i've had mine for 2 years now. when it works and dialed it, it's great. but it is so dang finicky. if you sneeze around it, something will fail.
anyways, i had a heck of a time with test c failing. I finally figured it out. it was a bad batch of c reagent. replaced it, recalibrated it and it fixed it. then low and behold, couple weeks later, did a reagent change and now test b is showing CA 0. finally figured it out as my cuvette was dirty. cleaned it out, recalibrated it and now back on track. But I am realizing the older the unit gets, the more prone it is to failing at something. I am contemplating sending the unit in for maintenance, but so far it is back on track.
any of you guys also experience this?
anyways, i had a heck of a time with test c failing. I finally figured it out. it was a bad batch of c reagent. replaced it, recalibrated it and it fixed it. then low and behold, couple weeks later, did a reagent change and now test b is showing CA 0. finally figured it out as my cuvette was dirty. cleaned it out, recalibrated it and now back on track. But I am realizing the older the unit gets, the more prone it is to failing at something. I am contemplating sending the unit in for maintenance, but so far it is back on track.