AN ACCURATE AND PRECISE METHOD FOR MEASURING IODINE

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
ABOUT THE ARTICLE:

The article describes an accurate and precise method and it's development, for measuring the level of Iodine in a saltwater aquarium. It is a follow-up to an article written in 2021. The current article describes an improved method but using the same basic ideas as the 2021 method. The details of the method development are included for those who might be interested. It is an informative example on how methods development can be challenging. That being said if you are just interested in the "Bottom Line" you can skip to Section VII "THE PROCEDURE". This is where the detailed procedure is described. There is also a video link there to see the procedure preformed. I hope this is helpful and you enjoy the reading.

Rick



1674675953336.png
I. INTRODUCTION

In December of 2019 I posted an article “USING THE HANNA CHECKER HI-707 TO TEST FOR IODINE”… You can read it here:


https://www.reef2reef.com/ams/using-hanna-checker-hi-707-to-test-for-iodine.743/

It might be helpful to read it because it has some additional background information.

In using this test method I found, although it worked, it was not satisfactory for the lower levels of Iodine (< .02-.03 ppm). The procedure put the measurement results up near the upper end of the meters measurement range (600ppm.) In addition while working on a separate project @taricha found the measured values to be off more than the test protocol predicted. This would have been the first hint that there is something protocol that was not solid.


This started us (@Dan_P, @taricha and me) on a quest to modify the existing procedure, which I thought would be a “piece of cake”. Little did we know that this “piece of cake” would take us down a rabbit hole that has taken many interesting turns and detours!!

I worked on the initial efforts by working with the reaction time and the levels of the Red Sea Reagent. This helped us to determine the time and reagent amounts that would move measurement range closer to the center of the range. Table 1 and Chart 1 show the results. (See REAGENT AMOUNTS SECTION for details). What follows are the account and some details of the procedure development and information necessary to do this test yourself.

1674679657017.png

TABLE 1



1674679719485.png


CHART 1
If you wish to join us in a trip down the rabbit hole, keep on reading. If not you can skip to the end and find a detailed step by step procedure to run the test for Iodine…Here we go!

1674679875472.png


II. INITIAL PROCEDURE DEVELOPMENT

In this section we will describe the steps in the initial procedure development. These will include:


  • Analysis of the variables of the Red Sea reaction
  • How the standards used in the test development were prepared
  • Sample measurement and data collection
  • Regression generation
  • Procedure replication and Validation
We will not go into great detail on this, but will cover some of the key work.


THE RED SEA REACTION VARIABLES

TIME:

The Red Sea reaction is very time dependent. In order to get accurate and precise results close attention to the reaction time is important. Chart 2, provided by @taricha shows the time dependence of the reaction at various ppm levels of iodine.

1674680179745.png
CHART 2

Here is a link to a time laps video showing the reaction over time. It plots Absorbance over time. It also shows a visual of the reaction over time.




TEMPERATURE

The reaction is temperature dependent as can be seen from Chart 3. We will cover this in a bit more detail in section on MODIFIED PROCEDURE DEVELOPMENT. The outcome of this temperature dependency is a need to know the temperature of the sample being tested.

1674680664367.png

CHART 3

Here is a link to a video showing the reaction at different temperature. I plots change in Absorbance over time at 3 different temperatures. You can also see a visual of the reaction overt time compared to the initial color of the sample



SALINITY

There is a very slight salinity effect but the impact on the measurement and the calculation is minimal and I did not include it in the calculation. Chart 4 shows the salinity effect. The differences in the measurement due the salinity effect were small enough that there is no adjustment made in the regression.

1674681294288.png
CHART 4

REAGENT AMOUNTS

The Red Sea reaction requires 2 reagents. When both reagents are added an orange color develops. The intensity of this orange color is dependent on the amounts of the reagents used. By reducing the amounts to 75% (3/4) their initial levels with a 4 minute reaction time I was able to obtain a more reasonable measurement range. This can be seen in Chart 5 as well as Chart 1 in the introduction. The orange trend line has a much broader range of meter measurements than the other trend lines which represents different reaction times and reagent levels.

