PART 2-----COLORIMETRIC VISUAL TESTING METHODS (The Chemistry)

As we covered in Part 1 in the “Understanding Errors” section, the visual testing method consists of essentially two steps, Evaluation and Assessment. Evaluation is the test procedure (The Chemistry) and Assessment is the determination of the results. We will take a close look at each of these steps and the components of each and look at some potential areas where errors might occur and how we might minimize them.

TESTING PROCEDURES (EVALUATION)

Most test kits consist of a series of reagents* that are mixed together according to a prescribed procedure. The procedure generally requires that the reagents be measured out by “drops”, ml or some type of measuring spoon. They are added to the sample being tested according to a series of steps in a prescribed manor. There is usually a time element associated with the procedure…Mix for 1 min…wait for 30 seconds…etc. This is the case in both a titration test (add drops till color changes) or just a reaction where the reagents are mixed and a resulting color is obtained.

  • Note on Safety: Some of the reagents we use can be hazardous…Nitric Acid…Sulfuric Acid just as examples. Exercise caution when using these reagents so as not to get then in your eyes or on your skin. Eye protection is a good practice. Some kit manufactures provide Safety Data Sheets, include precautions in their instruction sheet or label the containers accordingly. If you are unable to find this information assume it is hazardous! Keep them out of the hands of children!
ASSESSMENT

Once the procedure is complete the result is compared visually to a “Standard”. The standard is usually some type of color card or a colored liquid. As many of us can attest to, evaluating results of a measurement using visual color change or subtle color difference can sometimes be a challenge.
SOURCES OF ERROR IN THE TESTING PROCEDURE (EXAMINATION)

The first area we will explore where errors occur is in the measurement of the volumes of liquids and solids reagents required to do the chemistry.
  • Inaccurate reagent measurement
    • Inaccurate measuring devices
      • Droppers
      • Syringes/Pipettes
      • Measuring Spoons
      • Other
  • Procedural error in reagent measurement
    • Not holding droppers vertical for dispensing
    • Air bubbles in drops when dispensing
    • Measuring Spoon over/under fill
    • Not properly reading the measurement meniscus (Parallax Error)
EXPERIMENTAL EXAMPLES INACCURATE REAGENT MEASUREMENT

EXAMPLES OF INACCURATE REAGENT MEASUREMENT
Here are some examples of some of measurement sources of error (Tables 1-6) as well as some Best Practices to help reduce the errors.
It is important to note that some tests are more critical as to the amount of reagent added and others less so. This fact is not always made clear by the test kit instructions. So I tend to land on the cautious side and treat them all as if the amounts are important to accurate and repeatable results.

DROPPER BOTTLE ERRORS
Many of our test kits require the use of dropper bottles and or droppers to deliver the required volume of reagent for the test. A general rule would be 1 drop = .05mL. Thus 5 drops should yield .25mL. This seems to hold true for all of the test kits I have evaluated. Some areas where errors can creep in are:

1) Incorrect position of the bottle or dropper during delivery
2) Air bubbles in the drops
3) Inconsistent volume delivery of different bottles from new test kit for the same test (Replacement Kits)
Here are some experiments to illustrate these.
1575556548504.png
1575556564556.png
AIR IN DROP PROPER POSITION FOR DROP ADDITION
Fig 4
DROPPER DELIVERY VOLUME TEST (Same Dropper Bottle multiple times)
TEST #
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
ml DELIVERED
5 DROPS*
.22
.24
.21
.19
.23
12.8 %
8.8 %
LESS CARE TAKEN IN MEASUREMENT
TEST #
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
ml DELIVERED
5 DROPS*
.24
.23
.25
.23
.23
5.6%
3.8%
MORE CARE TAKEN IN MEASUREMENT
* Measured with volume weight of distilled water (Known Value .25mL)
TABLE 1
LESS CARE—Dropper was not held vertically and no observation for air bubbles was made.

MORE CARE—Dropper was held vertically and care was taken that there was no air in the drops.

