10, that is how calibrating works :)

fascinating!
 

Alexander1312

Active Member
View Badges
Joined
Oct 10, 2022
Messages
152
Reaction score
110
Location
San Francisco Bay Area, California, United States
Rating - 0%
0   0   0
These steps fixed the issue. Reading 696 in ph7 and 996 in pH10.

For the ones saying that it should not read the buffer after calibration I recommend these steps because your ph is off. After calibrating it should read exactly or very close to 10.0 in pH10, that is how calibrating works :)
How do I get access to APEX local? Seems like this is a great fix but hard to follow all steps. A bit more details would be great. Thanks!
 
OP
OP
piranhaman00

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
How do I get access to APEX local? Seems like this is a great fix but hard to follow all steps. A bit more details would be great. Thanks!

I used the iOS app for all steps.
 