1674681452673.png

CHART 5
REAGENT VARIABILITY

This was not evaluated in the initial Procedure Development. We assumed this would not be an issue, which turned out to be a very bad assumption. (See PROCEDURE REPLICATION section)

STANDARDS PREPARATION

Three sets of standards were prepared. Two we made with potassium iodide (KI) and one was made with potassium iodate KIO3. Four levels were prepared: 0 ppm, .02 ppm, .04 ppm, .06 ppm and .1 ppm. They were prepared according to the procedure described in the December 2019 article. https://www.reef2reef.com/ams/using-hanna-checker-hi-707-to-test-for-iodine.743/



SAMPLE MEASUREMENT (These were the prepared Standards)

The samples were measured using the Hanna Checker HI-707 according to the following procedure:

  • Sample temperatures were measured.
  • The adjusted levels of reagent A (.38mL) and B (.6mL) were added to 10 mL sample
  • Measurement was taken at the end of a 4 minute reaction time
  • Each level of the standards was measured 4 times then an additional 4 time on second day (Total 8 measurements at each level)
The potassium iodide standards were measured a second time on a different days with a total of 8 measurements at each level. The potassium iodate was also measured 4 times. It was only measured 1 time with a total of 4 samples at each level. The potassium iodate sample set was used to validate the iodide standards,

DATA COLLECTION

The measurement data was collected from the measurement sets and recorded. See Table 2 for the results. Only the results of the potassium Iodide are given.

1674681896100.png

TABLE 2

REGRESSION GENERATION

Using the data from Table 2 and the Excel Data Analysis Regression function, the regression function was calculated. Chart 6 is a plot of 3 regressions. The red is the new procedure using a 4 minute reaction time with 3/4 reagent level (Data from Table 2). The green is the old regression and the blue is the new procedure but using potassium iodate (Red Sea STD) as the standards. The red and blue regressions are generated from additional data sets not included here.

It can be seen from these results that there is good agreement between the two different standards ….Potassium Iodide and Potassium Iodate.

1674682078202.png
CHART 6

The regression equation selected was: (Includes Temperature Correction)


Y = (-0.0003339038*(((TEMP-79)*3.5) +X)) +.1452396

This was generated using the program MyCurve Fit with the same data set.




These are the results that we were attempting to replicate in the PROCEDURE REPLICATION/ VALIDATION step. This is where things went off the rail and opened up a whole new rabbit hole!


III. PROCEDURE REPLICATION/VALIDATION


As is my usual practice I presented my findings along with the procedure to @Dan_P and @taricha for review and replication….This was the entrance to quite a deep rabbit hole!

Although it was expected that “tweaking” might be necessary, it was not expected to run into so many difficulty replication my results. It would have been acceptable to have a 5 or even 10% deviation from my results, but there was a 25-35% deviation….Not acceptable. The puzzling thing was it was very consistent. It was in the same direction and the same magnitude every time. This indicated a systematic error…But where? This lead us to look at everything from standards preparation, step by step procedure validation, reagent variability, time and temperature effects, instrument validation and a host of others. We exchanged samples, standards and procedural notes and videos all searching find where the variability came from.

To shorten the story it ended up that there were 2 culprits…Reagent variability (See Chart 7) and temperature variations (See Chart 3 & 8). The temperature variation was manageable. The previous procedure had a temperature correction component so we expected this modified procedure would need one also. However, the reagent variability presented a big hurdle because this was outside of our control. The variation is large enough that without a solution to the problem the test would not be useful. One solution was to “test” the regression every time there was a new lot, and if it was not acceptable generate a new regression using a 2 point calibration. This is really not practical and makes the tests much more cumbersome….Seemed as though we were stuck...Ah…The story continues... @taricha to the rescue!!


1674682733046.png
CHART 7

1674682899844.png

CHART 8
The solution to the reagent variability problem came while @taricha was pondering the reaction as a function of time and noted that rate at which the reaction proceeded (loss of color—color destruction) from one set time to another was proportional to the iodine concentration. The logic can be seen in Charts 9-11

1674683185628.png

CHART 9
SPECTRAL ABSORBANCE OF 3 DIFERENT LOTS OF REAGENT

Chart 9 illustrates lot to lot variability. Looking at the absorbance value of each lot at the 240 seconds at 3 different iodine concentration levels shows the lot variation to be so large that the test is not useable.

1674683431614.png

CHART 10
REACTION TIME SLOPES

@taricha looked at the Time Slopes (Abs units/sec) (Chart 10) the results tightened up but not tight enough to yield solid testing results.


The breakthrough came when he realized the reaction in each lot was “destroying” close to the same proportion of the initial maximum color (0 ppm) per second. He took the initial measurement subtracted the end point measurement and divided this by the initial measurement. This produced a ratio representing the portion of the color “destroyed” (loss of color), which we later designated “Fraction of Color Destroyed” FCD. Chart 11 shows the results, almost perfect agreement...we could now proceed!