1575557084723.png


DROPPER DELIVERY VOLUME TEST (5 Different Dropper Bottles Same Test Kit)

DROPPER #
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
ml DELIVERED
5 DROPS*
.28
.29
.34
.23
.31
16%
14.0%
* Measured with volume weight of distilled water (Known Value .25mL)

TABLE 2
One way of improving the accuracy and precision is to switch to delivering the required volume (.25mL) by properly using a good syringe or single channel Micro Pipette (Fig 7). These will eliminate the difference between dropper bottles and the dropper errors. See table 3 below.

SYRINGE & PIPETTE DELIVERY VOLUME TEST

SYRINGE TEST #
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
ml DELIVERED .25 ml VOLUME*
.24
.24
.25
.25
.24
2.4%
2.2%
PIPETTE TEST #
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
ml DELIVERED .25 ml VOLUME*
.25
.24
.25
.25
.25
.8%
1.8%

TABLE 3
As you can see from comparing the results in Table 3 to those in Table 1 & 2, there is a marked improvement in the % Relative Error & % Relative Standard Deviation (STDEV) by using a syringe vs. drops. The Single Channel Micro Pipette does even better.

SYRINGES ERRORS

Many test kits require the use of syringes for measuring reagent or for titrations. When correctly used they can provide good accuracy and precision. Here are some things to watch for than can introduce errors.

1) Incorrectly reading the syringe
2) Air bubbles
3) Poor quality syringes
4) Assuming “0” is at the bottom stop of the syringe

1575557575976.png
1575557597416.png

READING THE SYRINGE CORRECTLY
FIG 5

1575557829997.png
1575557842023.png
1575557854945.png


AIR BUBBLES IN THE SYRINGE COLUMN
FIG 6
To eliminate errors in the use of a syringe be familiar with the proper reading location for the particular syringe being used. The instructions provided with most kits will give you this information.

To eliminate the bubbles in the syringe column draw the reagent out of the container with the syringe with the tip fully submerged. Then expel it back into the container fully depressing the syringe. Do this several times to be sure to push out all of the air. On the last draw pull the plunger past the top mark so column is completely full. Hold the syringe tip up and tap the column to release any remaining air. Then slowly depress the plunger until a small amount of liquid comes out of the top. Then eject the extra liquid back into the container until the plunger is at the required mark. Do a visual inspection to be sure there are no remaining air bubbles.

An additional consideration would be to invest in single channel micro pipettes. (See Fig 7)Their cost is around $50 from Amazon for a good one. They are easy to use and quite accurate and precise. Over time I have purchased several different sizes to deliver a range of volumes and now do all liquid measurements with them, including measuring out the volume of tank water to be tested. An additional advantage is they speed up the testing. They; like any instrument, need to be calibrated to keep the accuracy and precision high. I calibrate using mass--- .25mL=.25g of RODI water. I use a Jewelry scale that cost about $20 (Amazon). This is the recommended procedure by the manufacturer.

1575558023511.png

FIG 7
Here is where I purchase the ones I use https://www.amazon.com/s?k=single+c...x=single+channel+,aps,292&ref=nb_sb_ss_i_8_15

Below in Table 4 & 5 are a set of experiments to demonstrate the above ideas

TEST #
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
ml DELIVERED 1 ml VOLUME*
.93
.9
.95
.89
.97
9.2%
3.6%
LESS CARE TAKEN
TEST #
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
ml DELIVERED 1 ml VOLUME*
.97
.99
1
.96
.96
2.4%
1.9%
MORE CARE TAKEN
* Measured with volume weight of distilled water (Known Value 1mL)
TABLE 4

PIPETTE* VOLUME TEST –Fig 7 (Same pipette multiple times)

TEST #
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
ml DELIVERED 1 ml VOLUME*
.993
.991
1.023
.995
.993
.1%
1.35%
* Measured with volume weight of distilled water
TABLE 5

* SCIENTIFIC SINGLE CHANNEL MICRO PIPETTE VARIABLE VOLUME 100-1000µL

As you can see from Table 4 there is quite an improvement in accuracy and precision by applying some best practices in their use. Table 5 shows the gain that can be made using Micro Pipettes.

