Trident Only testing with a manual test

Johny_Dough

Community Member
View Badges
Joined
Sep 12, 2021
Messages
60
Reaction score
29
Location
Los Angeles
Rating - 0%
0   0   0
Trident stopped auto testing. I have confirmed the Mode is set to automatic.

No other error it just says "no test in 24 hours" then I can manual select Combined test and it runs. Looking to see if there is an easy solution before I contact support.
 

nothing_fancy

Active Member
View Badges
Joined
Oct 6, 2022
Messages
435
Reaction score
367
Location
usa
Rating - 0%
0   0   0
If you haven't trident initializing it, or doing a hard reset, basically unplug it for a few minutes, contact support. I hit them for an issue I had with my trident once and they were really attentive to it, pretty quick response.
 
OP
OP
Johny_Dough

Johny_Dough

Community Member
View Badges
Joined
Sep 12, 2021
Messages
60
Reaction score
29
Location
Los Angeles
Rating - 0%
0   0   0
If you haven't trident initializing it, or doing a hard reset, basically unplug it for a few minutes, contact support. I hit them for an issue I had with my trident once and they were really attentive to it, pretty quick response.
I did a restart...but not a unplug from wall and wait 3 minutes...I guess I'll try that...
 

areefer01

5000 Club Member
View Badges
Joined
Jun 28, 2021
Messages
5,718
Reaction score
5,933
Location
Ca
Rating - 0%
0   0   0
You mentioned it stopped auto testing. Is this a new install or existing? Was it running tests prior?

I would contact support regardless using the email option/ticket. This way it starts the ball rolling.

1681942435919.png
 

Attachments

  • 1681942458533.png
    1681942458533.png
    144.7 KB · Views: 74

Matt L

10K Club member
View Badges
Joined
May 15, 2021
Messages
10,746
Reaction score
41,835
Location
US
Rating - 0%
0   0   0
Check your tests remaining. The Trident will stop testing when it thinks you are out of reagent. Check all your bottles, and tap on the trident graph in fusion until you see tests remaining. If it says close to zero, reset that reagent if there is some left in the bottle, or change the bottle then reset. The Trident will not give an alarm when the reagent is empty until you program your "email" alarm to.
 
OP
OP
Johny_Dough

Johny_Dough

Community Member
View Badges
Joined
Sep 12, 2021
Messages
60
Reaction score
29
Location
Los Angeles
Rating - 0%
0   0   0
It's been running over a year now. Regants are 89 percent full.

The Apex head unit did just get swapped out with a refurbished unit, Just about the time this started. It also coincide with the recent OS update.
 

Roloilly

Community Member
View Badges
Joined
Jun 13, 2022
Messages
45
Reaction score
7
Location
Miami
Rating - 0%
0   0   0
Need help. I reprogrammed the trident for one combined test and two Alk per day. It's been working flawlessly for the past few weeks. The last 3 days it's been doing a a combine test back to back within 15 minutes apart. The setting I'm using for combine is

Set OFF
If Time 08:00 to 08:01 Then ON

Not sure why.its doing this. Any help would be appreciated.
 

TOP 10 Trending Threads

HOW DO YOU ADJUST YOUR CUC AS ALGAE DISAPPEARS?

  • Capture and re-home CUC

    Votes: 8 7.1%
  • Increase white light/hours in tank to spur algae growth to feed CUC

    Votes: 7 6.2%
  • Feed nori to support CUC

    Votes: 39 34.5%
  • Feed herbivore pellets to support CUC

    Votes: 40 35.4%
  • Allow attrition to balance CUC and algae

    Votes: 48 42.5%
  • Provide macro algae to feed CUC

    Votes: 8 7.1%
  • Introduce CUC predators

    Votes: 1 0.9%
  • Other (please explain)

    Votes: 11 9.7%
Back
Top