Here's what I do: Use Apex calibration solution. Put the bottle at the same height as the end of the black tubing as it is in the tank, but put the bottle outside of the tank. Then put the tube into the calibration bottle. The tube has to be close to operating configuration because the back pressure on the line is important for accuracy.
Prime the sample line twice before calibration. Then enter the numbers from the bottle on the calibration screen, and calibrate.
After calibration is complete, put the black tubing back into the tank where it was. Prime the sample line twice. Run a test, or wait till next test.