MEASURING SPOONS ERRORS

1575558449937.png
1575558462000.png

FIG 8

Some tests require us to add a prescribed number of measuring spoon of a dry reagent to the reaction vessel. If you read the instructions it will often say “Add One Leveled measuring spoon to the vial” The word that is often missed or ignored is “Leveled”. Many tests provide pictorial instructions that show the steps of the chemistry to be preformed. If a measuring spoon is required they just show the reagent being added with the spoon…They don’t show the “Leveling” part. You have to read the instructions to find that part. Not leveling the spoon will result in the addition of too much or too little reagent and introduce variability in the process resulting in reduced accuracy and precision. Leveling the spoon increases accuracy and precision which is demonstrated by the results in Table 6

MEASURED AMOUNT WITH SAME MEASURING SPOON MULTIPLE TIMES
TEST #
NO LEVELING
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
MEASURE SPOON 1
.103
.094
.1
.134
.127
11.6%
15.9%
MEASURING SPOON 2
.196
.172
.166
.178
.181
10.7%
6.3%

MEASURED AMOUNT WITH SAME MEASURING SPOON MULTIPLE TIMES
TEST #
LEVELING
1
2
3
4
5
% RELATIVE ERROR
% RELATIVE STDEV
MEASURE SPOON 1
.108
.097
.11
.098
.097
2%
6.3%
MEASURING SPOON 2
.194
.196
.206
.189
.187
2.8%
3.8%
TABLE 6
Statistical Target Value for Measuring Spoon # 1 is .1g

Statistical Target Value for Measuring Spoon # 2 is .2g
Parallax Errors

Parallax error is the shift in apparent position of an object due to different viewing position. It is an error caused by humans, while measuring a quantity if your eye is not at the proper angle to the scale you're reading, it will cause parallax error. It only depends upon the line of sight. These errors can occur both when measuring out reagent or the solution to be tested as well as interpreting results (mLs used). See Fig 9 & 10

1575558831190.png
1575558844232.png

Fig 9 (Graduated Container) Fig 10 (Syringe/Pipette)

Parallax Errors can be avoided by proper positioning of the device so the meniscus (bottom of the curve of the liquid in the device) is perpendicular to your line of sight and on the same level as your line of sight.

2) Procedural Errors
a. Order of addition not followed​
b. Mixing procedures not followed​
i. Mixing/Reaction Times
ii. Mixing intensity not followed (Gentle vs. Vigorous)​
c. Equipment contamination (Dirty vials, syringes etc.)​
(Note on contamination: 1ppm is roughly 7 drops in 60 gal so that means a very-- very small amount is required to contaminate a 10 ml test vial… .000001mL)

d. Temperature considerations not followed (Environment)​
e. Expired test reagent​
f. Bad batch of Test Reagent (It does happen)​
g. Not properly reading the measurement meniscus (Parallax Error)​

The first 5 procedural errors (a-e) can be avoided with four simple steps.

1) Read and follow the instructions for the test kit carefully. Pay attention to the “tips” often included in the instructions.​
2) Keep the equipment (syringes, vials, pipettes, etc.) clean and rinsed! Three complete rinsing after use is recommended, including cap or lid.​
3) Do the test the same way every time, paying close attention to wait times and mixing times and methods (gentle vs. vigorous)​
4) Check and note reagent expiration date.​

To check for the possibility of a “bad” reagent save back enough reagent from the previous batch to do at least 3 additional tests. Run a test using the previous batch then run a test using the new batch. You should get a measured value close (within your standard % Relative error for the test. If your measured value is not acceptable repeat the test two additional times to confirm. I have on two separate occasions found bad reagent so it does happen.

