GHL KHD Reading Low

  • Thread starter Thread starter Cheesus
  • Start date Start date
  • Tagged users Tagged users None
Yeah it is a constant 6.8 but my tank is at a 7.7 because I had a dosing pump go crazy on me earlier today...so that was fun.

Not sure of any other reason I would be reading so far off. If the heads were worn out some even after calibration would that cause an issue like this?
 
yeah I did the calibration with a test tube, recorded the numbers and saved it.
I am super confused.
 
None, the reagent sits directly under the doser.
I did just change that though, it was sitting to the side, running a test now.
 
It’s the pinch valve Normal. I mean you are getting consistent results. Maybe calibrate the probe again. Also make sure you refill with some saltwater since you’re spending 80 mls a test. If your rodi is running you don’t want to drop your sg
 
It’s the pinch valve Normal. I mean you are getting consistent results. Maybe calibrate the probe again. Also make sure you refill with some saltwater since you’re spending 80 mls a test. If your rodi is running you don’t want to drop your sg
Yeah I have the free dose head dosing 80 ml of saltwater lol

I can try that out tomorrow, Its a pain to calibrate.
 
It’s the pinch valve Normal. I mean you are getting consistent results. Maybe calibrate the probe again. Also make sure you refill with some saltwater since you’re spending 80 mls a test. If your rodi is running you don’t want to drop your sg
Hey what is your firmware version? Mine is from 2020 I think
 
Where exactly do you measure for the sample tube volume or how many ml” , from
The filter to tube inserting into the dosing beige tubing ?
 
Where exactly do you measure for the sample tube volume or how many ml” , from
The filter to tube inserting into the dosing beige tubing ?
You measure all of it, from the sump all the way to where it goes into the KHD.
 
Mine is version 1.00 I don’t think there is an updated version
So this morning I recalibrated the sample line, recalibrated the probes, same results.

I saw something about moving the filter to the sump, so I cut new sample lines and when I went to put the filter on I noticed that it said honda..interesting. Decided to test without it, and have had the closest reading yet, KHD 6.6 Salifert 7.1.

I just ran one test but there's one on the schedule for 12 ill let run and see if it improves.
 
Aaaand it went back down to 6.2. I put in a support ticket yesterday so
Hopefully I’ll hear something soon. This is incredibly frustrating.
 
I hope so. Any other suggestions I can do?
You can open up the box and double check that the reagent line has no air bubbles. Also make sure the end of the line that goes into the bottle isn’t getting stuck to the bottom of the bottle in a vacuum. To me it seems there isn’t enough reagent getting into the sample. Or too much sample. How big is your tank?
 
That’s an interesting theory. I’m gonna go test that. It’s a 125 with probably a 30 gallon sump
 

TOP 10 Trending Threads

WHAT FIRST SPARKED YOUR INTEREST IN REEFING?

  • I saw a reef tank that hooked me

    Votes: 11 28.9%
  • I was already keeping freshwater aquariums

    Votes: 13 34.2%
  • A particular fish got me interested

    Votes: 7 18.4%
  • Corals and inverts were the draw

    Votes: 6 15.8%
  • A friend or family member introduced me

    Votes: 5 13.2%
  • The science and ecosystem challenge appealed to me

    Votes: 6 15.8%
  • I stumbled into it… and here we are

    Votes: 3 7.9%
  • Other (Post in the comments)

    Votes: 6 15.8%
Back
Top
Home
Post thread…
Market
What's new