How is the Mastertronic working for the lucky ones that have been able to buy it?
I feel comfortable with most of my results but I think that the discrepancy in the Magnesium tests is too big for my taste. Specially as it gets close to the upper limit and you lose track of the trend.
Below you can see charts of the results of the last 50 days. They include the Mastertronic automatic tests and the other test kits I have tried along with 2 ICP tests (yellow line on Calcium and Magnesium).
Note: On September 11 I ran a descaling program using Hanna's probe cleaning solution. You can see a dip in the charts (improvement) for Calcium and Magnesium and more correlation with other test kits.
I feel comfortable with most of my results but I think that the discrepancy in the Magnesium tests is too big for my taste. Specially as it gets close to the upper limit and you lose track of the trend.
Below you can see charts of the results of the last 50 days. They include the Mastertronic automatic tests and the other test kits I have tried along with 2 ICP tests (yellow line on Calcium and Magnesium).
- Nitrate results are frequently around 5ppm lower than Hanna and the time I compared it with Nyos and Redsea (September 18) it fell in between. Descaling the test chamber on September 12 didn't show much change . The discrepancy is high (4.5ppm) but I think it is consistent and doesn't worry me too much.
- Calcium was initially around 60ppm higher than ICP and 34ppm higher than Hanna but got really close to the ICP result (14ppm) and the Hanna result (25ppm) after I descaled the test chamber. Now I'm very comfortable with the accuracy.
- Phosphate has shown a difference of less than 0.01ppm every time I have compared it against Hanna so I'm pretty confident with it too.
- Magnesium is the one that makes me uncomfortable, for some reason up until I descaled the test chamber with probe cleaning solution, Magnesium was always shown between 1578 and 1579 ppm. The consistency seemed amusing until I realized that it was the upper limit :S
Before descaling the test chamber, the difference between Mastertronic and ICP was 207 ppm. On the second ICP after the descaling, the difference decreased to 147 ppm, better but its too big for my taste. I have also tested using Red Sea (manually) and with Aquaforest. Interestingly, the results of the last two seem consistent but being manual titration I would not discard human bias in perceiving the color change (although I don't see the measurement until the test is finished).
Note: On September 11 I ran a descaling program using Hanna's probe cleaning solution. You can see a dip in the charts (improvement) for Calcium and Magnesium and more correlation with other test kits.