1674683646829.png

CHART 11

When we used the FCD the two reagent lots were almost identical in their prediction of the level of iodine in the sample. This opened the door for a meaningful method to modify the Initial Procedure Developed and get excellent replication results.

During this stage we modified the time segment to an end point of 3 minutes (180 Sec.). This shortened the testing time and moved the end point away from the low meter reading at high iodine concentrations. Chart 12 illustrates this.

1674683889668.png

CHART 12

IV. MODIFIED PROCEDURE DEVELOPMENT

This section will cover; Sample Measurement, Data Collection and Regression Generation. The findings we uncovered during our trip down the rabbit hole of “Procedure Replication” will be applied to the sample measurement. The new independent variable will now be the “Fraction of Color Destroyed” (FCD). The total reaction time will be reduced to 3 minutes. All the other variables will remain the same. See comparative descriptions below to get a better picture.


1674684422045.png
1674684478571.png
NOTE: This is not a detailed step by step of the procedure. That can be found in section VII

This modified procedure was used to generate the information that follows.




STANDARDS PREPARATION

The standards were prepared using reagent grade Potassium Iodide at the following levels in PPM. 0, .03, .06, .09, .12. This was changed from the initial procedure development to better handle data for procedure replication. Again more details can be found in the December 2019 article here: . https://www.reef2reef.com/ams/using-hanna-checker-hi-707-to-test-for-iodine.743/

SAMPLE MEASUREMENT (These were the prepared Standards)

The samples were measured using the Hanna Checker HI-707 according to the following procedure:


  • Sample temperatures were measured.
  • The adjusted levels of reagent A (.38mL) and B (.6mL) were added to 10 mL sample
  • Timer set for 3 minutes is started
  • Measurement was taken at the end of a 1 minute reaction time and the data was recorded. This would be measurement 1-- M1
  • The checker was re-zeroed
  • The test sample is placed back into the Checker
  • At the end of the 3 min time a second measurement is made---M2
  • Each level of the standards was measured 4 times.
  • This procedure was repeated a second time on a different day for a total of 8 measurements
  • CALCULATION OF FCD RATIO ---FDC = (M1-M2)/M1


DATA COLLECTION


The data Table 3 below is a tabulation of the data generated from the sample measurements.

1674684621608.png

TABLE 3
REGRESSION GENERATION

Using the data from Table 3 and the Excel Data Analysis Regression function, the regression function was calculated. Chart # 13shows the Excel Regression output.
1674684783035.png
CHART 13

The selected regression equation was generated from the combined data sets of 11/2 and 11/15. The data from the 11/2 by itself sample set was also included to see the repeatability of the procedure. As you can see the agreement is quite good.

The resulting regression is Y= .2544 X -.0005 (No temperature correction applied)

Y = the PPM Iodine in the sample


X = the FDC Ratio

The next section shows the step by step calculations.

VII. STEP BY STEP PROCEDURE


EQUIPMENT REQUIRED

1) Hanna HI-707 Nitrite LR Checker

2) Red Sea Iodine Pro Reagents A & B

3) Two Hanna Checker Vials (These should be matched vials) (1)

4) Timer

5) Accurate Pipette(s) or Syringe capable of measuring .38 mL and .6 mL

6) RODI Water for rinsing vials & pipettes

7) Accurate Thermometer

8) Sample Water Container (I use a beaker or plastic bottle)



The procedure itself may appear to be a little complicated, but the reality is there are only 5 components to the test and takes only about 7-9 min to do.

1674684918409.png


PROCEDURE STEP BY STEP (See Video Link below)


  • Fill vial to be used as the Blank with water to be tested.
  • Place the blank into the checker
  • Rinse second vial with water to be tested
  • Fill second vial with water to be tested to the 10 mL line
  • Measure and record sample temperature
  • With the blank in the checker press the button on the Hanna Checker…When “C1” appears press again….”C2” will now be on the display.
  • Remove the Blank
  • Set Timer to 3 Minutes
  • Add .38 mL of Red Sea reagent “A” to second vial containing the test water.
  • Add .6 mL of Red Sea reagent “B’ to the second test vial
  • IMMEDIATELY START TIMER!!
  • Invert the vial 5-10 times to mix reagents
  • Immediately place the test vial into the checker
  • When the timer reaches 2 min (1 min has passed) press the button on the checker and record the Measurement. This will be your M1 value
  • Press the checker button to turn off. Remove the test vial and replace it with the Blank.
  • Press the checker button to turn it back on. “C1”. Press a second time to re-measure blank.
  • Remove the Blank and replace it with the test vial. Wait for time to reach “0”
  • When the timer reaches “0” IMMEDIATELY PRESS THE BUTTON ON THE CHECKER TO GET THE MEASUREMENT…This will be your M2 value
  • Calculate Fraction of Color Destroyed (FCD)
FCD (Fraction of Color Destroyed) = (M1-M2)/M1

Calculate the Iodine PPM using Measured Temp and FCD and regression equation.


