OK! Obviously an update long time overdue!
First, and as I said in my first post from 2024-09-20 (and I quote): "But, realize that my reason to get automatic testing is that my calendar is very aggressive, and so I can't make any promise how or at what rate this build thread will progress!". Well, I kept that promise (that is, I kept the no promise...)!
Here is a recap (dates in ISO 8601 standard so only North Korea, Palau, Tuvalu, South Sudan and some other small island nations should not understand; I do my best and my part to unify the planet...!) - please see original posts above for details:
2024-09-10 Delivery of my GHL ION Director
2024-09-20 Installation (and first post) - basically no problem and easy, BUT requiring multiple tightening of the sensor attachment the following days to avoid leaks
2024-09-21 Immediately obvious that the unit is drinking reference fluid like an elephant on the savanna
2024-09-22 First hard facts (compared to Hanna):
* Low accuracy, but very high precision.
* Reference fluid consumption extremely high (in average per measurement as of the first 10 measurements 42.0 ml Ref A and 40.2 ml Ref B)
* Graphs in app highly misleading
2024-09-23 Recalibration and new confirming Hanna tests: same result
2024-09-25 I offset the values according to Hanna - then good values (consistent with a state of low accuracy, high precision). BUT rapidly degrading sensor (after 4 days and 10 measurements down to 72%). This despite troubleshooting according to manual
2024-09-26 Sensor suddenly went good (96%)
2024-09-28 Sensor again dropping (83%) (and reference fluid consumption remaining ridiculously high)
2024-10-06 Ordered a new sensor from GHL (estimated delivery "approx. 6-8 weeks"). Figured that even if the current sensor was replaced, I would soon need a new - judging from other users input...
2024-10-09 The sensor died completely (0%) - thus some three weeks and with whooping reference consumption, low accuracy, but high precision. Ordered more reference fluid, and also had received some IOND Multi-Reference -thus true reference fluid to measure on - but had no functioning sensor...
2024-10-26 Filed a ticket with GHL regarding the dead sensor.
2024-10-26 Immediate response from GHL requesting further information
2024-11-23 A very delayed response from me giving the additional information (including pictures)
---------------- New Updates ----------------
2024-12-04 No response so I wrote a reminder to GHL
2024-12-29 No response so I wrote a reminder to GHL
Gave up...
2025-01-20 Suddenly and without warning the sensor I ordered 2024-10-06 was delivered! Instead of the stated 6-8 weeks, it took 106 days (or some 15 weeks).
2025-02-11 (today) Got a mail from GHL apologizing for the late reply and asking for my shipping address to replace the erroneous sensor. Great! But a little late...
So, here are my conclusions:
IOD-wise:
1. Very high reference consumption - some 40 ml of each reference fluid per measurement
2. Questionable accuracy - significant and consistent difference to Hanna and ICP-OES (Reef Factory) (sorry, have no ICP-MS)
3. High precision with very consistent values between measurements
4. Extremely short sensor life time
GHL-wise:
1. Extremely slow resolution of my ticket despite reminders - yes, definitely also delay from me
2. Extremely slow delivery of newly ordered sensor - given even normal and anticipated sensor life time this would be a time-limiting factor
Personal-wise:
1. I've come to realize that the time I spent (even on a well functioning system) is well beyond the time I spend with my Hannas.
2. Even if everything is working like a charm, I still want to confirm with Hannas and ICP, just to make sure/confirm that everything is indeed working like a charm. So what is then the use of automated testing?
Conclusion:
1. I've given up the idea of automated testing. It costs (moneywize, timewize, energywize, everythingwize) more than it gives and more compared to Hannas and ICP (even ME) once a week.
2. My tank doesn't need it! During this time something has happened to my tank - it's matured! Everything is thriving and growing, all parameters good and stable no matter what I throw on them - so (I'm sorry to say) scrap this forever and constant testing - I haven't graduated to SPS anyhow... And the day I do graduate to SPS I understand I will have the time doing Hannas as I'm going/have to stay with my tank constantly anyhow...
Conclusion of the conclusion:
Anybody want a GHL IOD with two brand new sensors (if the replacement is shipped as promised, otherwise just one), 1000 ml IOND Multi-Reference fluid and 4000 ml each of Reference A and B? Yours for free, but two terms:
1. You pay tracked shipping (from Sweden, to where ever you live)
2. You continue the evaluation and post the results here