I posted this on Neptune System Community, I got a few views but no replies
I recently added a sump and have been running the Apex system for about 8 months, while making the transition I soaked the probes in vinegar and water, setup the sump and switched the probes and began to calibrate. The salinity probe would not calibrate in automatic mode (probably due to being outside of parameters), so off to the forums I went to investigate. here is what I found
I manually calibrated the probe and switched to TC factor 2.2, result was a reading of 68 ppt and upon recalibrating 45 ppt, all testing was done at sump temperature.
I used a hanna salinity tester (calibrated to insure accuracy) the sump reading was 35.5 ppt
I made sure that there were no air bubbles in the probe and placed it in the return bin of the sump, Apex reading currently reading 42.8 ppt actual 34.9 per Hanna
I purchased another probe in case I damaged the original during the transition, it is due to arrive this week.
Now my suggestion, it seems that several posts relating to the salinity probe have had the same issues of salinity readings being off by an unacceptable margin. A possible future feature for Fusion could be the ability to manually set the calibration to where the user has verified its actually value, much like that in calibrating the temperature probe.
I recently added a sump and have been running the Apex system for about 8 months, while making the transition I soaked the probes in vinegar and water, setup the sump and switched the probes and began to calibrate. The salinity probe would not calibrate in automatic mode (probably due to being outside of parameters), so off to the forums I went to investigate. here is what I found
I manually calibrated the probe and switched to TC factor 2.2, result was a reading of 68 ppt and upon recalibrating 45 ppt, all testing was done at sump temperature.
I used a hanna salinity tester (calibrated to insure accuracy) the sump reading was 35.5 ppt
I made sure that there were no air bubbles in the probe and placed it in the return bin of the sump, Apex reading currently reading 42.8 ppt actual 34.9 per Hanna
I purchased another probe in case I damaged the original during the transition, it is due to arrive this week.
Now my suggestion, it seems that several posts relating to the salinity probe have had the same issues of salinity readings being off by an unacceptable margin. A possible future feature for Fusion could be the ability to manually set the calibration to where the user has verified its actually value, much like that in calibrating the temperature probe.