When I took down my tank 8 years ago, I got rid of nearly everything. One thing that remained in the fish room cupboard is a ceramic mug containing a bunch of pH probes, some of which are pretty high end. For example, two are Ross pH electrodes. Fisher Scientific currently is selling them for $722 each (I didn't originally buy them, but got them used from places I worked for that no longer needed them).


Once upon a time they were in some type of water in the mug. I don't recall what it was. Maybe tank water, but probably tap water.

A couple of weeks ago, I went to figure out if anything was salvageable. I tried to pick up the mug by the handle and the mug literally broke into a dozen pieces. Of all the things I thought I might find, that was not among them. It was from a family reunion long ago. They probably bought the cheapest mugs available. lol

Of course, there was no water left, just dried salts. Even inside the probes they looked entirely dried out.

The glass bulb of a typical pH probe is a very important piece of hydrated thin glass that allows measurement of pH. The way it does that is surprisingly complicated, and I discuss it here for anyone interested in more details:


This little section from it summarizes the effect:

The glass used in a pH electrode has a composition that includes certain metals in addition to the usual silicon and oxygen of SiO2 glass. These metals can include lithium, barium, lanthanum, sodium, and calcium, among others. At both the inside and outside surfaces of the glass bulb (and also in a thin (50 nm) swollen region of hydrated glass on each side), there is some substitution of H+ for these cations. How much substitution depends on how much H+ is in solution, and hence on pH. The internal pH does not change, but the external pH does, so the amount of substitution on the outer surface changes while on the inside it does not, setting up a potential difference across the glass membrane that depends on pH.

Note in that description is the mention of a hydrated layer on the glass. That is why one does not want them to dry out. A pH electrode that has been sitting dry for years has certainly changed the nature of the hydrated glass, and it may well never be recoverable.

But why not try?

So this is a long term experiment to see if I can get any of these probes suitably working again.

First step was to refill the insides of them. All (IIRC) were refillable through a hole in the body. Some pH electrodes are sealed, but not these. I only had one filling solution available (also sitting on the fish room shelf). Some different pH electrodes use different filling solutions, but I'm not sure what happens if they are not exactly matched. Most are a fairly concentrated potassium chloride solution, and some are saturated with silver chloride as well. I think it was a Ross filling solution I used, but I'll have to double check that.

After filling them all, the next step was to try to rehydrate them on the outside. I'm not sure if that will work or how long it might take. I placed them all into a container of tap water and started them off soaking.

In the near future, I'll be seeing if I can get an old pH meter working again (found a meter behind a box in the old fish room but so far, not the cord lol) and then see if these are at all responsive to pH.

Assuming they do respond to pH, then I'll see if they can be used with actual calibration fluids to give useful values.

So no answers yet, just a long slow work in progress.

Happy reefing. :)
 

Miami Reef

Reef Chem Enthusiast
View Badges
Joined
Sep 8, 2017
Messages
17,255
Reaction score
29,618
Location
Miami
Rating - 0%
0   0   0
Following
 

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
7,223
Reaction score
10,586
Rating - 0%
0   0   0
if you want to make a detail-oriented person crazy, give them 3-4 cheap, neglected pH probes and some opened undated calibration buffer solutions.
 
OP
OP
Randy Holmes-Farley

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,259
Reaction score
92,294
Location
Massachusetts, United States
Rating - 0%
0   0   0
if you want to make a detail-oriented person crazy, give them 3-4 cheap, neglected pH probes and some opened undated calibration buffer solutions.
lol
a recipe for frustration
 

Miami Reef

Reef Chem Enthusiast
View Badges
Joined
Sep 8, 2017
Messages
17,255
Reaction score
29,618
Location
Miami
Rating - 0%
0   0   0
if you want to make a detail-oriented person crazy, give them 3-4 cheap, neglected pH probes and some opened undated calibration buffer solutions.
My eye just twitched.
 