PPM IODINE = ((0.2544-(0.0061*(Temp-79)))*FDC)-0.0005

REGRESSION EQUATION


HERE IS THE LINK TO THE VIDEO OF THE PROCEDURE ---

VIII. CONCLUSION

Well that about covers it. I think having an accurate and repeatable test for Iodine is an important tool to have for Reef Keeping. There are a number of us who dose iodine. Having a method to measure is very important. This test should fit the bill. Thank you for taking the time to read this article…Your comments are welcomed.

This project was a collaborative effort with @Dan_P and @taricha. Their efforts were key in bringing this project to a successful conclusion. There were times when we were at an impasse and it appeared we not going to be successful, but their innovative spirit and tenacity pulled us through. We are most fortunate to have them as contributors to our R2R family
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Nice job Rick. Your article covers a ton of work which most people wouldn’t think of just giving it away. Thank you.

Which element is next?

Dan
Thank you Dan...The reality is without your and @taricha support I would still be in the Rabbit Hole
 

radiata

Valuable Member
View Badges
Joined
Nov 2, 2015
Messages
1,089
Reaction score
764
Rating - 0%
0   0   0
Great piece of work, especially when RHF says so! Thank you.
 

GBRsouth

Valuable Member
View Badges
Joined
Jul 9, 2017
Messages
1,279
Reaction score
3,572
Location
N.S.W. AUSTRALIA
Rating - 0%
0   0   0
Thank you for the time and effort you have put into this. It is so much easier to read a result as a digital number for those of us with a range of eye conditions. I will certainly make us of your efforts here in the future.
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Thank you for the time and effort you have put into this. It is so much easier to read a result as a digital number for those of us with a range of eye conditions. I will certainly make us of your efforts here in the future.
Totally my pleasure...let me know how it goes.....I also want to point out that @Dan_P and @taricha put a lot of effort into this work...
 
Last edited:

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,535
Reaction score
10,082
Rating - 0%
0   0   0
@Rick Mathew thanks for putting the time in wrestling with this one.
This is an element that is in low but measureable concentrations in natural sea water, is quickly depleted in our tanks, and with blind dosing according to a schedule can get too high.
It's also one that depletes somewhat in shipping tubes to ICP and ICP vendors generally have a tough time agreeing on what the values are (I and P spectra interfere).
So it's one that really needed a good chemical test.
Nicely done!
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
@Rick Mathew thanks for putting the time in wrestling with this one.
This is an element that is in low but measureable concentrations in natural sea water, is quickly depleted in our tanks, and with blind dosing according to a schedule can get too high.
It's also one that depletes somewhat in shipping tubes to ICP and ICP vendors generally have a tough time agreeing on what the values are (I and P spectra interfere).
So it's one that really needed a good chemical test.
Nicely done!
Thank you...It was your contributions and efforts that pushed this project over the goal...many thanks to you @taricha
 

MnFish1

10K Club member
View Badges
Joined
Dec 28, 2016
Messages
22,804
Reaction score
21,929
Rating - 100%
1   0   0
Nice article
 

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,535
Reaction score
10,082
Rating - 0%
0   0   0
Wanted to show off using this in my tank.
My system runs essentially zero Iodine when I don't add it.

I measured baseline tank water, and then added Red sea Trace A (Iodine) that should raise Iodine to 0.060ppm Measured again 30 min later and again 4hr later.

Hours60s180sFCDtempcalc ppm
0.03843810.007868.20.002
0.83342640.209668.40.066
4.33402810.173568.40.055
This was using the recommended checker, the hi707.


...but you can use one of the hanna checkers at 525nm green LED, like the PO4 checkers and you get almost the same thing. Because it works by measureing the fraction of color removed, so the actual values of the readings don't matter.
So here's same water samples tested with a different checker with a green LED.
60s180sFCDtempcalc ppm
117900.230768.50.073
1251030.176068.50.056


So here's what it looks lijke comparing the two. Almost the same regardless of which checker and whether the LED is blue or Green.
Screen Shot 2023-02-03 at 10.41.01 PM.png
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
Wanted to show off using this in my tank.
My system runs essentially zero Iodine when I don't add it.

