Confirming Salinity Between Tools - Hanna Salinity Tester

Slider162

Community Member
View Badges
Joined
Jan 11, 2020
Messages
29
Reaction score
46
Location
Madison, WI
Rating - 0%
0   0   0
I like the Hanna because of the convenience and ease of use. But right now, it has me questioning the salinity in my new tank using Instant Ocean.

I have:
  • A Salinity Refractometer (New)
  • A Calibrated Neptune Apex Probe (Old)
  • A Calibrated Hanna Salinity Tester (New)
I'm on my second Hanna Salinity Tester because the first was exchanged for the same issue. The replacement is identical to the first.

The refractometer and Neptune probe have always agreed with each other.

Right now:
  • Neptune = 34.2ppt
  • Refractometer = 1.025
  • Hanna = 29.8ppt
Then I read the Reef Aquarium Salinity: Homemade Calibration Standards and dissolved 6.2g of table salt with 182ml RODI. Everything matched. Neptune was at 35ppt, Hanna was ballpark 34.4ppt, and refractometer was just a hair above 1.025.

So why would the Hanna Salinity Tester read so low in aquarium water?
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Just an fyi, those devices will not perfectly match in a single diy fluid. There are different recipes for different types of devices. You are using all three different types.

That does not explain your issue with the Hanna, however. It also is not necessarily the Hanna that is off.

Try measuring the conductivity meter standard and the tank water in a cup at the same room temperature using the Hanna.
 

exnisstech

Grumpy old man
View Badges
Joined
Feb 11, 2019
Messages
18,920
Reaction score
30,365
Location
Ashland Ohio
Rating - 0%
0   0   0
I've had two Hanna salinity checker's and both would read 0.002 low when calibrated using Hanna calibration solution. They were compared to TM hydrometers and refractometer calibrated with Randy's solution. Mine would also drift so I scrapped them. From reading it seems its not an uncommon issue having the Hanna read low.
 
OP
OP
S

Slider162

Community Member
View Badges
Joined
Jan 11, 2020
Messages
29
Reaction score
46
Location
Madison, WI
Rating - 0%
0   0   0
Out of curiosity, I calibrated the Hanna with the home made solution and the results were not any better. Aquarium water is still 4-5ppt low on the Hanna.
 
OP
OP
S

Slider162

Community Member
View Badges
Joined
Jan 11, 2020
Messages
29
Reaction score
46
Location
Madison, WI
Rating - 0%
0   0   0
It hasn't been 30 days, so I'm asking to exchange for a Milwaukee. Hate to see someone starting out and only using the Hanna. I posted here because I needed a sanity check.
 

W31Olds

Valuable Member
View Badges
Joined
Dec 13, 2024
Messages
1,784
Reaction score
1,529
Location
Timonium
Rating - 0%
0   0   0
I would measure some newly mixed SW. Hanna is a conductivity meter so I would image it could be affected by particulates and metals in the Water.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Out of curiosity, I calibrated the Hanna with the home made solution and the results were not any better. Aquarium water is still 4-5ppt low on the Hanna.

Did it read the standard correctly after calibration? Was the temp the same?

I just do not understand how a conductivity meter can read two solutions with the same conductivity at the same temperature differently.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
I would measure some newly mixed SW. Hanna is a conductivity meter so I would image it could be affected by particulates and metals in the Water.

No, that is not the issue. Metals are conductive ions, but at very low concentration and nonconductive particulates are mostly ignored by conductivity meters unless there are so many that they occupy a significant fraction of the water volume (and they do not in new seawater, even if cloudy.
 
OP
OP
S

Slider162

Community Member
View Badges
Joined
Jan 11, 2020
Messages
29
Reaction score
46
Location
Madison, WI
Rating - 0%
0   0   0
I actually use the Hanna to mix my water. I just mix it at 30ppt and it shows up in my tank on the Neptune as 34-35ppt.

The standard reads 38.4 after calibrating in the home brew. Both have been sitting at room temp overnight since I started this thread.

After calibrating again with the standard and dipping in the home brew this morning is 31.9ppt. Tank now reads 28.0ppt.