This concludes our look at the first step in The Visual Testing Methods---“Doing the Chemistry (Examination)”. In PART 3 we will look into the Assessment step of Visual Testing…”Looking at the sample”. We will explore human color vision and its impact on the results of our measurements as well as some conclusions and tips to reduce errors following our Quality System Model…Measure----> Evaluate----> Improve---> New Method

QUALITY CYCEL-page-001.jpg









 

TheGulag

Community Member
View Badges
Joined
Jul 22, 2019
Messages
70
Reaction score
60
Location
TN
Rating - 0%
0   0   0
To eliminate the bubbles in the syringe column draw the reagent out of the container with the syringe with the tip fully submerged. Then expel it back into the container fully depressing the syringe. Do this several times to be sure to push out all of the air. On the last draw pull the plunger past the top mark so column is completely full. Hold the syringe tip up and tap the column to release any remaining air. Then slowly depress the plunger until a small amount of liquid comes out of the top. Then eject the extra liquid back into the container until the plunger is at the required mark. Do a visual inspection to be sure there are no remaining air bubbles.

First off really good read and I appreciate you writing this and the past posts, lots of good information all in one location.

One question I have though was on the quoted part above as I have heard and experienced differently. The few times I have done a sudo test it seemed to confirm that by doing what you mentioned I would end up with more fluid in the syringe than anticipated. Compared to making sure the syringe is fully depressed, fully insert into fluid, pull to required mark at a semi slow pace.

The sudo test as I called it was taking 5 and 10mL syringes and filling them both ways(the way you mentioned and the way I mentioned) and measuring the volume via graduated cylinder. Again I didn't do this multiple times and with the multiple syringes that I have for various test kits. Very easily could have been one style of type of syringe that has thrown off my perspective.
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
First off really good read and I appreciate you writing this and the past posts, lots of good information all in one location.

One question I have though was on the quoted part above as I have heard and experienced differently. The few times I have done a sudo test it seemed to confirm that by doing what you mentioned I would end up with more fluid in the syringe than anticipated. Compared to making sure the syringe is fully depressed, fully insert into fluid, pull to required mark at a semi slow pace.

The sudo test as I called it was taking 5 and 10mL syringes and filling them both ways(the way you mentioned and the way I mentioned) and measuring the volume via graduated cylinder. Again I didn't do this multiple times and with the multiple syringes that I have for various test kits. Very easily could have been one style of type of syringe that has thrown off my perspective.


Thanks for your reply...very interesting observation worth checking it out...thanks for the feedback
 

Dempsey941

Valuable Member
View Badges
Joined
Nov 26, 2018
Messages
1,824
Reaction score
3,856
Location
Frisco
Rating - 0%
0   0   0
Thanks for this! I had to have my mom, the nurse, show me how to do this when I got my hanna kits. She was very perplexed over why their syringes are "backwards", so she just brought me a bunch of 1ml and 10ml syringes from work.
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Thanks for this! I had to have my mom, the nurse, show me how to do this when I got my hanna kits. She was very perplexed over why their syringes are "backwards", so she just brought me a bunch of 1ml and 10ml syringes from work.

Thanks for your reply....That is an interesting question...never really thought about it...
 

Daniel@R2R

Living the Reef Life
View Badges
Joined
Nov 18, 2012
Messages
37,507
Reaction score
63,952
Location
Fontana, California
Rating - 100%
1   0   0

Dan_P

5000 Club Member
View Badges
Joined
Sep 21, 2018
Messages
6,675
Reaction score
7,170
Rating - 0%
0   0   0
Great article and really good illustrations on how things can go wrong.
 

