Do you make a fresh batch each time? Or would this be able to be stored and used for a while if in a sealed container?Tank Salinity
I made my 35 ppt/53 mS/cm conductivity standard using table salt:
dissolve 1 teaspoon (6.20 grams) of Morton's Iodized Salt in 182 mL (182 g) of fresh water (making a total volume of about 184 mL after dissolution of the salt). This solution can be scaled up as desired.
![]()
Reef Aquarium Salinity: DIY Calibration Standards
Reef Aquarium Salinity: DIY Calibration Standards By Randy Holmes-Farley One of the most important issues facing marine aquarists is providing a suitable environment for their aquaria's inhabitants. Among the important properties for a marine...www.reef2reef.com
I first checked my Pinpoint Salinity device in it. It read a bit low (which I knew). It read 51.1 mS/cm in the 53 mS/cm solution. Since I cannot calibrate it, I will just multiply all readings with it by 53/51.1 =1.037.
I measured tank water to be 49.8 mS/cm. Multiplying up that gives 51.65 mS/cm, which is equivalent to 34.1 ppt or sg = 1.0257.
I’ll ensure that new salt water for AWC has at least 35 ppt in it, then just watch the tank salinity for a while to see if there’s a trend.