Dan_P

7500 Club Member
View Badges
Joined
Sep 21, 2018
Messages
9,785
Reaction score
9,638
Rating - 0%
0   0   0
When I took down my tank 8 years ago, I got rid of nearly everything. One thing that remained in the fish room cupboard is a ceramic mug containing a bunch of pH probes, some of which are pretty high end. For example, two are Ross pH electrodes. Fisher Scientific currently is selling them for $722 each (I didn't originally buy them, but got them used from places I worked for that no longer needed them).


Once upon a time they were in some type of water in the mug. I don't recall what it was. Maybe tank water, but probably tap water.

A couple of weeks ago, I went to figure out if anything was salvageable. I tried to pick up the mug by the handle and the mug literally broke into a dozen pieces. Of all the things I thought I might find, that was not among them. It was from a family reunion long ago. They probably bought the cheapest mugs available. lol

Of course, there was no water left, just dried salts. Even inside the probes they looked entirely dried out.

The glass bulb of a typical pH probe is a very important piece of hydrated thin glass that allows measurement of pH. The way it does that is surprisingly complicated, and I discuss it here for anyone interested in more details:


This little section from it summarizes the effect:

The glass used in a pH electrode has a composition that includes certain metals in addition to the usual silicon and oxygen of SiO2 glass. These metals can include lithium, barium, lanthanum, sodium, and calcium, among others. At both the inside and outside surfaces of the glass bulb (and also in a thin (50 nm) swollen region of hydrated glass on each side), there is some substitution of H+ for these cations. How much substitution depends on how much H+ is in solution, and hence on pH. The internal pH does not change, but the external pH does, so the amount of substitution on the outer surface changes while on the inside it does not, setting up a potential difference across the glass membrane that depends on pH.

Note in that description is the mention of a hydrated layer on the glass. That is why one does not want them to dry out. A pH electrode that has been sitting dry for years has certainly changed the nature of the hydrated glass, and it may well never be recoverable.

But why not try?

So this is a long term experiment to see if I can get any of these probes suitably working again.

First step was to refill the insides of them. All (IIRC) were refillable through a hole in the body. Some pH electrodes are sealed, but not these. I only had one filling solution available (also sitting on the fish room shelf). Some different pH electrodes use different filling solutions, but I'm not sure what happens if they are not exactly matched. Most are a fairly concentrated potassium chloride solution, and some are saturated with silver chloride as well. I think it was a Ross filling solution I used, but I'll have to double check that.

After filling them all, the next step was to try to rehydrate them on the outside. I'm not sure if that will work or how long it might take. I placed them all into a container of tap water and started them off soaking.

In the near future, I'll be seeing if I can get an old pH meter working again (found a meter behind a box in the old fish room but so far, not the cord lol) and then see if these are at all responsive to pH.

Assuming they do respond to pH, then I'll see if they can be used with actual calibration fluids to give useful values.

So no answers yet, just a long slow work in progress.

Happy reefing. :)
There is rehydrating the glass and there is cleaning the surfaces. What is the plan for cleaning?
 
OP
OP
Randy Holmes-Farley

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,259
Reaction score
92,294
Location
Massachusetts, United States
Rating - 0%
0   0   0
There is rehydrating the glass and there is cleaning the surfaces. What is the plan for cleaning?

I’ll likely soak the tips in an acid if it seems promising. :)
 

Johnd651

Getting back in after 14+ years
View Badges
Joined
Oct 29, 2019
Messages
4,866
Reaction score
7,008
Location
Upstate NY
Rating - 100%
1   0   0
So not over years, but we had left probes out in various labs I worked for, over a period of days, weeks, maybe even a month or two. We were usually able to rehydrate them, and a lot of times would use a 3 point calibration, so they ended up working. How far the drift was after, I never looked.
 

