Out of the box after the software update and factory reset, my nitrates are spot on, but the phosphate reading is extremely inaccurate. After spending way too much time troubleshooting, I have identified two significant problems and sources of error.
1. The thin black water sample tubing grows an internal biofilm that significantly reduces the phosphate level of the sample water before it reaches the tester. In some cases the tubing can reduce the phosphate level of the sample water to essentially zero before it reached the tester.
2. There is a fixed offset or shift of the tester’s phosphate reading that seems to be different with every batch of reagents. For my first batch I had to subtract 0.425 ppm from the reading to get the correct result. With the second batch, I have to subtract 0.035 ppm.
The good news is as long as the tester has a good water sample, and you subtract the offset, you can seemingly get an accurate phosphate reading. The tester consistently and accurately reports a 0.100 ppm phosphate spike to the sample, and this is very encouraging.
The ball is in Neptune’s court. They need to come up with a fix to get an uncompromised water sample into the tester, and they should allow users to calibrate the device like the original Trident.