Sample your Waste Water to ensure the unit is plumbed correctly and you're not pumping Waste Water through the RO Membrane and Product Water down the drain.
If the numbers tell you it's plumbed correctly, next release the system pressure and take the Membrane housing apart.
Remove the Membrane and inspect the o-rings for damage. You want to ensure Waste Water is not bypassing the Membrane.
Re-install the Membrane and push it down hard into the housing to ensure it's properly seated, then re-install the cap.
Re-start the unit WITHOUT the DI and flush the system for 15 mins or so. Hopefully you have a flush valve on the unit.
After flushing, run the unit for 5 mins or so, then take a TDS reading - remember this is without the DI included.
With your source water of 137, and a 98% rejection rate, you should be getting close to less than 3 TDS from the Membrane.
Post your results after testing