I just opened a ticket. I have run into what feels like a larger doser 3 issue. It may be of my own making or misunderstanding.
@KStatefan @Lasse, thoughts?
Below is the text of the ticket (typos and all) that I am waiting for response on.
"
Hi - I have upgraded my doser 2s to 3s using the upgrade kit.
1- firmware 1.55
2 - Set to doser 3 in PAB configuration.
Observations: At slow speeds some heads still visibly stick/stall momentarily and chatter before breaking free. This is random. Some heads do not exhibit this behavior. I have tried breaking in the tubing sets by simply running RO water for a few liters. It helped on most heads, but I still have issues on others.
Testing:
One of the heads that does not appear outwardly to be stalling: I have setup a 500ml container and set a nightly scheduled dose of 500ml.
Speed = 0 slow
Calibration with RO water. Dosing pump 11
1 - Set 100ml actual and desired. Clicked "Save" (I will explain below in notes)
2 - Close and re-open doser 11 to ensure settings are being used.
3 - Click "calibrate" and dose into an appropriate graduated cylinder zeroed on a .001 balance.
4 - record actual calibration dosage in dosing pump 11 and SAVE. This is 11ml min with the current suction and discharge tubing.
5 - Set "desired dose rate" to 1ml min
The dose runs, but of the 500ml prescribed dose, only ~250 ml of it is delivered over the expected over the 8.33 hour dosing run.
This is a tremendous error. Ignoring the single dosing run, this error would be the same if it was 1ml 500 times over a period of days or weeks.
I this a math issue with division of steps? Is this a doser accuracy issue compounding? Is this a Speed 0 issue vs Speed 1 and math/dosing calculation?
I have repeated this test numerous times and the results are basically the same. I have not measured exactly how much is delivered or left or attempted to work backwards to step math.
NOTE: from above. If "calibrate" is clicked with a "desired fkow rate" lower than actual then the pump appears to run at that rate and not "0 slow" full RPM. This is not intuitive, but appears to allow one to calibrate at the "desired flow rate". However that begs the question. If I set 11 mL/min actual and 1 mL/min desired and the flow test produces .88 ml in one minuted, there is no way to adjust for that. Changing the ACTUAL would then again change the the output and this would be an exercise in fudging "actual" to get "desired to match.
I am at a loss here at many levels. Can you please offer some insight?
I am afraid to upgrade my actual prime dosers for Alk/Ca/Mag/Etc and the doser for the KH director at this point. "
Hopefully this is just me selecting the wrong dosing speed and the math not working out with steps or something like that. As for the sticking heads, I don't know and I don't know if the head that I am using is not visibly sticking but still significantly missing steps.
Hopefully I get answers soon.