Salinity / Mag Issues (help)

Hallowhead

2500 Club Member
View Badges
Joined
Mar 27, 2019
Messages
3,015
Reaction score
1,476
Location
New Jersey
Rating - 0%
0   0   0
Howdy,

I'm having some sort of issue here.

I noticed my magnesium was dropping daily at excessive rates too which I've done my research indicating it's impossible for my mag to drop that much.

I then ordered an ICP test to confirm that my test kits were not the issue. My Mag / calcium and alk all tested in line with my salifert test kits.

The outlier was salinity which is still causing confusion for me.

I have two refractometers and the Hanna handheld salinity tester. My refractometer is calibrated with juice from BRS and the handheld is calibrated using Hanna tester calibration fluid. All three test kits show 35ppt however the ICP test came back at 31.5....

Someone help lol
 

Pod_01

Valuable Member
View Badges
Joined
Jan 10, 2022
Messages
1,522
Reaction score
1,435
Location
Waterloo
Rating - 0%
0   0   0
From my experience drop in Mg is related to salinity decrease. My experience…

Since you are like me and collect all types of test devices and kits you should add this to your selection:
1739582260161.jpeg


Once I got it I retired my refractometers. :beaming-face-with-smiling-eyes:
 
OP
OP
Hallowhead

Hallowhead

2500 Club Member
View Badges
Joined
Mar 27, 2019
Messages
3,015
Reaction score
1,476
Location
New Jersey
Rating - 0%
0   0   0
From my experience drop in Mg is related to salinity decrease. My experience…

Since you are like me and collect all types of test devices and kits you should add this to your selection:
1739582260161.jpeg


Once I got it I retired my refractometers. :beaming-face-with-smiling-eyes:
lol I was looking at one of these. It works well?
 

Pod_01

Valuable Member
View Badges
Joined
Jan 10, 2022
Messages
1,522
Reaction score
1,435
Location
Waterloo
Rating - 0%
0   0   0
lol I was looking at one of these. It works well?
Yes that is my gold standard, I calibrate my GHL conductivity probe against it. All ICP lab results for salinity are always inline as well so I have high confidence in it.
Top it off it is easy to use and consistent.

After each use just clean it with tap or RO water so nothing is sticking to the glass.
 

UMALUM

Valuable Member
View Badges
Joined
Sep 7, 2023
Messages
1,063
Reaction score
1,344
Location
Fl.
Rating - 0%
0   0   0
Make sure your not testing warm water with cold instruments. Seems odd that 3 devices are showing similar results.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
77,779
Reaction score
77,338
Location
Massachusetts, United States
Rating - 0%
0   0   0
I’m wary of trusting any hobby brand salinity standards. I recommend either a science standard, or better yet, a diy using a decent scale.

 
OP
OP
Hallowhead

Hallowhead

2500 Club Member
View Badges
Joined
Mar 27, 2019
Messages
3,015
Reaction score
1,476
Location
New Jersey
Rating - 0%
0   0   0
Purchased in 1999 and still works like a charm.

IMG_1616.JPG
Grabbed one of these bad boys as a last ditch check. It's coming in at like a low 34ish ppt but not 31.5 like the ICP test. Obviously I'd like to believe the ICP test but now I've had 4 (granted hobby grade) test kits against one lab grade
 

gbroadbridge

2500 Club Member
View Badges
Joined
Mar 25, 2021
Messages
4,957
Reaction score
5,375
Location
Sydney, Australia
Rating - 0%
0   0   0
Grabbed one of these bad boys as a last ditch check. It's coming in at like a low 34ish ppt but not 31.5 like the ICP test. Obviously I'd like to believe the ICP test but now I've had 4 (granted hobby grade) test kits against one lab grade
ICP does not directly measure salinity.

It may calculate salinity based on the elemental composition of the water, or they may be using an EC device.

Best way to check your measurement instrument is by making a DIY standard as Randy stated above.
 
OP
OP
Hallowhead

Hallowhead

2500 Club Member
View Badges
Joined
Mar 27, 2019
Messages
3,015
Reaction score
1,476
Location
New Jersey
Rating - 0%
0   0   0
ICP does not directly measure salinity.

It may calculate salinity based on the elemental composition of the water, or they may be using an EC device.

Best way to check your measurement instrument is by making a DIY standard as Randy stated above.
It's on my list today !! Happy presidents day y'all !
 
OP
OP
Hallowhead

Hallowhead

2500 Club Member
View Badges
Joined
Mar 27, 2019
Messages
3,015
Reaction score
1,476
Location
New Jersey
Rating - 0%
0   0   0
I followed this exactly, however using regular salt vs iodized salt and I can't even calibrate my refractometers to make them show 35pt. Could it be using salt vs iodized ? I used 2 liters exactly measured with a large measuring cup for liquids and used a 1/4 cup used for baking for the salt. I can try again measuring 73.1g instead using a scale I have but I want to confirm salt vs iodized doesn't matter.

That amount roughly corresponds to ¼ cup (73.1 g) of Morton's Iodized Salt dissolved into 2 liters (2000 g) of water (giving very slightly more than 2 L of total volume).
 

Shirak

2500 Club Member
View Badges
Joined
Nov 3, 2019
Messages
2,739
Reaction score
1,582
Location
Thousand Islands, NY
Rating - 0%
0   1   0
regular table salt is fine. Doesn't need to be iodized. If you have a decent electronic scale I would use that. I reduce the recipe and use 3.65g into 100ml(g)

edit I have the decimal in the wrong spot sorry.. correct now
 
OP
OP
Hallowhead

Hallowhead

2500 Club Member
View Badges
Joined
Mar 27, 2019
Messages
3,015
Reaction score
1,476
Location
New Jersey
Rating - 0%
0   0   0
I don't get why my prior mix per Randy's directions resulted in crazy high salinity water though
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
77,779
Reaction score
77,338
Location
Massachusetts, United States
Rating - 0%
0   0   0
I don't get why my prior mix per Randy's directions resulted in crazy high salinity water though

I don’t know what happened, but iodized vs not is insignificant.
 
OP
OP
Hallowhead

Hallowhead

2500 Club Member
View Badges
Joined
Mar 27, 2019
Messages
3,015
Reaction score
1,476
Location
New Jersey
Rating - 0%
0   0   0
Okay so after making two more batches to just confirm, my brs refracto juice appears to be about a point or two high which makes my swing arm about two points low.

My tank is within acceptable range and the ICP test was without a doubt wrong, which is the most concerning. Tank water is 34ppt and ICP claimed 31.5ppt
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
77,779
Reaction score
77,338
Location
Massachusetts, United States
Rating - 0%
0   0   0
Okay so after making two more batches to just confirm, my brs refracto juice appears to be about a point or two high which makes my swing arm about two points low.

My tank is within acceptable range and the ICP test was without a doubt wrong, which is the most concerning. Tank water is 34ppt and ICP claimed 31.5ppt

Sounds good. :)
 

gbroadbridge

2500 Club Member
View Badges
Joined
Mar 25, 2021
Messages
4,957
Reaction score
5,375
Location
Sydney, Australia
Rating - 0%
0   0   0
Okay so after making two more batches to just confirm, my brs refracto juice appears to be about a point or two high which makes my swing arm about two points low.

My tank is within acceptable range and the ICP test was without a doubt wrong, which is the most concerning. Tank water is 34ppt and ICP claimed 31.5ppt
FWIW, the local lab that does the Triton ICP always gives me nonsense values for Salinity (at least 2-3ppt too high)

No idea why, I just ignore it.
 