DaJMasta

Valuable Member
View Badges
Joined
Feb 3, 2022
Messages
1,175
Reaction score
1,431
Location
Maryland
Rating - 0%
0   0   0
It was a few years ago and I haven't been able to find the source of the method, but I remember trying to revive a dry YSI pH probe (sealed, glass ball) with a soak in a strong acid - 1M HCl. Not friendly stuff nor something to recommend to people, and even while it appeared clean (most recommendations seem to be for dissolving residue on them). What I tried was DI to rinse, HCl to soak, and then either storage buffer or DI again (can't remember, but the instructions for the sensor assembly is to keep distilled water in the cup on the sensor head, not a storage buffer), and it was responsive to changes in pH appropriately, though it tended to read a little lower than expected after two or three point calibration. Checking a reference solution after calibration showed it wasn't perfectly repeatable, but it wasn't too far off (would be usable for hobby testing).

Don't know about the lifetime of it as I eventually just replaced the probe.
 

Red_Beard

I already did
View Badges
Joined
Feb 17, 2019
Messages
6,798
Reaction score
18,960
Location
Utah
Rating - 0%
0   0   0
Following. I have 2 Hannah ph meters that have been sitting for 2 years. Going to play copycat if if works.
 

vahegan

Active Member
View Badges
Joined
Sep 9, 2019
Messages
247
Reaction score
181
Rating - 0%
0   0   0
How long do you think it might take to rehydrate 50 nanometers of glass?
 

BeanAnimal

7500 Club Member
View Badges
Joined
Jul 16, 2009
Messages
9,284
Reaction score
15,575
Rating - 0%
0   0   0
But why not try?

So this is a long term experiment to see if I can get any of these probes suitably working again.

So no answers yet, just a long slow work in progress.

Happy reefing. :)

You are literally 2 days too late.

I had a few highish end probes that I got along with some walchem readers from a biotech bankruptcy a decade or so ago. I was cleaning out the fish room Sunday and found them in drawer in their wet storage holders, dried up.

I actually wondered if they could be brought back to life, but in trying to turn over a new leaf and scale back on "stuff" -- I threw them out. Now I have garbage remorse. Thanks Randy, I can only hope that you come back in three weeks and says "this failed" - otherwise I will be at the dump searching for my pH probes.
 
Last edited:

vahegan

Active Member
View Badges
Joined
Sep 9, 2019
Messages
247
Reaction score
181
Rating - 0%
0   0   0
I actually wondered if they could be brought back to life, but in trying to turn over a new leaf and scale back on "stuff" - I threw them out. Now I have garbage remorse. Thanks Randy, I can only hope that you come back in three weeks and says "this failed" - otherwise i will be at the dump searching form my pH probes.
That's what always happens to me! I didn't know that it was scientifically named garbage remorse ;) But as soon as I put something in the garbage, even though it was littering my very limited space for decades and I finally decided I won't ever use it, I will suddenly discover an immediate use exactly for the thrown part on the very next day, or at least within a week after the event...
 

Red_Beard

I already did
View Badges
Joined
Feb 17, 2019
Messages
6,798
Reaction score
18,960
Location
Utah
Rating - 0%
0   0   0
Lol, garbage remorse.. :rolling-on-the-floor-laughing:
Nolan North Hoarder GIF by RETRO REPLAY

I can totally relate though, and it isnt hoarding if it is still potentially useful...
 

TOP 10 Trending Threads

WHAT AMOUNT OF LIVE ROCK AND SAND SHOULD BE PRIORITIZED FOR OPTIMAL BIODIVERSITY/FILTRATION?

  • 100% live rock + bagged sand

    Votes: 37 27.6%
  • 100% dry rock + 100% live sand

    Votes: 45 33.6%
  • 50/50 live/dry rock, 50/50 live/bagged sand

    Votes: 30 22.4%
  • 75% live rock, 25% live sand

    Votes: 12 9.0%
  • 25% live rock, 75% live sand

    Votes: 10 7.5%
Back
Top