Hanna Marine Magnesium Checker HI783 Problems/Issues

Users Who Are Viewing This Thread (Total: 1, Members: 0, Guests: 1)

Are you experiencing issues with your Marine Magnesium Checker (HI783)?

  • Yes - Describe Below

    Votes: 31 88.6%
  • No

    Votes: 4 11.4%

  • Total voters
    35

trevorhiller

Reefing Nurse
View Badges
Joined
Oct 22, 2021
Messages
1,052
Reaction score
752
Location
Williamsport
Rating - 100%
1   0   0
There are several threads accumulating on the forums of people having issues w/new marine magnesium checker HI783. There are also some similar reports on the AskBRS facebook page--results several hundred points higher. I wanted to start a thread here in Hanna Instruments sub-forum to give Hanna a chance to see the reports of similar issues and address. I personally am despite about a dozen repeated tests using both factory instructions and varied testing procedures to ensure accuracy.


If you are having/have had issues w/this new device please post here.
 

sublime275

Community Member
View Badges
Joined
Jul 20, 2013
Messages
89
Reaction score
59
Location
Dallas, TX
Rating - 0%
0   0   0
I’ve posted in one of those threads how the new tester gave me inaccurate (high) results using the Fauna Marin reference solution for the sample water. The magnesium level that was printed on the reference solution was confirmed by my Salifert and Aquaforest magnesium test kits. I also experienced widely varying results on the same sample. I’ve used almost all of the reagents trying to get somewhat accurate and consistent results.
 

Coolcasino

Well-Known Member
View Badges
Joined
Apr 29, 2020
Messages
713
Reaction score
348
Rating - 0%
0   0   0
I just bought mine at Sea Creature Aquarium in North Providence. First test 1650, second test 1680 3rd test 1690. Tested 3 times with salifert and got 1460,1450 and 1450. Used red sea and got 1450, 1450, 1440. My trident is reading 1458. Ive tried all the tricks and tips. I have your Alk, Phos, Nitrate checkers so I know how to use it. Also have your salinity checker. The syringes are barley tight on all 3. The airline tubing you have in the tips is to wide and the tips pop off when you push the liquids out. I drew the liquids pushing the tips tightly into the syringe. No liquid was leaking. Did the same when I was putting it in the curvette. I dont think this checker is up to the standards that im use to from a Hanna product.
 

Sean Clark

7500 Club Member
View Badges
Joined
May 16, 2019
Messages
8,055
Reaction score
31,577
Location
Michigan
Rating - 0%
0   0   0
I just bought mine at Sea Creature Aquarium in North Providence. First test 1650, second test 1680 3rd test 1690. Tested 3 times with salifert and got 1460,1450 and 1450. Used red sea and got 1450, 1450, 1440. My trident is reading 1458. Ive tried all the tricks and tips. I have your Alk, Phos, Nitrate checkers so I know how to use it. Also have your salinity checker. The syringes are barley tight on all 3. The airline tubing you have in the tips is to wide and the tips pop off when you push the liquids out. I drew the liquids pushing the tips tightly into the syringe. No liquid was leaking. Did the same when I was putting it in the curvette. I dont think this checker is up to the standards that im use to from a Hanna product.
Well that is the tightest that I have seen... off the mark by 200ish but consistent.
 

Coolcasino

Well-Known Member
View Badges
Joined
Apr 29, 2020
Messages
713
Reaction score
348
Rating - 0%
0   0   0
Well. Looks like it just keeps testing higher.
 

Attachments

  • 20220603_211839_HDR.jpg
    20220603_211839_HDR.jpg
    61.5 KB · Views: 63

piranhaman00

2500 Club Member
View Badges
Joined
Jun 24, 2019
Messages
4,875
Reaction score
4,828
Rating - 0%
0   0   0
I don’t understand the point of this checker. It’s error is 5% meaning at 1600ppm we are taking 1520 to 1680 ppm, if you only care to be that accurate run a simple salifert test
 
OP
OP
trevorhiller

trevorhiller

Reefing Nurse
View Badges
Joined
Oct 22, 2021
Messages
1,052
Reaction score
752
Location
Williamsport
Rating - 100%
1   0   0
I don’t understand the point of this checker. It’s error is 5% meaning at 1600ppm we are taking 1520 to 1680 ppm, if you only care to be that accurate run a simple salifert test
This is true. But that’s also assuming you’re off by 5% in opposite directions on each test. Ideally that’s not the case and while you might be +/- up to 5% of the actual value, one would hope that the tester is consistently off by +/-x% in the same direction-leaving your trends atleast valid if testing procedure is consistent. However, that’s doesn’t seem to be the case with this particular checker as results are all over the place on the higher end.