stanleo

Valuable Member
View Badges
Joined
Jan 11, 2014
Messages
1,508
Reaction score
1,976
Location
Statesville, NC
Rating - 0%
0   0   0
Some tests require us to add a prescribed number of measuring spoon of a dry reagent to the reaction vessel. If you read the instructions it will often say “Add One Leveled measuring spoon to the vial” The word that is often missed or ignored is “Leveled”. Many tests provide pictorial instructions that show the steps of the chemistry to be preformed. If a measuring spoon is required they just show the reagent being added with the spoon…They don’t show the “Leveling” part. You have to read the instructions to find that part. Not leveling the spoon will result in the addition of too much or too little reagent and introduce variability in the process resulting in reduced accuracy and precision. Leveling the spoon increases accuracy and precision which is demonstrated by the results in Table 6
Great article and very informative. I have a question though. When leveling the spoons, is there a best practice to follow? I tap the spoon on the inside of the bottle and look at the level at eye level to determine if it is level but I am never quite sure if it is, I always shrug and say "close enough". Should use a straight edge to scrape along the rim of the spoon? Should the reagent be packed down?
 

stanleo

Valuable Member
View Badges
Joined
Jan 11, 2014
Messages
1,508
Reaction score
1,976
Location
Statesville, NC
Rating - 0%
0   0   0
2) Keep the equipment (syringes, vials, pipettes, etc.) clean and rinsed! Three complete rinsing after use is recommended, including cap or lid.
What do you mean by "Three complete rinsing after use"?

I use the Red Sea KH/mag/cal kit for those. I rinse the bottles in tap water, then poor RO water in and shake it a few times and dry it out with a paper towel. The syringes I pump with RO water and pump dry a few times to get all the water out. I have never rinsed the caps because I never use them during the testing. The spoons I swirl in RO water and dry with a paper towel. Is this the right way? What is the right way?

Also one of the syringes sticks when plunging it during the titration so it's impossible to add one drop at a time with it. I stopped using it and just use one syringe for calcium and magnesium, rinsing it with the method I mentioned before between uses. I use a different one for the KH and don't rinse that one because I use it every day. Is there a way to fix that or can I just keep using it?
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
What do you mean by "Three complete rinsing after use"?

I use the Red Sea KH/mag/cal kit for those. I rinse the bottles in tap water, then poor RO water in and shake it a few times and dry it out with a paper towel. The syringes I pump with RO water and pump dry a few times to get all the water out. I have never rinsed the caps because I never use them during the testing. The spoons I swirl in RO water and dry with a paper towel. Is this the right way? What is the right way?

Also one of the syringes sticks when plunging it during the titration so it's impossible to add one drop at a time with it. I stopped using it and just use one syringe for calcium and magnesium, rinsing it with the method I mentioned before between uses. I use a different one for the KH and don't rinse that one because I use it every day. Is there a way to fix that or can I just keep using it?

Thanks for your reply...On the rinsing of the test vials and syringes...3 complete rinsing for me is after the test I empty out the remains of any reaction...I fill the vial to the top with RODI water then empty it out...I fill it a second time...not quite to the top...then place the lid or cap on the vial and shake it 10 times or so and then empty it out....Then I fill it a 3ed time all the way to the top and then empty that out and I then let it air dry I don't dry them with a cloth or towel ....some tests actually instruct you to clean the vial with some type of acid...Vinegar.... .1N HCL or citric acid...(example Salfert Potassium Test Kit)
I actually conducted a test which will be in Part 4 of the Article that illustrates the need for at least 3 rinsing

I rinse my syringes a minimum with RODI water 3 times ...I remove the tip and blow air through it because a small amount of rinse water remains in many of the tips...Since I went to using the Single Channel Pipettes I pitch the tips (except for the 10 mL Which I rinse 3-4 times with RODI Water) I do use a standard pipette for the Salfert Mag Test which I rinse according to the above procedure.

As for the "problem" Syringe...I would ditch it and replace it with another....

My practice on leveling the spoon is to use a straight edge as you suggested...I do not pack down the reagent.

Just a note...It helped my testing repeatably to rinsing out the test vial with the water to be tested 1-2 times before you actually conduct the test.

