This is a follow-up to my post regarding my solution to the apparent drift over time of the Hanna Alkalinity Test. I believe I witnessed the drift years ago, and began using small 10ml glass vials to store the reagent. With that change the drift seemingly disappeared.
In an attempt to recreate the drift, I performed a full reagent bottle of tests (24) over the last three months, and compared the results of the reagent stored in the Hanna bottle to the results from reagent stored in my 10ml glass vials. Both methods used reagent from Lot 1293, and a commercial alkalinity standard was used at the beginning and end of the test for calibration and verification.
The number of days into the test, test date, alkalinity of each test, and difference was recorded. The difference between the tests verses time was plotted. The test ran until the reagents were consumed.
At no point during the test was there any significant difference between the two test methods. All of the tests tracked within 0.1 dKH of each other. Verification of both tests to the alkalinity standard at the end of the test revealed that neither reagent drifted.
The cause of the apparent drift remains a mystery.