Hello,
I work in an environment that requires precise and accurate measurements of certain variables. This is only possible if the instruments used to take those measurements are calibrated before the project commences and continuously verified during the project.
This brings me to the subject of the ICP test. They have always bothered me because of the above paragraph. I have done a fair amount of reading about these instruments, primarily OES versions. However I am aware of the methodology used for both the OES and MS ICP tests.
They are impressive but in saltwater the matrix is complex, which causes most of the problems with the test (OES).
These can include:
If your salinity, Mg or Ca is slightly off, the corrections can misrepresent minor elements.
My conclusions are for what they are worth:
I work in an environment that requires precise and accurate measurements of certain variables. This is only possible if the instruments used to take those measurements are calibrated before the project commences and continuously verified during the project.
This brings me to the subject of the ICP test. They have always bothered me because of the above paragraph. I have done a fair amount of reading about these instruments, primarily OES versions. However I am aware of the methodology used for both the OES and MS ICP tests.
They are impressive but in saltwater the matrix is complex, which causes most of the problems with the test (OES).
These can include:
- Interference with the detection of trace metals such as vanadium, manganese or zinc.
- Suppression or enhancement of emission lines of target elements.
- Spectral overlaps, for example calcium lines overlapping with strontium or magnesium.
- Elements can read artificially high or low depending on the calibration and software correction.
- Metal from collection containers syringes or even human contact can skew results.
- Some mail-in kits use cheap plastic vials that are not pre-acid-washed and can leach elements.
- Cross-contamination in laboratories that run hundreds of samples a day is also possible.
- Certain ions (e.g. Fe, Al, Mn or even PO₄ bound to CaCO₃ particles) can adhere to particles or precipitate in the sample bottle before it reaches the laboratory.
- This means the true dissolved concentration may be lower than what the laboratory measures (if the particles are digested) or higher if some are lost.
- Bacteria can alter redox states (Fe²⁺ → Fe³⁺).
- Precipitates can form with carbonate or phosphate.
- Some laboratories require acidification of samples, others do not – this affects stability.
If your salinity, Mg or Ca is slightly off, the corrections can misrepresent minor elements.
My conclusions are for what they are worth:
- You can have high confidence in Ca, Mg, Na, K, Sr, B, Br, Li measurements.
- Use to track for trends in Fe, Zn, Cu, Mo, Mn, I, Al, Si.
- Ignore absolute numbers for P, Cr, Ti, heavy metals (unless alarmingly high).
- Always compare changes over time, not single snapshots.