I am just getting back into reefing after being out for ~15 years. I bought a generic refractometer from my LFS a couple of months ago and something I wasn't aware to watch out for was how out of calibration a new refractometer can be. Bad (old) newbie mistake, to be sure.
Anyway - in trying to fix the problem, I bought some Fritz calibration fluid and thought that would be that - boy was I wrong!
The first thing I noticed was that my generic refractometer appears to be utter garbage. Readings for the Fritz solution would fall out of line within hours of calibration. This is including waiting for temperatures to equalize and a rigorous rinse/cleaning both before calibration and before readings. The readings swung by more then .003-4 within hours. So - I decided to remedy that problem.
I bought a used Milwaukee seawater digital refractometer off of Ebay - for a bargain price (~$55) and thought that would be that. When it got here I was amazed at the simplicity of using it etc. but then I got the bright idea of measuring my Fritz calibration solution. It showed 1.025. I understand from some other reading on here that the Milwaukee aims for a +-.002 level of tolerance for readings, so that is within margin of error. That said, it was disconcerting to say the least... so...
I bought more calibration solution just to see what is going on. I purchased a bottle of Brightwell Refractometer calibration Standard and BRS Refracto Juice for comparison (not knowing that the BRS solution was manufactured by Brightwell... more on that below).
So... both of those new ones arrived today and I went and tested again. Here are the results:
Fritz: 1.025 / 34ppt
Brightwell: 1.025 / 34ppt
BRS: 1.024 / 33ppt
These measurements were repeatable too. I ran the test at least twice for each solution.
Soooooo... none of them come to the target number, and oddly enough, the BRS and Brightwell have different results despite being the same manufacturer and probably the same recipe/composition.
This all just leaves me wondering - why is this so hard? and from here... should I just count on the Milwaukee to be .002 lower than expected, so aim for 1.028 if I really want a measurement of 1.024?
Lastly - I can't believe that my LFS sold me that junk _without_ calibration solution. I want to support them, but that seems like a pretty bad move. I think I caught it in time for it not to be a major issue with my newish tank, but not a fun experience to go through. I know I bear a lot of the blame for being under-educated there, but still.
Anyway - in trying to fix the problem, I bought some Fritz calibration fluid and thought that would be that - boy was I wrong!
The first thing I noticed was that my generic refractometer appears to be utter garbage. Readings for the Fritz solution would fall out of line within hours of calibration. This is including waiting for temperatures to equalize and a rigorous rinse/cleaning both before calibration and before readings. The readings swung by more then .003-4 within hours. So - I decided to remedy that problem.
I bought a used Milwaukee seawater digital refractometer off of Ebay - for a bargain price (~$55) and thought that would be that. When it got here I was amazed at the simplicity of using it etc. but then I got the bright idea of measuring my Fritz calibration solution. It showed 1.025. I understand from some other reading on here that the Milwaukee aims for a +-.002 level of tolerance for readings, so that is within margin of error. That said, it was disconcerting to say the least... so...
I bought more calibration solution just to see what is going on. I purchased a bottle of Brightwell Refractometer calibration Standard and BRS Refracto Juice for comparison (not knowing that the BRS solution was manufactured by Brightwell... more on that below).
So... both of those new ones arrived today and I went and tested again. Here are the results:
Fritz: 1.025 / 34ppt
Brightwell: 1.025 / 34ppt
BRS: 1.024 / 33ppt
These measurements were repeatable too. I ran the test at least twice for each solution.
Soooooo... none of them come to the target number, and oddly enough, the BRS and Brightwell have different results despite being the same manufacturer and probably the same recipe/composition.
This all just leaves me wondering - why is this so hard? and from here... should I just count on the Milwaukee to be .002 lower than expected, so aim for 1.028 if I really want a measurement of 1.024?
Lastly - I can't believe that my LFS sold me that junk _without_ calibration solution. I want to support them, but that seems like a pretty bad move. I think I caught it in time for it not to be a major issue with my newish tank, but not a fun experience to go through. I know I bear a lot of the blame for being under-educated there, but still.