I measured baseline tank water, and then added Red sea Trace A (Iodine) that should raise Iodine to 0.060ppm Measured again 30 min later and again 4hr later.

Hours60s180sFCDtempcalc ppm
0.03843810.007868.20.002
0.83342640.209668.40.066
4.33402810.173568.40.055
This was using the recommended checker, the hi707.


...but you can use one of the hanna checkers at 525nm green LED, like the PO4 checkers and you get almost the same thing. Because it works by measureing the fraction of color removed, so the actual values of the readings don't matter.
So here's same water samples tested with a different checker with a green LED.
60s180sFCDtempcalc ppm
117900.230768.50.073
1251030.176068.50.056


So here's what it looks lijke comparing the two. Almost the same regardless of which checker and whether the LED is blue or Green.
Screen Shot 2023-02-03 at 10.41.01 PM.png
AWESOME...thanks for sharing!
 

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,535
Reaction score
10,082
Rating - 0%
0   0   0
Wanted to show off using this in my tank.
My system runs essentially zero Iodine when I don't add it.

I measured baseline tank water, and then added Red sea Trace A (Iodine) that should raise Iodine to 0.060ppm
two more days of measurements...
with the preferred checker (hi707)
Daysat 60sat 180sFraction Color Destroyed
(FCD)
Tempppm
1.233863580.072565.10.024
2.173983820.040265.10.013

and with a more common green LED checker (hi764)
Daysat 60sat 180sFraction Color Destroyed
(FCD)
Tempppm
1.231231150.065066.20.021
2.171341280.044866.20.014

And so now you get a good look at the depletion rate of the 0.06ppm Iodine in my tank.

Screen Shot 2023-02-05 at 9.42.16 PM.png
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
67,160
Reaction score
63,520
Location
Arlington, Massachusetts, United States
Rating - 0%
0   0   0

And so now you get a good look at the depletion rate of the 0.06ppm Iodine in my tank.

Screen Shot 2023-02-05 at 9.42.16 PM.png

Excellent. Thanks for posting that.
 
OP
OP
Rick Mathew

Rick Mathew

Valuable Member
View Badges
Joined
Mar 21, 2016
Messages
1,472
Reaction score
4,736
Location
North Central Florida
Rating - 0%
0   0   0
two more days of measurements...
with the preferred checker (hi707)
Daysat 60sat 180sFraction Color Destroyed
(FCD)
Tempppm
1.233863580.072565.10.024
2.173983820.040265.10.013

and with a more common green LED checker (hi764)
Daysat 60sat 180sFraction Color Destroyed
(FCD)
Tempppm
1.231231150.065066.20.021
2.171341280.044866.20.014

And so now you get a good look at the depletion rate of the 0.06ppm Iodine in my tank.

Screen Shot 2023-02-05 at 9.42.16 PM.png
Very informative...It got me to thinking...which can be dangerous :flushed-face: ...I had a iodine doser malfunction some time ago and pushed the iodine up to >.1 ppm. There were no visible damage to my critters but I turned off the dosing. I continued to test parodically over 30 days for...Most of which were tested with the new procedure...so I plotted the results below

1675698318297.png

essentially the same shape but different rate as your system
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
67,160
Reaction score
63,520
Location
Arlington, Massachusetts, United States
Rating - 0%
0   0   0
Mine was more like taricha's. A few days from 0.06 ppm to undetectable by Salifert.

With no dosing at all, ICP by Triton said 0.022 ppm, but I don't know what form.
 

taricha

5000 Club Member
View Badges
Joined
May 22, 2016
Messages
6,535
Reaction score
10,082
Rating - 0%
0   0   0
There were no visible damage to my critters but I turned off the dosing. I continued to test parodically over 30 days for...Most of which were tested with the new procedure...so I plotted the results below
Some recent results show Iodine well above target levels in a couple of salt mixes (which is fine - it depletes)
I.O. and Reef Crystals checked in at 0.20ppm Iodine here.
https://www.reef2reef.com/threads/the-ultimate-salt-test.949805/post-10794608

(Also my I.O. made fresh is above 0.10 ppm Iodine too, so the results in the thread are consistent with what I've seen as well.)

Did you ever see a bump up just from a water change, and were there many water changes during this depletion curve?
1675698318297.png

essentially the same shape but different rate as your system
 