Tank is 9.2F warmer than room temp solutions. But if we believe the marketing, the Hanna is supposed to be temperature compensated.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Well that is strange. Why would the 35 ppt standard not read 35 ppt after calibration with it?

I don’t know how the electronics of the Hanna calibrates. Does it let you tell it what the salinity is when calibrating?
 
OP
OP
S

Slider162

Community Member
View Badges
Joined
Jan 11, 2020
Messages
29
Reaction score
46
Location
Madison, WI
Rating - 0%
0   0   0
Maybe I misread the question. After calibrating with the standard, the standard tests at 34.4ppt.
 

NanoSteam

ANTI 100% Blues Only Club
View Badges
Joined
Dec 23, 2019
Messages
2,887
Reaction score
13,341
Location
SoCal
Rating - 0%
0   0   0
Did it read the standard correctly after calibration? Was the temp the same?

I just do not understand how a conductivity meter can read two solutions with the same conductivity at the same temperature differently.

Mine also had issues with temperature correction... It read anywhere from 2 to 4 degrees F low so I wonder if that's also causing some issues.

Two tanks with 3 different thermometers including a standard glass thermometer that match but Hannah would be 2 to 4 degrees lower with no explanation.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Maybe I misread the question. After calibrating with the standard, the standard tests at 34.4ppt.

Ok, still strange it not as much so. Do you tell it the standard is 35 ppt? It’s almost as if the standardization process did not actually change the meter.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Mine also had issues with temperature correction... It read anywhere from 2 to 4 degrees F low so I wonder if that's also causing some issues.

Two tanks with 3 different thermometers including a standard glass thermometer that match but Hannah would be 2 to 4 degrees lower with no explanation.

Maybe. I think it should not matter if it also reads the standard low, unless it electronically requires a specific temp for the standard.

The correction is a fixed change in conductivity vs degree F (or C).
 

NanoSteam

ANTI 100% Blues Only Club
View Badges
Joined
Dec 23, 2019
Messages
2,887
Reaction score
13,341
Location
SoCal
Rating - 0%
0   0   0
Maybe. I think it should not matter if it also reads the standard low, unless it electronically requires a specific temp for the standard.

The correction is a fixed change in conductivity vs degree F (or C).
Well in my case it didn't read the standard low... that was fine but putting it in the tank that's verified at 78F and it reads 74F is problematic I'd think. If I open a fresh solution packet it'll read 35ppt.

This is with tank salinity verified through calibrated refractometer that agrees with the TM glass hydrometer.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Well in my case it didn't read the standard low... that was fine but putting it in the tank that's verified at 78F and it reads 74F is problematic I'd think. If I open a fresh solution packet it'll read 35ppt.

This is with tank salinity verified through calibrated refractometer that agrees with the TM glass hydrometer.

Temperature correction for a conductivity meter should only be an issue if it sometimes corrects and sometimes not (like a reading vs standardization). If it just always read temps 4 degrees too low, that would not cause an issue. The temperature correction of the standard and the sample is the same.

But it is prudent to have the standard and tank sample at the same temp and close to 25 deg C if you think there may be a temp correction issue.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Why not use the TM hydrometer as the reference and calibrate to a known dependable instrument that isn’t influenced by calibration fluids of unknown accuracy. J

You assume it is accurate. I would not assume it is accurate without testing it against a fluid of known density. There are high quality seawater standards used by labs. Reefers just like to cheap out.

I’d have a lot more confidence in a standard I made using a calibrated scale than an inexpensive and untested hydrometer in which the paper may have slipped, the guy making it got a text at the wrong moment, etc.

Once validated with a test, I agree that is a good way to check other equipment.
 

TOP 10 Trending Threads

WHAT AMOUNT OF LIVE ROCK AND SAND SHOULD BE PRIORITIZED FOR OPTIMAL BIODIVERSITY/FILTRATION?

  • 100% live rock + bagged sand

    Votes: 34 28.1%
  • 100% dry rock + 100% live sand

    Votes: 41 33.9%
  • 50/50 live/dry rock, 50/50 live/bagged sand

    Votes: 27 22.3%
  • 75% live rock, 25% live sand

    Votes: 11 9.1%
  • 25% live rock, 75% live sand

    Votes: 8 6.6%
Back
Top