I've been sending out some ICP's and thought it'd be worth doing some manual tests at the same time to compare the accuracy. There's a lot that goes into the accuracy of both testing measures but I am at least consistent w my manual tests and try to do things as accurate as possible. These are the results of my most latest (5/8 sample, 5/19 results) ICP/manuals at the same time, but all prior tests are pretty much the same which speaks to my personal methods in manual testing (or so I like to think lol)
Here's what I found-
I'm pretty happy with everything overall. Particularly happy with the TM Hydrometer :)
Anyways, I thought this was cool information and will continue to track it.
Here's what I found-
| Param | Manual Result | ICP Result | Manual Brand/Type |
| NO3 | 25.2 | 28.33 | Hanna / Oceamo-MS |
| PO4 | 0.17 | 0.124 | Hanna / Oceamo-MS |
| Alk #1 | 8.23 | 8.22 | Hanna / Oceamo-MS |
| Alk #2 | 8.40 | 8.22 | Alkatronic / Oceamo-MS |
| Alk #3 | 8.42 | 8.22 | Trident / Oceamo-MS |
| Ca #1 | 450 | 409 | API tritration / Oceamo-MS |
| Ca #2 | 421 | 409 | Trident / Oceamo-MS |
| Sg #1 | 1.0261 / 34.64 | 34.4 | Tropic Marin Hydrometer / Oceamo-MS |
| Sg #2 | 32.1 | 34.4 | Hanna Tester / Oceamo-MS |
I'm pretty happy with everything overall. Particularly happy with the TM Hydrometer :)
Anyways, I thought this was cool information and will continue to track it.