@Dogtown - I can help a little on the Tritation takes too long, as I was having the same error message
@Vinny@GHLUSA and
@Matthias Gross helped me fix the issue.
You need the latest version(Version 1..1.1.6) of GCC and you do not need to upgrade the firmware to use it.
If you connect to your P4 with the latest version of GCC you will see on the KH Screen something new:
What you need to do is first re calibrate your PH probe, you will see the ADC1 and ADC2 will adjust this is more just to make sure your PH program is calibrating and that when you see the PH read 4 and 7 in the display when your doing your calibration.
Your next going to do a test, you should see your PH go up in Actual Value, and when your Reagent pump beings to dose the internal chamber the PH will go down. You need to pay attention to this, if it does not go down means that not enough reagent is making into the chamber :).