I wanted to chime in on this conversation because the user’s situation with the Hanna ULR Checker reading 0.43 ppm sounds very familiar to something that happened to me.
My Hanna was also consistently reading MAX (or very high), which worried me a lot. However, when I compared it with a Salifert kit and, more importantly, with an ICP-OES test, both came back showing zero PO₄.
I found out that the cause of this huge discrepancy was a high silicate level in my water. The silicates were interfering with the chemistry of the Hanna ULR Checker and causing the phosphate readings to be falsely high.
My question to the OP (Tom Boyle) and those of you running your tanks at 0.4+:
• Have you tried to measure your silicates?
• If you’re only using the Hanna ULR, have you compared that high reading with another type of kit (like Salifert) or, ideally, an ICP test?