Hope this helps
 

stanleo

Valuable Member
View Badges
Joined
Jan 11, 2014
Messages
1,508
Reaction score
1,976
Location
Statesville, NC
Rating - 0%
0   0   0
Thanks for your reply...On the rinsing of the test vials and syringes...3 complete rinsing for me is after the test I empty out the remains of any reaction...I fill the vial to the top with RODI water then empty it out...I fill it a second time...not quite to the top...then place the lid or cap on the vial and shake it 10 times or so and then empty it out....Then I fill it a 3ed time all the way to the top and then empty that out and I then let it air dry I don't dry them with a cloth or towel ....some tests actually instruct you to clean the vial with some type of acid...Vinegar.... .1N HCL or citric acid...(example Salfert Potassium Test Kit)
I actually conducted a test which will be in Part 4 of the Article that illustrates the need for at least 3 rinsing

I rinse my syringes a minimum with RODI water 3 times ...I remove the tip and blow air through it because a small amount of rinse water remains in many of the tips...Since I went to using the Single Channel Pipettes I pitch the tips (except for the 10 mL Which I rinse 3-4 times with RODI Water) I do use a standard pipette for the Salfert Mag Test which I rinse according to the above procedure.

As for the "problem" Syringe...I would ditch it and replace it with another....

My practice on leveling the spoon is to use a straight edge as you suggested...I do not pack down the reagent.

Just a note...It helped my testing repeatably to rinsing out the test vial with the water to be tested 1-2 times before you actually conduct the test.

Hope this helps
Thank you for you response, that does help. I will do it this from now on.
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
First off really good read and I appreciate you writing this and the past posts, lots of good information all in one location.

One question I have though was on the quoted part above as I have heard and experienced differently. The few times I have done a sudo test it seemed to confirm that by doing what you mentioned I would end up with more fluid in the syringe than anticipated. Compared to making sure the syringe is fully depressed, fully insert into fluid, pull to required mark at a semi slow pace.

The sudo test as I called it was taking 5 and 10mL syringes and filling them both ways(the way you mentioned and the way I mentioned) and measuring the volume via graduated cylinder. Again I didn't do this multiple times and with the multiple syringes that I have for various test kits. Very easily could have been one style of type of syringe that has thrown off my perspective.

So I checked out your question with regards to the syringe filling procedure. I use 2 different syringes. One 1 mL and one 2 mL. The 1 mL was a higher quality syringe and the 2 mL was one that came with a test kit...not as good a quality. I mesured the results by weight not volume...1 mL would = 1 gram of distilled wate....from what I understand this is a more accurate way of gaging the accuracy of a syringe or pipette...The Pipette manufactures recommend it.

The attached table are the results...based on this it looks as if it is 6 of one half a dozen of the other...What do you think?...The main reason I use the multi draw method is some syringes are more prone to trapping air and the multi draw method tends to purge it out...But the single draw is much easier and quicker

syrange filling procedure test-page-001.jpg
 

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,545
Reaction score
10,102
Rating - 0%
0   0   0
Excellent work Rick! Some of these errors are things always in the back of my mind - are all drops really the same? And others are errors I hadn't considered. Thanks for answering them all in great detail.
 

Dan_P

5000 Club Member
View Badges
Joined
Sep 21, 2018
Messages
6,675
Reaction score
7,170
Rating - 0%
0   0   0
Just remembered to tell you that your ideas for quality have influenced my cooking. I started moving to weighing ingredients on a digital scale to avoid measuring volumes. I expect consistency to improve! Your lessons don’t inform us about the downsides of a glass of wine during meal preparation though.
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Just remembered to tell you that your ideas for quality have influenced my cooking. I started moving to weighing ingredients on a digital scale to avoid measuring volumes. I expect consistency to improve! Your lessons don’t inform us about the downsides of a glass of wine during meal preparation though.

Down sides...What down sides... :oops:
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Excellent work Rick! Some of these errors are things always in the back of my mind - are all drops really the same? And others are errors I hadn't considered. Thanks for answering them all in great detail.

Thank you Sir...appreciate you help and support of doing this...The work you and Dan P are doing is inspiring!!!
 