Their Alk checker is also +/-5% or 0.3 dKH. So an actual of 8.0 could be 7.7-8.3. But in my experience, I get way more consistent results with the Alk checker.
 

Viet658

Community Member
View Badges
Joined
Mar 31, 2017
Messages
85
Reaction score
96
Rating - 0%
0   0   0
The Hanna isn’t giving me consistent results. Results also appear to be high compared to my aquaforest test kit. First test was flashing 1800 which means results are out of range. Second test indicated 1695. Aqua forest testing at 1470.
 

Brewtoszek

Community Member
View Badges
Joined
Sep 3, 2020
Messages
61
Reaction score
85
Location
Bethlehem
Rating - 0%
0   0   0
Having the same problems, first test on the tank was out of range(high)second 1780. Tested freshly mixed reef crystals to 35ppt and got 1750, box says it should be 1345 on average.
 

Coolcasino

Well-Known Member
View Badges
Joined
Apr 29, 2020
Messages
713
Reaction score
348
Rating - 0%
0   0   0
I just mixed regular old IO and tested it 5 times and got 1690, 1680, 1650, 1700, 1690. Tested with Red Sea and got 1360 tested with Salifert and got 1350.
 

yanetterer

Active Member
View Badges
Joined
Jul 26, 2019
Messages
251
Reaction score
326
Location
Atlantis
Rating - 0%
0   0   0
I've used mine about 10 times now and all but two results have been 1800+. The two times I've been able to get a reading were the first time I used it and when testing a freshly mixed batch of Reef Crystals. Both of those tests were in the 1600s which is higher than the ~1300 that my tank is normally at and that Reef Crystals normally mixes to. An ATI sample was at 1320, Salifert was in the high 1300s to low 1400s, and Red Sea was in the 1400s to 1500s. I tried ensuring that I am adding everything slowly to minimize bubbles, rinsed thoroughly with RODI and let dry, but still can't get the thing to stop reading 1800.
 

Coolcasino

Well-Known Member
View Badges
Joined
Apr 29, 2020
Messages
713
Reaction score
348
Rating - 0%
0   0   0
What's the point of the plastic tips? We aren't doing titration like in other tests.
 

Coolcasino

Well-Known Member
View Badges
Joined
Apr 29, 2020
Messages
713
Reaction score
348
Rating - 0%
0   0   0
I presume it’s to make it easier to get the reagents out of the bottles.
I think this causes more harm than good. Its hard to get a good seal on these. Im going to try to run a test without the tips.
 

jrill

Valuable Member
View Badges
Joined
Jun 21, 2018
Messages
1,114
Reaction score
1,191
Location
Crossville, TN
Rating - 0%
0   0   0
Not a happy camper. Aqua Forest tested at 1420. Hanna is at 1650. Even the Aqua Forest calibration fluid is reading in the 1600 range With hanna when it should be 1310.
 
OP
OP
trevorhiller

trevorhiller

Reefing Nurse
View Badges
Joined
Oct 22, 2021
Messages
1,052
Reaction score
752
Location
Williamsport
Rating - 100%
1   0   0
Not a happy camper. Aqua Forest tested at 1420. Hanna is at 1650. Even the Aqua Forest calibration fluid is reading in the 1600 range With hanna when it should be 1310.
I replied to the email I received back from Hanna requesting them to send me a return number to send the tester in. I also mentioned in my email that multiple people are posting similar issues to mine and included a link to this thread in my email. Well see if they acknowledge that part of my email.

I was hoping to tag Hanna instruments in this thread for a solution to this problem, but their name is “Hanna Instruments” on here and won’t come up when you type the @ symbol to tag them.
 

Sean Clark

7500 Club Member
View Badges
Joined
May 16, 2019
Messages
8,055
Reaction score
31,577
Location
Michigan
Rating - 0%
0   0   0
I replied to the email I received back from Hanna requesting them to send me a return number to send the tester in. I also mentioned in my email that multiple people are posting similar issues to mine and included a link to this thread in my email. Well see if they acknowledge that part of my email.

I was hoping to tag Hanna instruments in this thread for a solution to this problem, but their name is “Hanna Instruments” on here and won’t come up when you type the @ symbol to tag them.
This entire thread is under the Hanna Instruments sponsor thread. They should be seeing all of the posts. It is odd that the @ is not working though.
 