My Trident has been working flawlessly for about a year until two days ago. It was reading Alk at about 8.0 - 8.1 and then the next test showed Alk at 13.3 and has been reading high since. What I've done so far:
1. Replaced reagents.
2. Calibrated with solution.
3. Tested with Hannah - tank reads 8.3
4. Tested calibration fluid with Hannah- was spot on
5. Removed black sample hose and ran rodi through it with a syringe
6. Checked reagent lines for kinks / disconnected metal tube
7. Primed sample line twice
8. Trident now reads 13.37
Any advice on what to do next would be greatly appreciated!